CPAP use on overnight flights

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Stan in Seattle
Posts: 1
Joined: Fri Oct 28, 2011 1:41 pm

CPAP use on overnight flights

Post by Stan in Seattle » Fri Oct 28, 2011 1:51 pm

I have a REMstar Auto A-Flex and would like to use it on an upcoming overnight flight. My machine is not battery operated. According to the British Airways site: "Maximum power output from the laptop points is 75 Watts and, if your equipment draws more than this, the outlet will be automatically deactivated".

I am having a problem determining the wattage for my machine. Presumably, it depends on the use of the heater. I would be happy to turn the heat off or down very low for the duration of the flight. Any advice would be appreciated.

Thanks
Stan

davelikesbeer
Posts: 250
Joined: Tue Jun 21, 2011 11:45 pm
Location: California

Re: CPAP use on overnight flights

Post by davelikesbeer » Fri Oct 28, 2011 2:25 pm

Stan in Seattle wrote:I have a REMstar Auto A-Flex and would like to use it on an upcoming overnight flight. My machine is not battery operated. According to the British Airways site: "Maximum power output from the laptop points is 75 Watts and, if your equipment draws more than this, the outlet will be automatically deactivated".

I am having a problem determining the wattage for my machine. Presumably, it depends on the use of the heater. I would be happy to turn the heat off or down very low for the duration of the flight. Any advice would be appreciated.

Thanks
Stan
I don't have any experience with this at all, but I'm curious where you would put your APAP. You may want to research BA's web site for specific information about CPAP inflight use.

Looking at the Specifications section in http://respironicsremstars.respironics. ... Manual.pdf it says 12 VDC * 5.0 A which would be 60 watts maximum if I'm not mistaken (someone more knowledgeable about electricity please correct me). I can't imagine using your APAP with humidifier set to 0 or not connected would be greater than 75 watts.
CPAP for the rest of your life.

Guest

Re: CPAP use on overnight flights

Post by Guest » Fri Oct 28, 2011 3:37 pm

Thanks for the quick reply, dave. Fortunately, I am flying in business class, so there will be room for the machine. I think your calculation is correct. I called my local dealer and they came up with a much larger number.

BA does allow the use of a CPAP as long as it does not exceed 75W (which would trip the power outlet). They do not require prior approval, which other airlines do.

From the BA website:
CPAP machine

Medical clearance is not required for the carriage or use of CPAP machines (used in the treatment of sleep apnoea) as fitness to travel will not be in doubt.
CPAP machines can be plugged into a laptop power-point, where available, with a suitable adaptor.
Maximum power output from the laptop points is 75 Watts and, if your equipment draws more than this, the outlet will be automatically deactivated.
We cannot guarantee that any of the laptop points will be serviceable. If you intend to use one in flight, we recommend that you use a dry-cell battery-operated device.


Stan

User avatar
billbolton
Posts: 2264
Joined: Wed Jun 07, 2006 7:46 pm
Location: Sydney, Australia

Re: CPAP use on overnight flights

Post by billbolton » Fri Oct 28, 2011 5:09 pm

Stan in Seattle wrote:Presumably, it depends on the use of the heater. I would be happy to turn the heat off or down very low for the duration of the flight.
You definitely do need to turn the humidifier off but you may also exceeed the 75W power limit if you are running your CPAP at higher pressures. I know from my own experience with BA, using a Resmed S8 this was anything much above about 12cms H20 with the altitude setting adjusted to cabin pressure (~2400 metres altitude).

Cheers,

Bill

_________________
MachineMask
Additional Comments: Airmini, Medistrom Pilot 24, CMS 60C Pulse Oximeter, ResScan 6

User avatar
LinkC
Posts: 3154
Joined: Wed Oct 29, 2008 1:06 pm
Location: Amelia Island, FL

Re: CPAP use on overnight flights

Post by LinkC » Fri Oct 28, 2011 8:18 pm

Many of us who travel a lot find it's more hassle to use CPAP, and not worth it. Sleeping in a semi upright position usually reduces apneas.

_________________
MachineMaskHumidifier
Additional Comments: 11-14 cmH2O
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: CPAP use on overnight flights

Post by idamtnboy » Fri Oct 28, 2011 10:05 pm

Guest wrote:
CPAP machine

CPAP machines can be plugged into a laptop power-point, where available, with a suitable adaptor.
Stan
From what I have seen in the past when buying a power adapter for my laptop the power point in an airplane is not the same as in automobiles. Do you have, or can you get, a power adapter for your CPAP that will plug into an airline power point?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
islandboy5150
Posts: 301
Joined: Fri Oct 15, 2010 9:14 pm
Location: Temecula, California
Contact:

Re: CPAP use on overnight flights

Post by islandboy5150 » Sat Oct 29, 2011 1:12 am

Linc is right on the money, CPAP is just too much a PITA to use in flight. Hose getting trampled, finding a place to set it up where it won't get kicked, knocked off, knocked over, plus the power cord getting all tangled and hung up on everything anndbecoming a trip hazard, then your really up a creek if you need to get up to go to the lavatory. You have to unhook, try to put it all away enough that you don't drag the unit down the isle with you. It's just a real PITA on a plane. Not worth doing anything but sliding it under the seat in front of you in it's case. Don't try putting it up in the overhead either, people don't respect others belongings and will cram/smash the shit out of it to get their crazy big wheelie bag in with it. I've even had one lady pull it out and let it drop and fall on the isle floor to make room for her crap. That's when I learned to just put the darn thing under the seat in front of me to keep an eye on it and keep it protected. It's too valuable to do anything else in my opinion.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: PAP-CAP for Headgear, Hose Cozy and Hose Boss

User avatar
cnew
Posts: 39
Joined: Tue Sep 13, 2011 8:15 am

Re: CPAP use on overnight flights

Post by cnew » Sat Oct 29, 2011 6:52 am

I'm planning a trip to Italy over the holidays, and am reading this thread with interest. I hadn't planned to try to use my CPAP it in flight, and am reassured to know that sleeping upright may help mitigate apneas.

What I am wondering about is getting to carry it onboard uncounted as my allowed carry-on if I have it in a bag with my netbook. I really consider the netbook part of my medical equipment in the sense that it will allow me to continue monitoring my AHI while I am away, but the airline might not see it that way. I plan to have my roll-on bag, my CPAP bag and my purse. Any thoughts about this?

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Started 9/15/2011, Using Sleepyhead open source software, PR Premium Chinstrap.
The little cutie I use for my avatar is orphan squirrel I had in rehab. Some day, when my OSA is under control, I will be bright-eyed and bushy-tailed, too. :-)

User avatar
GumbyCT
Posts: 5778
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: CPAP use on overnight flights

Post by GumbyCT » Sat Oct 29, 2011 7:42 am

1st. I would say that your DC cable comes w/5A fuse so it's not likely the machine w/HH will draw more than that. Always keep (n carry) spare fuses to fit your DC cable. I bought mine at an auto parts store.

2. Do not even attempt to use water in the humidifier in flight as that can ruin your equipment at a very bad time for you.

3. IF you ever do spill water into your machine simply do NOT apply power to it until absolutely, totally dry inside, even if that is 3 days or more. IF you do you will likely burn it up and it may take weeks to get it replaced, so 3 days will seem like nothing.

4. Always try your DC cable and hook up BEFORE you need your machine to make sure you have everything and know how the pieces go together and that they work. Even if that means going out to the car to see if the machine powers up correctly for at least 3-5 minutes.

5. Travel with a spare or backup machine if you can. Should this machine get stolen or damaged at least you will have another machine. I read a post on here where a cpap was stolen from a rental car. Presumably someone thought it was a laptop or something. Who'da thunk it?

6. Always, Always travel with a copy of your prescription. IF you have ignored any of the above and need to buy a mask or machine while away from home you WILL need a prescription. Even dealing with a national company the person you encounter probably won't be able to access any of your info because of company security.

As a backup you can always upload your prescription to cpap.com and buy from them if/when needed, also they can ship overnite. Just be sure you aren't traveling to a country they can't ship to.

HTH

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

User avatar
idamtnboy
Posts: 2186
Joined: Mon Nov 01, 2010 2:12 pm
Location: Idaho

Re: CPAP use on overnight flights

Post by idamtnboy » Sat Oct 29, 2011 7:54 am

cnew wrote:I'm planning a trip to Italy over the holidays, and am reading this thread with interest. I hadn't planned to try to use my CPAP it in flight, and am reassured to know that sleeping upright may help mitigate apneas.

What I am wondering about is getting to carry it onboard uncounted as my allowed carry-on if I have it in a bag with my netbook. I really consider the netbook part of my medical equipment in the sense that it will allow me to continue monitoring my AHI while I am away, but the airline might not see it that way. I plan to have my roll-on bag, my CPAP bag and my purse. Any thoughts about this?
Do a search on CPAP and airlines. This has been discussed quite a bit during the past year. The CPAP will not be counted as a carry on, but it must not have anything in the bag except the direct CPAP equipment.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Hose management - rubber band tied to casement window crank handle! Hey, it works! S/W is 3.13, not 3.7

User avatar
GumbyCT
Posts: 5778
Joined: Fri Sep 14, 2007 6:22 pm
Location: CT
Contact:

Re: CPAP use on overnight flights

Post by GumbyCT » Sat Oct 29, 2011 8:02 am

idamtnboy wrote:Do a search on CPAP and airlines. This has been discussed quite a bit during the past year. The CPAP will not be counted as a carry on, but it must not have anything in the bag except the direct CPAP equipment.
Combine with medicines which are also exempt. Computers are not exempt. But then ALL of this varies with who you encounter and how well they slept lastnite.

_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: New users can't remember they can't remember YET!
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember
;)
If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!

AnotherEngineer
Posts: 1
Joined: Mon Nov 14, 2011 7:50 am

Re: CPAP use on overnight flights

Post by AnotherEngineer » Mon Nov 14, 2011 7:57 am

I shall be flying back from Tornoto or Montrreal to UK next year on BA so a Boeing 777.

The use of my Remstar is allowed but I am unsure whether it can be connected to the 15v DC supply.

Has anyone else done this successfully or can a voltage reducer be bought anywhere.

Thanks,

mark

User avatar
araminta
Posts: 392
Joined: Wed Jan 21, 2009 7:30 pm
Location: USA

Re: CPAP use on overnight flights

Post by araminta » Fri Nov 18, 2011 11:33 pm

i just did a long haul flight and found that the angle of the seat allowed me to sleep adequately and not have apneas. Fortunately i was able to get a flight that arrived in the evening so I could just get a bite to eat and crawl into bed.

User avatar
JointPain
Posts: 261
Joined: Thu Apr 28, 2011 10:41 am
Location: Houston, TX

Re: CPAP use on overnight flights

Post by JointPain » Sat Nov 19, 2011 12:19 am

idamtnboy wrote:
cnew wrote:I'm planning a trip to Italy over the holidays, and am reading this thread with interest. I hadn't planned to try to use my CPAP it in flight, and am reassured to know that sleeping upright may help mitigate apneas.

What I am wondering about is getting to carry it onboard uncounted as my allowed carry-on if I have it in a bag with my netbook. I really consider the netbook part of my medical equipment in the sense that it will allow me to continue monitoring my AHI while I am away, but the airline might not see it that way. I plan to have my roll-on bag, my CPAP bag and my purse. Any thoughts about this?
Do a search on CPAP and airlines. This has been discussed quite a bit during the past year. The CPAP will not be counted as a carry on, but it must not have anything in the bag except the direct CPAP equipment.
These are the rules for within the USA and for to/from the USA. Within Europe these rules don't apply. A CPAP bag is just a bag. When I changed planes in Frankfurt last year, we were supposed to have only one carry on. I managed to carry on my CPAP plus one other small bag.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure is 11-14. Old CPAP was a Resmed S6 Lightweight. Also have Profile lite mask. ResScan is actually version 3.14. Now I use Sleepyhead.