Ok, so what *else* causes hose rain?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
AlanMacNeill
Posts: 29
Joined: Wed Sep 14, 2011 12:10 pm

Ok, so what *else* causes hose rain?

Post by AlanMacNeill » Tue Sep 27, 2011 8:45 am

This is weird...

I got my new machine about 11 days ago, and did my first weekly intensive cleaning this past Sunday, as per instructions (well...okay, not completely per instructions, as that was day 9 not 7, but my routine is weekly maintenance of stuff, not just the CPAP, is Sunday, so I figured I could slide a couple of days to get it onto the routine moving forward).

The humidifier chamber was filled with the 1:10 vinger:water solution for 10 minutes, emptied and let air dry inside (paper towel wipe off outside), the mask was broken down and all components washed with gentle dish soap and warm water, the hose was washed inside and out with gentle dish soap and water (because this machine has a heated hose, the instructions say quite dramatically NOT to use vinegar in the hose), then let hang dry for about half an hour before reassembly.

My humidifier is set at 80 degrees.

Sunday night I had significant rain in the tube, I mean we're talking hurricane hitting landfall levels of moisture coming out of the tube.

Same thing last night. I ultimately had to turn the humidifier temp down to 70 and literally take apart the mask and tube again and shake the living c**p out of the tube to get all the water out, then no rain from there, but I did have massive mask fit issues (despite not changing the headgear sizing at all that I'm aware of)

Everything I'm reading says that I've done my job correctly, and that the copper wire and sensors in the tube are supposed to prevent rain...

What should I take a gander at to figure out the cause? The 70 degree air is a little too chilly for my comfort, but rain is worse.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: (New machine 9/15)

Axxel
Posts: 76
Joined: Wed Sep 21, 2011 5:48 am

Re: Ok, so what *else* causes hose rain?

Post by Axxel » Tue Sep 27, 2011 9:15 am

I actually started not adding water to tank and turning humidifier off so it doesnt damage the machine. I like it this way. Air is humid here already so my nose doesnt get dry at all and no moisture or rain problems.

User avatar
Pugsy
Posts: 65302
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Ok, so what *else* causes hose rain?

Post by Pugsy » Tue Sep 27, 2011 9:25 am

I have read where someone has filled their tank to the desired full level and if they turn the machine on full blast and not be masked up that some of the water from the tank gets blown into the hose. Seems to depend on pressures used too.
Might experiment with having mask on before turning on machine or using slightly less water in the tank.

This problem has been mentioned here on the forum by others.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
Last edited by Pugsy on Tue Sep 27, 2011 9:45 am, edited 1 time in total.
I may have to RISE but I refuse to SHINE.

User avatar
xenablue
Posts: 1358
Joined: Tue Sep 07, 2010 12:52 pm
Location: Aussie living in balmy Wisconsin

Re: Ok, so what *else* causes hose rain?

Post by xenablue » Tue Sep 27, 2011 9:37 am

Sure you didn't still have a bunch of water in the hose after cleaning?

I hang mine on an "s" hook in front of a kitchen window, and I can still see water droplets in it several days later. In fact if I wash my gear on Saturday, it's usually around Wednesday evening before I see the hose is completely dry.

I cycle the hoses I have so I'm always using a clean, completely dry one. I guess you could run the machine with the wet hose on it and let it blow until dry if you don't have a spare.

Cheers,
xena

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Dx 10/14/10. Also a T2 diabetic. High night/fasting numbers prompted a sleep study and here I am :-)

User avatar
Bobby269
Posts: 803
Joined: Sat Nov 27, 2010 11:12 am
Location: San Antonio Texas

Re: Ok, so what *else* causes hose rain?

Post by Bobby269 » Tue Sep 27, 2011 9:50 am

Do you use a hose holder? http://www.amazon.com/s/ref=nb_sb_noss? ... ps&x=0&y=0.
Take a look at this information. http://www.apneasupport.org/about790.html
Hope that this helps.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Set to Apap 14.00/20. EPR 3, Climate line hose, Oscar for Mac 1.1.0. Betta.1

User avatar
apneawho
Posts: 609
Joined: Wed Jul 27, 2011 10:09 am
Location: Atlanta, GA, USA

Re: Ok, so what *else* causes hose rain?

Post by apneawho » Tue Sep 27, 2011 9:55 am

This was one of the threads from yesterday. Maybe these adjustments would help you.
If you are using the Climateline, there is a feature where you can go into settings and change the ability to regulate the temp of the hose vs the humidity. Where they are two separate functions to adjust.

http://www.cpaplibrary.com/machines.html

It'll show you different features on how it works.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed Climate line tubing, Hose cozy, PurSleep Clear aroma therapy, Using latest version Sleepyhead sofware-thanks Jedimark!

User avatar
thunderhand
Posts: 19
Joined: Sun Feb 13, 2011 8:22 am

Re: Ok, so what *else* causes hose rain?

Post by thunderhand » Tue Sep 27, 2011 12:14 pm

I have never had an issue with moisture in the tube..

don't all machines continue to run for a bit after shutting off to assist with drying out the tube ??

Is this caused by overfilling the tank or using a high pressure with a full tank ??

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS 50E pulse oximeter

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Ok, so what *else* causes hose rain?

Post by Janknitz » Tue Sep 27, 2011 12:44 pm

I never wash my hose! It seems nearly impossible to get all the moisture out, and I figure I'm in more danger from mold growing in the trapped water in the hose than the clean air blowing through it. And I have used the same hose forever since Crapria doesn't seem to understand what the heck a PERFORMANCE hose is--I don't want to have to fight the kinks in the crappy cheap hose they insist on sending out instead.

I haven't died yet--respiratory system is doing pretty well, actually.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
Pugsy
Posts: 65302
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Ok, so what *else* causes hose rain?

Post by Pugsy » Tue Sep 27, 2011 12:49 pm

thunderhand wrote: don't all machines continue to run for a bit after shutting off to assist with drying out the tube ??
No, only the ResMed machine.
thunderhand wrote:Is this caused by overfilling the tank or using a high pressure with a full tank ??
The other users who have had this problem stated that they were certain the tank was not overfilled.
Since it is not easily duplicated we aren't for sure if it was high pressure and a full tank.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Perrybucsdad
Posts: 834
Joined: Mon Sep 12, 2011 7:09 am
Location: Northeast Ohio

Re: Ok, so what *else* causes hose rain?

Post by Perrybucsdad » Tue Sep 27, 2011 1:08 pm

Pugsy wrote:I have read where someone has filled their tank to the desired full level and if they turn the machine on full blast and not be masked up that some of the water from the tank gets blown into the hose. Seems to depend on pressures used too.
Might experiment with having mask on before turning on machine or using slightly less water in the tank.

This problem has been mentioned here on the forum by others.
Pugsy, I have seen this first hand. Yes, if you fill your tank to full or if you turn the unit on and don't put your mask on right away, water will blow into the hose (at least on my S9 AutoSet model).

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ResScan v5.9; Sleepyhead v1.0.0-beta

User avatar
Pugsy
Posts: 65302
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Ok, so what *else* causes hose rain?

Post by Pugsy » Tue Sep 27, 2011 1:34 pm

Perrybucsdad wrote: Pugsy, I have seen this first hand. Yes, if you fill your tank to full or if you turn the unit on and don't put your mask on right away, water will blow into the hose (at least on my S9 AutoSet model)
Thank you for confirming this. It should be easy for someone to troubleshoot to see if this is the cause for copious amounts of water in the hose for Resmed S9 users. I have never seen it reported by PR S1 users and I have not seen this myself but then I mask up and then allow the machine to come on.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Perrybucsdad
Posts: 834
Joined: Mon Sep 12, 2011 7:09 am
Location: Northeast Ohio

Re: Ok, so what *else* causes hose rain?

Post by Perrybucsdad » Tue Sep 27, 2011 1:45 pm

Troublesome kids pouring water in the hose can be a problem too. I've seen that first hand too.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: ResScan v5.9; Sleepyhead v1.0.0-beta

greg-g
Posts: 219
Joined: Sun Dec 20, 2009 1:07 am
Location: Tasmania Australia

Re: Ok, so what *else* causes hose rain?

Post by greg-g » Tue Sep 27, 2011 5:44 pm

Back to first principals.
If everything is connected correctly and working, I don't think you can get rainout in a ResMed S9 heated hose.
What are the possible explanations.
Is the hose connected correctly to the machine. Push it on and rotate 90 deg. I suspect this is correct as otherwise you get a different screen display.
It is possible to get rainout in some masks. If this is the case the mask will have water but the hose will be dry. I have had problems with my Swift Fx.
It is possible as discussed above to blow water into the hose if you have a massive leak (Machine on mask off)
Personally as my hose is always dry I only wash it very occasionally (every 6 months or so). It would need washing weekly though if wet.
The optional humidifier chamber can be taken apart and is much easier to clean.
I run my temperature at 78 deg and have never had an issue
If all the above are OK I would suspect a problem with the hose or machine.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: CMS-50F, Night Vision camera, Hose hanger, ResScan 3.16, Modified Swift FX with head band and air diffuser.