Sleepyhead issues - Getting "Unable to load machine data"

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
greatnitz
Posts: 11
Joined: Tue Sep 20, 2011 12:10 pm

Sleepyhead issues - Getting "Unable to load machine data"

Post by greatnitz » Fri Sep 23, 2011 4:49 pm

Hello all,

Hopefully there is somebody out there fluent in the Sleepyhead software. Just got a remstar PRO w/ c-flex machine and actually only used it one night. I do see data in the SD card (path is D:\P-Series\P040894876962\p0\), I have some files such as '0000000004' which is 110 kb. However, when I attempt to import it, I browse to the p0 folder and click 'Select folder' I get an error saying that it couldn't find any machine data. Anybody know if I'm doing something wrong?

Thanks,

Darren

User avatar
Breathe Jimbo
Posts: 954
Joined: Sun Sep 12, 2010 3:02 pm
Location: Pasadena, CA

Re: Sleepyhead issues - Getting "Unable to load machine data"

Post by Breathe Jimbo » Fri Sep 23, 2011 5:22 pm

Try loading at the P-Series folder level.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Since 9/9/10; 13 cm; ResScan 3.16; SleepyHead 0.9; PapCap

greatnitz
Posts: 11
Joined: Tue Sep 20, 2011 12:10 pm

Re: Sleepyhead issues - Getting "Unable to load machine data"

Post by greatnitz » Fri Sep 23, 2011 5:27 pm

Just tried selecting the 'P-Series' folder, still says "could not find any machine data". Is there anything on the machine I need to do to capture data to the SD card?

User avatar
Pugsy
Posts: 65278
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead issues - Getting "Unable to load machine data"

Post by Pugsy » Fri Sep 23, 2011 5:31 pm

Select the drive where the SD card is as the folder.
If that doesn't work, post back and I will get screen shot for you.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Pugsy
Posts: 65278
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead issues - Getting "Unable to load machine data"

Post by Pugsy » Fri Sep 23, 2011 5:39 pm

For version 0.8.5 I have Windows 7 Home Premium

Don't go any farther than this to select the folder. Next step in is P Series folder and it won't work with latest SH version.
Image

For older versions (0.8.1 and older) of SleepyHead the P series folder works but not with 8.5 and 8.3 I forget which for 8.1.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
Last edited by Pugsy on Fri Sep 23, 2011 5:42 pm, edited 1 time in total.
I may have to RISE but I refuse to SHINE.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Sleepyhead issues - Getting "Unable to load machine data"

Post by archangle » Fri Sep 23, 2011 5:40 pm

Breathe Jimbo wrote:Try loading at the P-Series folder level.
Go up one directory from the P series folder.

Or go into the directory that contains "P-series," but do NOT click on P-series to highlight it. Just click on "Select Folder"

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

greatnitz
Posts: 11
Joined: Tue Sep 20, 2011 12:10 pm

Re: Sleepyhead issues - Getting "Unable to load machine data"

Post by greatnitz » Fri Sep 23, 2011 7:44 pm

That worked!, thank you! Now I just have to figure out what all this means

LSekunda
Posts: 2
Joined: Sat May 12, 2012 9:55 pm

Re: Sleepyhead issues - Getting "Unable to load machine data"

Post by LSekunda » Sat May 12, 2012 10:01 pm

Sleepyhead V0.9.2-1 is giving me the error "couldn't find any machine data"

Windows 7 x64 machine

I have tried every directory I can from root of the memory card down to the data files; copied memory card to hard drive with same results. Also tried "Just point to the root drive letter all by itself. Then click on select folder button." as suggested in this forum.

Not sure what will help anyone, but here is the information I can see:

Data file structure is listed below:
E:\
E:\P-Series
E:\P-Series\C022986650B2F
E:\P-Series\C022986650B2F\e
E:\P-Series\C022986650B2F\e\1.004 and so on through 16.004
E:\P-Series\C022986650B2F\p0
E:\P-Series\C022986650B2F\p0\0000000110.001 and so on through 0000000266.002
E:\P-Series\C022986650B2F\properties.txt
E:\P-Series\Clear0C022986650B2F
E:\P-Series\Clear0C022986650B2F\e
E:\P-Series\Clear0C022986650B2F\e\1.004 and so on through 16.004
E:\P-Series\Clear0C022986650B2F\p0
E:\P-Series\Clear0C022986650B2F\p0\0000000110.001 and so on through 0000000109.002
E:\P-Series\Clear0C022986650B2F\encore.txt
E:\P-Series\Clear0C022986650B2F\properties.txt
E:\P-Series\encore.txt
E:\P-Series\last.txt

This is the data that is in the E:\P-Series\C022986650B2F\properties.txt file:
SerialNumber=C022986650B2F
ModelNumber=1061401
ProductType=0x40
DataFormat=UDF
DataFormatVersion=2
Family=Ventilator
FamilyVersion=0
SoftwareVersion= 2.03
FirstDate=1324075838
LastDate=1336867200
ErrorKey=003f04009a4a
PatientFolderNum=0
PatientFileNum=312
EquipFileNum=16
SupportedLogs=1,2,4,6
ValidCheck=0b7b

This is the data that is in the E:\P-Series\Clear0C022986650B2F\properties.txt file:
SerialNumber=C022986650B2F
ModelNumber=1061401
ProductType=0x40
DataFormat=UDF
DataFormatVersion=2
Family=Ventilator
FamilyVersion=0
SoftwareVersion= 2.03
FirstDate=1289567195
LastDate=1324039324
ErrorKey=001e040009a2
PatientFolderNum=0
PatientFileNum=213
EquipFileNum=16
SupportedLogs=1,2,4,6
ValidCheck=3302

encore.txt files have identical content:
<?xml version="1.0" encoding="utf-16"?><encore ver="1.0"><guid>82149337-5a5b-4c10-a490-05e1fb327658</guid></encore>

Any ideas? Does anyone know if Sleepyhead copies this data and if so where? Is there a manual import procedure? From what i have read the file type and structure appears to be correct, just Sleepyhead does not recognize it.

User avatar
Pugsy
Posts: 65278
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Sleepyhead issues - Getting "Unable to load machine data"

Post by Pugsy » Sun May 13, 2012 7:39 am

LSekunda wrote:Any ideas? Does anyone know if Sleepyhead copies this data and if so where? Is there a manual import procedure? From what i have read the file type and structure appears to be correct, just Sleepyhead does not recognize it.
What model Respironics machine are you using?
For the Beta version the path goes only to the drive letter where the SD card is sitting. So when you tried only the drive letter....that should have worked. In the SD card and in the P0 folder there should have been 3 file types that are the data files .001 .002 and .005 files. The .005 files are the wave form files.

Wonder what would show with Encore? Give me a couple of minutes and check your private message box.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

LSekunda
Posts: 2
Joined: Sat May 12, 2012 9:55 pm

Re: Sleepyhead issues - Getting "Unable to load machine data"

Post by LSekunda » Mon May 14, 2012 6:54 pm

LSekunda wrote:Sleepyhead V0.9.2-1 is giving me the error "couldn't find any machine data"
Just a quick update:

Big thanks to Pugsy for his help.

The CPAP machine btw is a Philips Respironics BiPAP S/T C-Series

My wife was taking a nap when I first posted, so when she woke up I tried her card from her System One RemStar Pro with C-Flex+ (450P) machine, Sleepyhead read her card just fine, so not likely the install or computer. Assuming at this point that the Sleepyhead software just does not recognize the BiPap ST machine.

I can read my data with Encore Basic; but looking at my wife's data, Sleepyhead appears to provide better detail. Submitted my info to JediMark, so hopefully sometime in the future.

For now, this appears to be a model support issue; when I have some time I am going to see if I can figure a way to get the data to load. If I figure something out I will update again.