Do I need higher pressure if my AHI is still high?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
snoozebabe
Posts: 11
Joined: Tue May 03, 2011 6:48 pm

Do I need higher pressure if my AHI is still high?

Post by snoozebabe » Wed Aug 31, 2011 4:41 am

I have been using my CPAP for 5 months now and the AHI is averaging 10.5 / 11. This is still higher than the recommended range.

I have no problem with mask leaks. The leakage is 0%.

Does it mean I need to get my upper limit pressure adjusted even higher (currently at 14)?
And if tonight my soul may find her peace
in sleep, and sink in good oblivion,
and in the morning wake like a new-opened flower
then I have been dipped again in God, and new-created.
~D.H. Lawrence

User avatar
bdp522
Posts: 3378
Joined: Tue Jun 27, 2006 5:13 pm

Re: Do I need higher pressure if my AHI is still high?

Post by bdp522 » Wed Aug 31, 2011 5:39 am

Do you have the software so you can see your charts? What is the machines pressure set to? If you use it in auto mode and the low pressure is too low, you will have events until the pressure gets up to where it will stent the airway open. You may just need to raise the lower setting. If you don't have issues with runaway pressure or chasing snores, there is no reason to limit the upper pressure. The machine shouldn't go higher than needed. It is the lower pressure setting that seems to cause the most trouble.

Brenda

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Love my papillow, Aussie heated hose and PAD-A-CHEEKS! Also use Optilife, UMFF(with PADACHEEK gasket), and Headrest masks Pressure; 10.5

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Do I need higher pressure if my AHI is still high?

Post by avi123 » Wed Aug 31, 2011 6:29 am

Check this:

THE APNEA-HYPOPNEA INDEX: USEFUL OR USELESS?

Link:

http://www.respiratoryreviews.com/sep02 ... Index.html

“There is not much association between the AHI and anything else—sleepiness, muscle dysfunction,” or other markers for SDB, claimed Dr. Sullivan, who heads the Sleep Disorders Unit at the University of Sydney in Australia. The management of SDB should hinge on the history, examination, and clinical judgment," he asserted.

Not to cause other problems by raising the pressure it needs be done by someone who knows about it.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
BlackSpinner
Posts: 9742
Joined: Sat Apr 25, 2009 5:44 pm
Location: Edmonton Alberta
Contact:

Re: Do I need higher pressure if my AHI is still high?

Post by BlackSpinner » Wed Aug 31, 2011 6:47 am

avi123 wrote:Check this:

THE APNEA-HYPOPNEA INDEX: USEFUL OR USELESS?

“There is not much association between the AHI and anything else—sleepiness, muscle dysfunction,” or other markers for SDB, claimed Dr. Sullivan, who heads the Sleep Disorders Unit at the University of Sydney in Australia. The management of SDB should hinge on the history, examination, and clinical judgment," he asserted.

Not to cause other problems by raising the pressure it needs be done by someone who knows about it.
Ignore the man behind the curtain.
From the same article:
The AHI can mislead physicians about the severity of SDB, Dr. Sullivan added. At certain points in the menstrual cycle, for example, women with SDB may respond to apnea with a large rise in blood pressure (BP) rather than in the AHI.

AHI values correlate with symptoms of SDB. “[The correlation] is mediocre perhaps, but it is not absent,” stated Dr. Rapoport, countering Dr. Sullivan’s earlier assertion. Because the AHI is imperfect, Dr. Rapoport views it as a marker for the apnea-hypopnea syndrome and not as a definitive metric. He has found the AHI most useful for detecting severe apnea-hypopnea syndrome.
10 is too high. You need to adjust your pressure - especially the lower end. Do one small adjustment a week and keep a close eye on your AHI

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Additional Comments: Quatro mask for colds & flus S8 elite for back up
71. The lame can ride on horseback, the one-handed drive cattle. The deaf, fight and be useful. To be blind is better than to be burnt on the pyre. No one gets good from a corpse. The Havamal

snoozebabe
Posts: 11
Joined: Tue May 03, 2011 6:48 pm

Re: Do I need higher pressure if my AHI is still high?

Post by snoozebabe » Wed Aug 31, 2011 7:05 am

bdp522 wrote:Do you have the software so you can see your charts? What is the machines pressure set to? If you use it in auto mode and the low pressure is too low, you will have events until the pressure gets up to where it will stent the airway open. You may just need to raise the lower setting. If you don't have issues with runaway pressure or chasing snores, there is no reason to limit the upper pressure. The machine shouldn't go higher than needed. It is the lower pressure setting that seems to cause the most trouble.

Brenda

Hi Brenda, I don't have the software but I use the online SmartCode to generate the reports. The pressure for my machine is 6 to 14. Thanks for your advice!
And if tonight my soul may find her peace
in sleep, and sink in good oblivion,
and in the morning wake like a new-opened flower
then I have been dipped again in God, and new-created.
~D.H. Lawrence

snoozebabe
Posts: 11
Joined: Tue May 03, 2011 6:48 pm

Re: Do I need higher pressure if my AHI is still high?

Post by snoozebabe » Wed Aug 31, 2011 7:11 am

BlackSpinner wrote:
avi123 wrote:Check this:

THE APNEA-HYPOPNEA INDEX: USEFUL OR USELESS?

“There is not much association between the AHI and anything else—sleepiness, muscle dysfunction,” or other markers for SDB, claimed Dr. Sullivan, who heads the Sleep Disorders Unit at the University of Sydney in Australia. The management of SDB should hinge on the history, examination, and clinical judgment," he asserted.

Not to cause other problems by raising the pressure it needs be done by someone who knows about it.
Ignore the man behind the curtain.
From the same article:
The AHI can mislead physicians about the severity of SDB, Dr. Sullivan added. At certain points in the menstrual cycle, for example, women with SDB may respond to apnea with a large rise in blood pressure (BP) rather than in the AHI.

AHI values correlate with symptoms of SDB. “[The correlation] is mediocre perhaps, but it is not absent,” stated Dr. Rapoport, countering Dr. Sullivan’s earlier assertion. Because the AHI is imperfect, Dr. Rapoport views it as a marker for the apnea-hypopnea syndrome and not as a definitive metric. He has found the AHI most useful for detecting severe apnea-hypopnea syndrome.
10 is too high. You need to adjust your pressure - especially the lower end. Do one small adjustment a week and keep a close eye on your AHI
Thank you both for the helpful advice. Do you know whether it is possible for me to adjust the pressure myself on my machine?
And if tonight my soul may find her peace
in sleep, and sink in good oblivion,
and in the morning wake like a new-opened flower
then I have been dipped again in God, and new-created.
~D.H. Lawrence

snoozebabe
Posts: 11
Joined: Tue May 03, 2011 6:48 pm

Re: Do I need higher pressure if my AHI is still high?

Post by snoozebabe » Wed Aug 31, 2011 7:19 am

snoozebabe wrote:
BlackSpinner wrote:
avi123 wrote:Check this:

THE APNEA-HYPOPNEA INDEX: USEFUL OR USELESS?

“There is not much association between the AHI and anything else—sleepiness, muscle dysfunction,” or other markers for SDB, claimed Dr. Sullivan, who heads the Sleep Disorders Unit at the University of Sydney in Australia. The management of SDB should hinge on the history, examination, and clinical judgment," he asserted.

Not to cause other problems by raising the pressure it needs be done by someone who knows about it.
Ignore the man behind the curtain.
From the same article:
The AHI can mislead physicians about the severity of SDB, Dr. Sullivan added. At certain points in the menstrual cycle, for example, women with SDB may respond to apnea with a large rise in blood pressure (BP) rather than in the AHI.

AHI values correlate with symptoms of SDB. “[The correlation] is mediocre perhaps, but it is not absent,” stated Dr. Rapoport, countering Dr. Sullivan’s earlier assertion. Because the AHI is imperfect, Dr. Rapoport views it as a marker for the apnea-hypopnea syndrome and not as a definitive metric. He has found the AHI most useful for detecting severe apnea-hypopnea syndrome.
10 is too high. You need to adjust your pressure - especially the lower end. Do one small adjustment a week and keep a close eye on your AHI
Thank you both for the helpful advice. Do you know whether it is possible for me to adjust the pressure myself on my machine?
It's ok - I found the link to the manuals that show me the steps to adjust my pressure setting. Thanks everyone!
And if tonight my soul may find her peace
in sleep, and sink in good oblivion,
and in the morning wake like a new-opened flower
then I have been dipped again in God, and new-created.
~D.H. Lawrence

DaveLP
Posts: 133
Joined: Sat Jul 23, 2011 6:02 pm

Re: Do I need higher pressure if my AHI is still high?

Post by DaveLP » Wed Aug 31, 2011 7:43 am

I would get the software if you don't have it, and view the graphs to determine when events are occurring and what is happening. AHI is a combination of apnea and hypopnea. You need to see how much of each is being recorded. Apnea is an event where you may have a blocked airway or under 10% air flow. Hypopnea is a condition where air flow is 50% of normal or less.

Leaks and mouth breathing are two different things. EPI is a measure of those times when you exhale fully or partially through your mouth. Leaks are a measure of therapy pressure constantly leaving somewhere else, causing the machine to up the average pressure. All masks have a designed leak in the expiration port, some more than others. This prevents you from breathing out into the hose and re-breathing CO2. Mask leaks from poor seals will elevate this, but it won't be recorded as excessive leaks until it reaches 95%.

You can always go to the intellipap web site and input the last day code to see more detail. In my case, AI is low (under 1%)
but I have higher levels of hypopnea. My hypopnea counts increase with pressure while my apnea events stay the same. I get better numbers with limits on pressure increase. You might try adjusting the ranges but you need the software to see what your pressures are doing in reference to the events that are occurring.

The best measure of success is how you feel in the morning and whether you experience daytime sleepiness. Do you tend to nap watching afternoon TV or reading?

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Do I need higher pressure if my AHI is still high?

Post by Pugsy » Wed Aug 31, 2011 8:04 am

Blackspinner gave you spot on suggestion.
Start with minimum pressure adjustment. With 6 minimum the machine may simply not be able to get to the pressure some events may need quickly enough. It is a common happening when using APAPs with ranges.
Give it with a bit of a head start by increasing the minimum a bit.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
bdp522
Posts: 3378
Joined: Tue Jun 27, 2006 5:13 pm

Re: Do I need higher pressure if my AHI is still high?

Post by bdp522 » Wed Aug 31, 2011 2:24 pm

6 is low for the low pressure setting. I would raise the pressure one whole point at a time. Be sure to keep that pressure for at least a few days before changing it again. It could take that long for you to adjust to the change. Most find it best to set an auto machine 2 above and 2 below the titrated pressure. So if you were titrated at 12 the top limit would be 14 and the bottom limit would be 10. Small changes one at a time are best.

Brenda

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Love my papillow, Aussie heated hose and PAD-A-CHEEKS! Also use Optilife, UMFF(with PADACHEEK gasket), and Headrest masks Pressure; 10.5

jnk
Posts: 5784
Joined: Mon Jun 30, 2008 3:03 pm

Re: Do I need higher pressure if my AHI is still high?

Post by jnk » Wed Aug 31, 2011 2:54 pm

snoozebabe wrote:I have been using my CPAP for 5 months now and the AHI is averaging 10.5 / 11. This is still higher than the recommended range.

I have no problem with mask leaks. The leakage is 0%.

Does it mean I need to get my upper limit pressure adjusted even higher (currently at 14)?
What are you diagnosed with, and what other medical conditions are you dealing with?

Self-tweaking may be fine if the only problem you are dealing with is plain obstructive sleep apnea. Self-tweaking may NOT be such a good idea if you are dealing with a form of central apnea, for example. People dealing with things beyond simple obstructive sleep apnea would especially do well to keep their doc and RT informed before considering any pressure adjustments.

Some with pain issues or drug interactions may not be able to get their home-machine-estimated AHI as low as others. That is due to those situations affecting the hypopnea estimates made by their home machine.

If you are dealing only with simple obstructive sleep apnea, I agree with the others that raising the minimum would likely be the thing to do in order to see if that helps the numbers, and more important, how you feel. Your doc may have no problem with your doing that. It is generally considered a courtesy to keep your medical professionals informed about changes to a medical therapy whenever possible, even when you know what you're doing. It keeps the "team" concept in play, if they happen to be team players.

Just sayin'.

johntee
Posts: 90
Joined: Sat Apr 14, 2007 5:36 pm

Re: Do I need higher pressure if my AHI is still high?

Post by johntee » Wed Aug 31, 2011 4:24 pm

Use caution though, if you are increasing pressures. I also (years ago) was "chasing the dragon" by gradually increasing pressures to chase away the remaining Apneas and Hypopneas. At a visit to the sleep clinic doctor (where I was being treated regularly for 1+ year because I wasn't feeling refreshed), she explained that too high of a pressure can actually cause Central Apneas (not good), which she thought was happening in my case; my recollection of her explanation was that apparently the brain is tricked into believing that enough airflow exists (because of the high machine pressure), so the brain decides it doesn't need to take a breath. Much different than your run-of-the-mill Obstructive Apneas, and from what I understand, more dangerous.

She reduced my pressures at that visit, and it actually improved the AHI. (I still didn't feel refreshed, and still don't, but that's another story, lol.)

Good luck!
John

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: Do I need higher pressure if my AHI is still high?

Post by DoriC » Wed Aug 31, 2011 4:51 pm

avi123 wrote:Check this:

THE APNEA-HYPOPNEA INDEX: USEFUL OR USELESS?

Link:

http://www.respiratoryreviews.com/sep02 ... Index.html

“There is not much association between the AHI and anything else—sleepiness, muscle dysfunction,” or other markers for SDB, claimed Dr. Sullivan, who heads the Sleep Disorders Unit at the University of Sydney in Australia. The management of SDB should hinge on the history, examination, and clinical judgment," he asserted.

Not to cause other problems by raising the pressure it needs be done by someone who knows about it.
This article is dated Sept 2002 and we've come a long way since then so disregard it. An AHI<5 is considered good treatment.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08

User avatar
avi123
Posts: 4509
Joined: Tue Dec 21, 2010 5:39 pm
Location: NC

Re: Do I need higher pressure if my AHI is still high?

Post by avi123 » Wed Aug 31, 2011 6:51 pm

Here is a post on this board from two weeks ago. See what can happen by raising CPAP pressure:

Air shoots out eye thru nasolacrimal duct when using cpap
by David II on Tue Aug 23, 2011 1:13 am

Hello. My name is Dave and I am brand new to this as I have just discovered cpaptalk.com this evening, so please be patient with a newbie I have severe obstructive sleep apnea and have for many years, and about 3 years ago went thru testing for this. In the testing process, the pressure they took me up to, which was 20, permanently blew out (for lack of a knowledge of better medical terminology) my left nasolacrimal duct, so that when I use a cpap machine anytime or even just hold my mouth and nose and pressurize, air will constantly squirt out of the corner of inner corner of my left eye. This made me so angry that I quit cpap treatment altogether. Well, it is 3 years later and I have calmed down and wish to do things right here and get on the machine, and my goal would be to stay on the machine until such time as I sufficiently lose enough weight that the doctor lets me off of it. Just FYI, I used a mask that covered both my mouth and nose because they could never get just a nasal mask or pillows to work due them not being able to get my mouth to shut when I slept no matter what they tried. I am writing this to see if anyone else has any experience with this problem and could offer any insight and/or possible solutions, and also to see if anyone has heard of a full face mask and if so would it potentially solve this problem. all help is appreciated. thanks, Dave



Well, who is the poster on this board who is ready to pay the consequences if his/her advice about raising CPAP pressure is not appropriate? Do posters on this board carry malpractice insurance as MDs do?

Re-edited.

_________________
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments:  S9 Autoset machine; Ruby chinstrap under the mask straps; ResScan 5.6
Last edited by avi123 on Thu Sep 01, 2011 6:15 am, edited 1 time in total.
see my recent set-up and Statistics:
http://i.imgur.com/TewT8G9.png
see my recent ResScan treatment results:
http://i.imgur.com/3oia0EY.png
http://i.imgur.com/QEjvlVY.png

User avatar
Pugsy
Posts: 65144
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Do I need higher pressure if my AHI is still high?

Post by Pugsy » Wed Aug 31, 2011 7:18 pm

Yes, people can have pressure induced centrals. It happens to a very small percentage of people.
If someone happened to have a lot of centrals in their sleep study an APAP with a wide range might not have been the best choice. If no history of centrals and an APAP with settings off 6 to 14 already has a wide range and the top is 15. No one has suggest increasing the max... only increasing the minimum to 7. Chances of that increase throwing a truck load of centrals is slim to none but chances of giving the machine a head start on preventing events is pretty darn good.

Snoozebabe it would be better if we could see one of your reports to see where the APAP wants to go.