Up late and Centrals? Coincidence?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ColCpap
Posts: 20
Joined: Tue Mar 22, 2011 3:25 pm
Location: Little Rock, AR

Up late and Centrals? Coincidence?

Post by ColCpap » Tue Aug 09, 2011 10:52 pm

Bottom Line Up Front: Does your level of tiredness affect your rate of centrals // can switching to nasal pillows increase centrals??

Details:
Started CPAP in March 11. I recently (about a month ago) switched to the Swift FX nasal pillow from the Comfort Gel [lots of little leaks and discomfort were the primary reasons]

ANYWAY... I noticed the first couple of nights on the pillows that my Centrals shot up a lot from averaging <1/hr to 4-5/hr. After approx 2-3 days things went back to normal and other than not getting enough sleep [job related] I was resting great.

Fast forward to this week and I have been wiped out. So I went and downloaded data to investigate. My AHI on Friday night was 10! [since being on CPAP 0.5-1.5AHI was average; prior to CPAP in the triple digits]. So I went back and looked at the last month and there is an interesting trend:

On "school nights" I am normally in bed around 12 and up at 6:30. But on Friday and Saturday nights I can be up til 1:30 or 2 and then sleep in till I wake up (~8 hrs or so).
What is alarming is on the nights I stayed up late I had an average AHI of 8.6, 75% CA and 20% HA [school nights its been around 2AHI]. The events are almost all grouped at the end of the night approx 6 hrs into it and/or aorund 7:30 or 8 am.

Any thoughts on why Fri and Sat are so bad??

NOTE: Behaviorally eveything is about the same other regardless of the night (caffiene, nicotine, alcohol, food, routine before bed, etc etc), I just normally seem to be up later watching a movie with the teen or hanging out with the wife.

Thanks for any inputs
Fletch

User avatar
FizzyWater
Posts: 78
Joined: Thu Jun 30, 2011 11:36 pm
Location: Columbus, OH

Re: Up late and Centrals? Coincidence?

Post by FizzyWater » Wed Aug 10, 2011 12:38 am

Not sure if there's a real correlation, but it seems to me the longer I'm asleep, the higher my numbers are. Could just be that you're sleeping longer, and in those last hours (which are not the deepest levels of sleep) you have more centrals/whatever...?

_________________
Mask: Opus 360 Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: backup machine: PR System One REMstar Pro CPAP Machine with C-Flex Plus; alternate mask: Swift FX Bella (love those loops!)
Began CPAP therapy 3/22/2011.

Some days, I guess there just aren't enough rocks -- Forrest Gump

User avatar
JointPain
Posts: 261
Joined: Thu Apr 28, 2011 10:41 am
Location: Houston, TX

Re: Up late and Centrals? Coincidence?

Post by JointPain » Wed Aug 10, 2011 8:24 am

I tend to have two phases during the night. During the first phase, I have lots of "events" (OA, H, Flow limitations) that cause the APAP to bump the pressure up. Towards the morning, my pressure line is much lower and flatter, but sometimes I'll have a few to a lot of short centrals (10-15s). Sometimes the two sleep phases are separated by a middle-of-the-night bathroom break.

If I'm tired, I tend to have more "events" in the first half of the night. I haven't noticed any effect of tiredness one day on centrals that night.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Pressure is 11-14. Old CPAP was a Resmed S6 Lightweight. Also have Profile lite mask. ResScan is actually version 3.14. Now I use Sleepyhead.

pap4life
Posts: 207
Joined: Tue Jan 25, 2011 12:08 pm
Location: Florida

Re: Up late and Centrals? Coincidence?

Post by pap4life » Wed Aug 10, 2011 8:36 am

Very interesting post..

Coincedentially, the second half of the night is when I have the most centrals also.

User avatar
Pugsy
Posts: 65093
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Up late and Centrals? Coincidence?

Post by Pugsy » Wed Aug 10, 2011 8:48 am

ColCpap wrote:The events are almost all grouped at the end of the night approx 6 hrs into it and/or aorund 7:30 or 8 am.
Is it possible that these groupings are related to semi awake stage? Sort of like sleep onset in reverse?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Otter
Posts: 1063
Joined: Sat Mar 19, 2011 5:02 pm

Re: Up late and Centrals? Coincidence?

Post by Otter » Wed Aug 10, 2011 10:21 am

Maybe it's not so much that you're up late as that you're sleeping later. The sun, noise from neighbors, or your habit of waking up earlier may be rousing you, and that could cause the centrals. Or maybe it's simply that your alarm is waking you up on weekdays, preventing you from scoring drowsy centrals as you wake up naturally.

Do you think the change to pillows is disturbing your sleep? If not, the wiped out feeling may be a good sign. Many people go through a stage where they feel tired again, perhaps even more tired than they did before starting therapy. I think this sometimes has to do with the need for periodic shots of adrenalin to keep breathing. The adrenal glands seem to get stuck partly on, but after a few months of, oxygen-saturated sleep, they calm down, and it becomes obvious that the years of sleep debt haven't been paid off yet. Serotonin levels returning to normal could have a similar effect.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Software: SleepyHead 0.9 beta

User avatar
Bobby269
Posts: 803
Joined: Sat Nov 27, 2010 11:12 am
Location: San Antonio Texas

Re: Up late and Centrals? Coincidence?

Post by Bobby269 » Wed Aug 10, 2011 2:51 pm

The less sleepy I am the more central clusters I have. I try to go to bed the same time each night. If I stay up an hour past that time I don't sleep as deep and
my central clusters are there in the morning. Wednesday and Thursday I get up very early. So Thursday night Friday morning I sleep very deep little or no
Apnea's
When I keep my sleep to no more than 7 1/2 hours a night and life is good.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Set to Apap 14.00/20. EPR 3, Climate line hose, Oscar for Mac 1.1.0. Betta.1

wolewyck
Posts: 150
Joined: Wed Sep 29, 2010 3:17 pm

Re: Up late and Centrals? Coincidence?

Post by wolewyck » Wed Aug 10, 2011 6:36 pm

Can't advise about the sleep timing thing, but it's true that some people have centrals increase when on CPAP or BiPAP therapy, even though the obstructive events are treated effectively (I am one of those folks). If problems persist, you may want to see about getting another sleep titration study, using an ASV machine, which can address both centrals and obstructive events very effectively.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: ResMed S9 VPAP Adapt SV