What do you think of Auto BiPAP BiFlex?
What do you think of Auto BiPAP BiFlex?
I'm a newbie with a respironics BiPAP Auto Bi-Flex machine
pressures running as high as 17 at times
the machine supposedly figures out and delivers whatever pressure you need to keep your airway open...instead of a continuous pressure
also...it backs off on pressure to make it easier to exhale
what are the positives and the negatives to this system? Is this a smart choice for someone who might need higher pressures?
thanks again!!!
evan
pressures running as high as 17 at times
the machine supposedly figures out and delivers whatever pressure you need to keep your airway open...instead of a continuous pressure
also...it backs off on pressure to make it easier to exhale
what are the positives and the negatives to this system? Is this a smart choice for someone who might need higher pressures?
thanks again!!!
evan
Re: What do you think of Auto BiPAP BiFlex?
Well,, first off, you are NOT going to get a BPAP w/o a script for a bi-level.
I was fortunate to use a PR SystemOne BPAP Auto w/Bi-Flex for some 6-7 months. I found it very comfortable, no problems exhaling, smooth transitions from IPAP to EPAP. I didn't use the Bi-Flex settings at all. I just didn't need it.
I was fortunate to use a PR SystemOne BPAP Auto w/Bi-Flex for some 6-7 months. I found it very comfortable, no problems exhaling, smooth transitions from IPAP to EPAP. I didn't use the Bi-Flex settings at all. I just didn't need it.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator |
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.
My computer says I need to upgrade my brain to be compatible with its new software.
Re: What do you think of Auto BiPAP BiFlex?
BiPap is the ultimate in exhale relief BUT it doesn't figure out your pressures and needs to be setup correctly.nmevan wrote:I'm a newbie with a respironics BiPAP Auto Bi-Flex machine
pressures running as high as 17 at times
the machine supposedly figures out and delivers whatever pressure you need to keep your airway open...instead of a continuous pressure
also...it backs off on pressure to make it easier to exhale
what are the positives and the negatives to this system? Is this a smart choice for someone who might need higher pressures?
thanks again!!!
evan
_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: New users can't remember they can't remember YET! |
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
Re: What do you think of Auto BiPAP BiFlex?
I've used mine for 2 nights, so I'm not an expert yet. I asked my sleep doc if I could try one for 2 weeks to see if it helped me. He agreed, so Apria loaned me one with 62 hours on it. I get to keep it if the numbers turn out to show that I have improved. It is set at 4/20 pressure. My Autoset was set at 7.8/12.4 pressure. I probably should change the settings on my bipap, but I'm afraid that if I mess with the settings, the DME will "hit the ceiling". Anyway, I am getting used to it, and hope I will be able to keep it. I love the fact that the humidifier self adjusts, and it seems to be just as quiet as my Autoset.
Jay
Jay
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack |
Additional Comments: MAD device last 18 years (when power is out) |
Resmed Vauto S Bilevel
Airfit n30
Airfit n30
Re: What do you think of Auto BiPAP BiFlex?
GumbyCT, the BPAP Auto w/Bi-Flex "can" be used for auto-titrating bi-level IPAP and EPAP pressure needs. Usually one would set the Pressure Support at 4 cms. IPAP and EPAP can then free range as needed according to their settings EXCEPT that they can not get any closer together than 4 cms (or whatever the PS is set at).
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator |
Last edited by Slinky on Fri Nov 05, 2010 4:23 pm, edited 1 time in total.
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.
My computer says I need to upgrade my brain to be compatible with its new software.
Re: What do you think of Auto BiPAP BiFlex?
IF it is setup correctly, sure. I wouldn't start at 4cm either.Slinky wrote:GumbyCT, the BPAP Auto w/Bi-Flex "can" be used for auto-titration bi-level IPAP and EPAP pressure needs.
_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: New users can't remember they can't remember YET! |
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
Re: What do you think of Auto BiPAP BiFlex?
Pressure Support is NOT the staring pressure. It is a "range" of pressure that IPAP and EPAP can not get closer to each other than its setting.
For instance, IPAP 17, EPAP 8, Presure Support 4. Pressure during the night could then vary any where between 8 and 17 excpet that IPAP and EPAP could never get closer together than 4 cms. Or to put it another way, IPAP could range anywhere from 13 to 17 as needed and EPAP could range anywhere from 8 to 12 as needed.
At least that's my understanding.
For instance, IPAP 17, EPAP 8, Presure Support 4. Pressure during the night could then vary any where between 8 and 17 excpet that IPAP and EPAP could never get closer together than 4 cms. Or to put it another way, IPAP could range anywhere from 13 to 17 as needed and EPAP could range anywhere from 8 to 12 as needed.
At least that's my understanding.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator |
Last edited by Slinky on Sat Nov 06, 2010 6:47 am, edited 1 time in total.
Women are Angels. And when someone breaks our wings, we simply continue to fly.....on a broomstick. We are flexible like that.
My computer says I need to upgrade my brain to be compatible with its new software.
My computer says I need to upgrade my brain to be compatible with its new software.
Re: What do you think of Auto BiPAP BiFlex?
I think that is never get farther apart than 4cm of course if you set pressure support at 6cm then 4cm apart would be within those parameters.Slinky wrote:Pressure Support is NOT the staring pressure. It is a "range" of pressure that IPAP and EPAP can not get closer to each other than its setting.
For instance, IPAP 17, EPAP 8, Presure Support 4. Pressure during the night could then vary any where between 8 and 17 excpet that IPAP and EPAP could never get closer together than 4 cms. Or to put it another way, IPAP could range anywhere from 13 to 17 as needed and EPAP could range anywhere from 8 to 23 as needed.
At least that's my understanding.
_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: New users can't remember they can't remember YET! |
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
- rested gal
- Posts: 12881
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
Re: What do you think of Auto BiPAP BiFlex?
You're right about "Pressure Support" not being the starting pressure.Slinky wrote:Pressure Support is NOT the staring pressure. It is a "range" of pressure that IPAP and EPAP can not get closer to each other than its setting.
However, I believe it is what prevents the IPAP/EPAP from getting farther apart from each other than whatever number the "pressure support" is set for. I don't think the "max press sup" setting is to prevent IPAP and EPAP from getting closer to each than that number. At least, that's my understanding of the way it is in Respironics machines. In ResMed machines the pressure support setting works differently.
As an aside...Jeff (jnk) described beautifully the difference in the way the two brands handle "pressure support" in the topic:
ResMed VPAP Auto 25 Clinician's manual
viewtopic.php?p=376749#p376749
Back to the Respironics BiPAP Auto and Philips Respironics System One BiPAP Auto, when the machine is being used in autotitrating bilevel mode, max IPAP and min EPAP can be set as close as 3 cm to each other, and they will actually operate as closely as 2 cm apart.
No matter how where you set the max IPAP / min EPAP pressures, the machine (in autotitrating bilevel mode) will begin each sleep session blowing an IPAP pressure only 2 cms higher than whatever the "min EPAP" was set for.
With the Respironics and Philips Respironics machines, IPAP and EPAP can get closer together than 4 cm. They can get as close as 2 cm to each other. That's where those two pressures start out (2 cm apart) each time the machine is used. They just can't get farther apart from each other than 4 cm if "Max Press Sup" is set for 4 cm.Slinky wrote:For instance, IPAP 17, EPAP 8, Presure Support 4. Pressure during the night could then vary any where between 8 and 17 excpet that IPAP and EPAP could never get closer together than 4 cms.
I think you typo'd IPAP/EPAP reversed, but even with those words corrected I don't think that's the range either of those pressures could use in the example of max IPAP 17, min EPAP 8, max press sup 4.Slinky wrote: Or to put it another way, IPAP could range anywhere from 13 to 17 as needed and EPAP could range anywhere from 8 to 23 as needed.
At least that's my understanding.
In that example settings scenario, I think it would go this way....
IPAP could range from 10 to 17. If IPAP did, on its own initiative (so to speak) go to 17, EPAP would have been dragged up to 13 -- in order to keep them no farther apart than the 4 cm that was set for "maximum pressure support."
10 would be as low as IPAP could go, assuming EPAP was rocking along down there at the min EPAP of 8.
17 would be as high as IPAP could go, since that's what "max IPAP" was set for. 17.
EPAP could range from 8 to 15. If EPAP did, on its own, go up to 15, IPAP would have been pushed up to 17, in order to keep no less than 2 cm between them.
IPAP and EPAP can operate as close as 2 cm to each other. They just can't range more than 4 cm apart from each other -- if the Max Press Sup is set for 4 cm. The word "Max" in the "Max Press Sup" setting of the Respironics BiPAP Auto machine is the key word. That setting governs the maximum difference that IPAP/EPAP can travel apart from each other. A built-in difference of 2 cm allows them to snuggle as closely as two cm with each other.
Two cms apart is the way the machine will start each session, no matter how high the max IPAP is set, and no matter how much you set the "Max Pressure Support" for. The most the Max Press Sup setting can be set for is 8 cm -- that is if the max IPAP and min EPAP are set far enough apart to allow 8 cm difference between them.
That's my understanding anyway, but I sure could be wrong or be stating it incorrectly.
Last edited by rested gal on Fri Nov 05, 2010 10:49 pm, edited 2 times in total.
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Re: What do you think of Auto BiPAP BiFlex?
you certainly have added a lot of information about the machine I have.
just wondering if you think it is a good machine? and for what reasons?
or would you suggest something else for someone whose sleep study showed 53 AHIs per hour and who needs a high pressure of around 17
at times
thanks
evan
just wondering if you think it is a good machine? and for what reasons?
or would you suggest something else for someone whose sleep study showed 53 AHIs per hour and who needs a high pressure of around 17
at times
thanks
evan
Re: What do you think of Auto BiPAP BiFlex?
The "best" machine is the one that gives you the best therapy. A machine can give best therapy to 95% of the people, but if you belong to the 5%, you need another one.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
- rested gal
- Posts: 12881
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
Re: What do you think of Auto BiPAP BiFlex?
evan, I don't actually need bilevel treatment, or even autotitration. But I like the nearly instant drop in pressure when I breathe out that bilevel gives. And I like autotitration for occasions when I might need more pressure than the straight "11" that would do a good job for me the majority of time.
I personally just like using a bilevel machine whether I set it for bilevel mode or for autotitrating bilevel mode.
Either brand (Respironics or ResMed) of bilevel or bilevel auto would suit me fine.
Everyone is different, so what type of machine and what mode of operation suits a particular person best can be different. Generally speaking, I'd think a bilevel or autotitrating bilevel (either one) would probably be more comfortable to breathe out against for most people who need a pressure of 17 at times.
I think you have a good BiPAP (Respironics trademark name for their bilevel machines.) So do the people who have "VPAP" machines (ResMed's trademark name for their bilevel machines.)
I personally just like using a bilevel machine whether I set it for bilevel mode or for autotitrating bilevel mode.
Either brand (Respironics or ResMed) of bilevel or bilevel auto would suit me fine.
Everyone is different, so what type of machine and what mode of operation suits a particular person best can be different. Generally speaking, I'd think a bilevel or autotitrating bilevel (either one) would probably be more comfortable to breathe out against for most people who need a pressure of 17 at times.
I think you have a good BiPAP (Respironics trademark name for their bilevel machines.) So do the people who have "VPAP" machines (ResMed's trademark name for their bilevel machines.)
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Re: What do you think of Auto BiPAP BiFlex?
Bless you, RG, for clarifying the Pressure Support information!
I've used both the Resmed VPAP Auto and the PR SystemOne BPAP Auto w/Bi-Flex and done well w/both devices. I can't even say that I prefer one over the other as there are features about both that I like. For "just" therapy, I really couldn't pick one over the other. There was a "learning curve" w/the VPAP Auto, but not w/the PR S1 (which I had the opportunity to use AFTER having used the VPAP Auto for some 12-13 months). My likes and dislikes had more to do w/the software, the data available, the method of the data availability via the LCD screen, the integrated humidifiers, etc.
I liked the additional data of the S1 but hated the limited availability of that data via the LCD screen. I liked the visibility of the VPAP's humidifier tank, I hated the extremely limited visibility of the S1 humidifier tank. (I don't empty and refill my humidifier tank each day, I just top the water level off). I preferred the ResScan software to the EncorePro software.
I had difficulties w/the VPAP Auto in Spontaneous (straight bi-level) mode w/the "comfort feature" settings being too complicated for me to find a smooth, easy transition between IPAP and EPAP pressures and my "professionals" not being knowledgeable enough about the new VPAP Autos to do so either. HOWEVER, that was easily remedied by using it in Auto mode and limiting the Pressure Support to the exact setting between IPAP and EPAP settings. This allowed the EasyBreathe technology to take over and the pressure transistions between IPAP and EPAP were comfortable, smooth and easy breathing. I had no such problems w/the S1. From the very beginning pressure transition in straight bi-level mode was smooth and easy and comfortable w/no need for any adjustments.
Frankly, the Mask Selection options of both devices were a .... PITA, IF you didn't use one of that manufacturerer's own masks. This was especially true w/the S1 BPAP rather than the VPAP as determining which mask was which setting was considerably more difficult w/the S1. IF you don't find and use the correct mask selection your data is one big fat lie and not reliable w/either device.
I've used both the Resmed VPAP Auto and the PR SystemOne BPAP Auto w/Bi-Flex and done well w/both devices. I can't even say that I prefer one over the other as there are features about both that I like. For "just" therapy, I really couldn't pick one over the other. There was a "learning curve" w/the VPAP Auto, but not w/the PR S1 (which I had the opportunity to use AFTER having used the VPAP Auto for some 12-13 months). My likes and dislikes had more to do w/the software, the data available, the method of the data availability via the LCD screen, the integrated humidifiers, etc.
I liked the additional data of the S1 but hated the limited availability of that data via the LCD screen. I liked the visibility of the VPAP's humidifier tank, I hated the extremely limited visibility of the S1 humidifier tank. (I don't empty and refill my humidifier tank each day, I just top the water level off). I preferred the ResScan software to the EncorePro software.
I had difficulties w/the VPAP Auto in Spontaneous (straight bi-level) mode w/the "comfort feature" settings being too complicated for me to find a smooth, easy transition between IPAP and EPAP pressures and my "professionals" not being knowledgeable enough about the new VPAP Autos to do so either. HOWEVER, that was easily remedied by using it in Auto mode and limiting the Pressure Support to the exact setting between IPAP and EPAP settings. This allowed the EasyBreathe technology to take over and the pressure transistions between IPAP and EPAP were comfortable, smooth and easy breathing. I had no such problems w/the S1. From the very beginning pressure transition in straight bi-level mode was smooth and easy and comfortable w/no need for any adjustments.
Frankly, the Mask Selection options of both devices were a .... PITA, IF you didn't use one of that manufacturerer's own masks. This was especially true w/the S1 BPAP rather than the VPAP as determining which mask was which setting was considerably more difficult w/the S1. IF you don't find and use the correct mask selection your data is one big fat lie and not reliable w/either device.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: PR SystemOne BPAP Auto w/Bi-Flex & Humidifier - EncorePro 2.2 Software - Contec CMS-50D+ Oximeter - Respironics EverFlo Q Concentrator |