resmed card reader not recognized any more

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
RMinOntario
Posts: 121
Joined: Sat Jun 06, 2009 2:40 pm
Location: Guelph, Ontario, Canada

resmed card reader not recognized any more

Post by RMinOntario » Sun Jun 27, 2010 10:46 am

Hi all. Coming up on my 1yr anniversary with CPAP this week. I haven't been around the forum for ages but now have a question. I never had a problem reading my data from the ACR reader into the 3.7 software until a week ago. Now there is no audible sound when I insert the reader into any of the USB ports (and they are all functioning properly). If I go ahead and try to run the software and download, I get the msg 'THERE ARE NO DEVICES AVAILABLE FOR DOWNLOADING DATA. CONNECT A FLOW GENERATOR OR DATA CARD AND TRY AGAIN.' I remember ages ago advice was to disable the wireless card and it would fix the problem. I tried that and still the same problem. I am using the same computer as always and have tried 2 different cards in the reader, but it really is that the reader is not recognized. Any ideas would be welcome.

_________________
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Additional Comments: AHI at sleep study - 39.6, Pressure 12.4 cm H2O, equipment trials 6/2009, my own equipment 7/2009

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: resmed card reader not recognized any more

Post by Wulfman » Sun Jun 27, 2010 10:56 am

Did it coincidentally occur after applying Microsoft patches or updates? That's where I'd be inclined to look. I'm not familiar with your ResMed software, but I know there were incidents in the past where Encore Pro software would fail after the user updated XP with Service Pack 3 (and possibly other updates).


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

RMinOntario
Posts: 121
Joined: Sat Jun 06, 2009 2:40 pm
Location: Guelph, Ontario, Canada

Re: resmed card reader not recognized any more

Post by RMinOntario » Sun Jun 27, 2010 11:08 am

I can't say if it was right after an update or not ... sorry. I have Vista on my laptop. Can't think of anything 'I' changed to caused this. If it did happen after an update, what can I really do about it? I select only the necessary updates to Windows to go through.

_________________
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Additional Comments: AHI at sleep study - 39.6, Pressure 12.4 cm H2O, equipment trials 6/2009, my own equipment 7/2009

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: resmed card reader not recognized any more

Post by Wulfman » Sun Jun 27, 2010 11:23 am

I don't know. I'm not a Vista user, so I'll defer to other users who may have any ideas for you.


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: resmed card reader not recognized any more

Post by jdm2857 » Sun Jun 27, 2010 11:42 am

Have you tried uninstalling and reinstalling the card reader driver?
jeff

Hose_Head
Posts: 804
Joined: Wed Mar 12, 2008 6:43 pm
Location: Ontario Canada

Re: resmed card reader not recognized any more

Post by Hose_Head » Sun Jun 27, 2010 12:07 pm

[quote="jdm2857"]Have you tried uninstalling and reinstalling the card reader driver?[/quote

Another idea: If you have access to another computer, you might try plugging into it to see if that computer recognizes the card reader. This will tell you whether the problem lies with the card reader or with your main computer.
I'm workin' on it.

vapat75
Posts: 21
Joined: Tue May 25, 2010 3:27 pm
Location: Virginia

Re: resmed card reader not recognized any more

Post by vapat75 » Sun Jun 27, 2010 12:20 pm

You may want to try this: when you get to no devices detected, just click ok; at the download; put your card in the reader and then plug in your card reader (don't plug your card reader and card before; I had to do this before I could get it to recognize the reader.

_________________
Mask

RMinOntario
Posts: 121
Joined: Sat Jun 06, 2009 2:40 pm
Location: Guelph, Ontario, Canada

Re: resmed card reader not recognized any more

Post by RMinOntario » Sun Jun 27, 2010 1:01 pm

Well, there is success ... thank you. I uninstalled / reinstalled the drivers on my Vista laptop (my main computer) and that worked. I guess it was corrupt. Prior to that, I booted up my XP desktop and the card reader was read successfully. It also read okay on my travel Vista netbook. After this test on my netbook, I went to apply the latest necessary Windows updates and there was an optional update for the ACR38 card reader. This did not show up in the updates for my laptop, possibly due to the fact that the driver had ? disappeared. I agreed to the update, and just noticed, that it has failed. I think I will leave well enough alone and not attempt another installation.

_________________
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Additional Comments: AHI at sleep study - 39.6, Pressure 12.4 cm H2O, equipment trials 6/2009, my own equipment 7/2009

User avatar
kteague
Posts: 7783
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Re: resmed card reader not recognized any more

Post by kteague » Sun Jun 27, 2010 1:59 pm

Glad you posted this question. I was going to post a similar one, and have it on good authority that Wulfman Den (not to be confused with D.J. Wolfman Jack) is "the man" on this kind of thing. My situation is a bit different. Changed computers, reinstalled Encore Pro, but card reader is not recognized. Updates were run on this computer, I think before the reinstall. I have a vague memory of there being instructions on here about going into Encore Pro's files and checking off something. Any clues Den or anyone else?

Not trying to hijack your thread, just thinking we may have the same solution.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: resmed card reader not recognized any more

Post by Wulfman » Sun Jun 27, 2010 2:37 pm

kteague wrote:Glad you posted this question. I was going to post a similar one, and have it on good authority that Wulfman Den (not to be confused with D.J. Wolfman Jack) is "the man" on this kind of thing. My situation is a bit different. Changed computers, reinstalled Encore Pro, but card reader is not recognized. Updates were run on this computer, I think before the reinstall. I have a vague memory of there being instructions on here about going into Encore Pro's files and checking off something. Any clues Den or anyone else?

Not trying to hijack your thread, just thinking we may have the same solution.
Please define "changed computers".
New computer? One you already had?
Which operating system?
Which version of Encore Pro?

Need a few more details, Kathy.

Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
kempo
Posts: 1023
Joined: Thu Jun 17, 2010 10:09 am

Re: resmed card reader not recognized any more

Post by kempo » Sun Jun 27, 2010 3:21 pm

You might have malware. If you don't have anti-malware go to CNET downloads and download "Malwarebytes" and do a complete scan of your computer. It's free and and does a damn good job of removing that nasty stuff. Malware will make your computer do crazy things.

_________________
MachineMask

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: resmed card reader not recognized any more

Post by Wulfman » Sun Jun 27, 2010 4:15 pm

kteague wrote:Glad you posted this question. I was going to post a similar one, and have it on good authority that Wulfman Den (not to be confused with D.J. Wolfman Jack) is "the man" on this kind of thing. My situation is a bit different. Changed computers, reinstalled Encore Pro, but card reader is not recognized. Updates were run on this computer, I think before the reinstall. I have a vague memory of there being instructions on here about going into Encore Pro's files and checking off something. Any clues Den or anyone else?

Not trying to hijack your thread, just thinking we may have the same solution.
If you're using EP 1.8, there's an option in the menus for selecting the type of card reader connection. Check to see if it's defaulting to serial......should be USB.

When you reinstalled EP, did you have your card reader connected to the computer? It should not be connected until after the software is installed and a re-boot is done. Then (I think) run the Setup.exe (from the install CD) to install the card reader driver and then connect the card reader.

There may be some other threads about installing 1.8, but I found this one from RG.

viewtopic.php?f=1&t=29934&p=257825&hili ... ll#p257825


Den
Last edited by Wulfman on Sun Jun 27, 2010 4:32 pm, edited 4 times in total.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
torontoCPAPguy
Posts: 1015
Joined: Mon Dec 28, 2009 11:27 am
Location: Toronto Ontario/Buffalo NY

Re: resmed card reader not recognized any more

Post by torontoCPAPguy » Sun Jun 27, 2010 4:17 pm

Just as a matter of interest you should know that the seating of the card in the reader is an issue that has never been addressed by ResMed; so make sure that your card is actually being seen when you plug it in. It's a matter of trial and error. If the card is not being seen the reader will not be seen either. A little wiggling usually resolves the issue.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Respironics Everflo Q infusing O2 into APAP line to maintain 95% SaO2; MaxTec Maxflo2 Oxygen Analyzer; Contec CMS50E Recording Pulse Oxymeter
Fall colours. One of God's gifts. Life is fragile and short, savour every moment no matter what your problems may be. These stunning fall colours from my first outing after surviving a month on life support due to H1N1.

User avatar
SleepyT
Posts: 775
Joined: Wed Jul 08, 2009 10:50 am

Re: resmed card reader not recognized any more

Post by SleepyT » Sun Jun 27, 2010 5:37 pm

I had this problem pop up yesterday...and I have tried your suggestions (except for the reinstalling one) and so no go. I didn't manually update anything...perhaps an automatic update did something to disable my card reader?

I cannot find the disk I used to install the card driver...not yet anyway. I have Windows XP.

This is driving me a little batty!
"Knowledge is power."

RMinOntario
Posts: 121
Joined: Sat Jun 06, 2009 2:40 pm
Location: Guelph, Ontario, Canada

Re: resmed card reader not recognized any more

Post by RMinOntario » Sun Jun 27, 2010 10:54 pm

SleepyT, the following website is where you can get the drivers. Choose the one to match your computer system for the ACR38.

http://www.acs.com.hk/index.php?pid=drivers&id=ACR38

The first time you run it, select to uninstall the driver, then run it again and do a fresh install. Drivers, like anything in life, can just hiccup.

_________________
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear
Additional Comments: AHI at sleep study - 39.6, Pressure 12.4 cm H2O, equipment trials 6/2009, my own equipment 7/2009