M Series Auto w/ AFEX Whistling and It's NOT Dixie

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
bjdare
Posts: 33
Joined: Tue Jan 06, 2009 9:43 am
Location: Manteca, CA

M Series Auto w/ AFEX Whistling and It's NOT Dixie

Post by bjdare » Sat May 29, 2010 2:31 am

My M-Series has started whistling and it's driving me crazy. Most nights it starts within minutes of putting my mask on and I have no idea how long it continues. I wake up (briefly) several times during the course of the evening and don't notice it at those times. However, this morning my husband woke up around 6 a.m. and said it sounded like a freight train. More often than not, it sounds like the back up "beep" on a car or truck. I noticed an earlier post by John10 that he was having a similar problem with the PR system. I've had the M-Series for 14 months and the "whistling" has just started within the last month. Other than the annoying sound, the machine is functioning fine. Has ANYONE out there had a problem with the M-Series doing this? Not sure if I should send it for service or try to teach it a tune.

Thanks for your help / comments.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Additional Comments: 13 cm/H2O

cflame1
Posts: 3311
Joined: Sat Mar 11, 2006 6:55 am
Location: expat Canadian in Kentucky

Re: M Series Auto w/ AFEX Whistling and It's NOT Dixie

Post by cflame1 » Sat May 29, 2010 3:00 am

have you checked for a hole in your hose? If there isn't does your mask fit correctly?

User avatar
PST
Posts: 986
Joined: Sat May 30, 2009 9:56 pm

Re: M Series Auto w/ AFEX Whistling and It's NOT Dixie

Post by PST » Sat May 29, 2010 10:13 am

I hear the same complaint from my wife sometimes. It always turns out to be a tiny, sporadic leak between mask and face squealing, not the flow generator.

User avatar
bjdare
Posts: 33
Joined: Tue Jan 06, 2009 9:43 am
Location: Manteca, CA

Re: M Series Auto w/ AFEX Whistling and It's NOT Dixie

Post by bjdare » Sat May 29, 2010 11:48 am

The hose is brand new - doubled checked it just in case - and my mask is sealing great - 0 leaks. Last night, 5-10 minutes after masking up I leaned into the machine and the "whistle" sounded as if it was coming from within the machine it's self. My husband also heard it again when he woke up this morning. My pressure (90% & mean) have been unusually high the last two nights (90% at 17 & 16 vs normal of 14, mean 14.3 & 13.8 vs normal of 13.1 - 13.4) and my AHI is up (2.0 vs normal of .7). I don't know if any of this is tied to the machine or just my body having a couple of "off" days (I feel fine). I've been aware of the whistle for the last month or so, but my husband just started hearing it the last two days. I'm really not sure what to make of it!!

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Additional Comments: 13 cm/H2O

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: M Series Auto w/ AFEX Whistling and It's NOT Dixie

Post by Wulfman » Sat May 29, 2010 11:56 am

I would suggest hooking up and running it during the daytime while you're awake......to see if you can diagnose where the "whistle" is coming from. (take the smart card out while you're performing this "diagnostic experiment")


Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
bjdare
Posts: 33
Joined: Tue Jan 06, 2009 9:43 am
Location: Manteca, CA

Re: M Series Auto w/ AFEX Whistling and It's NOT Dixie

Post by bjdare » Sat May 29, 2010 11:59 am

Thanks Den - will do that today!

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Additional Comments: 13 cm/H2O