change pressure on remstar choice and tranquility quest
Am probably borrowing an old machine from my mom rather than renting since mine is in the shop, so does anyone know how to adjust the pressure?
Re: change pressure on remstar choice and tranquility quest
If I remember right there is little door on the bottom right of the machine. If you open it there is a screwdriver slot that says pressure. You have to turn it and set it with a water manometer. Does not have pre programed pressure settings.
Jerry
Jerry
_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: 11cm/H2O, Encore Pro 1.8i, Pro Analyzer, Encore Viewer1.0 - 3 Remstar Pro2's, 1 Remstar Auto |
Life's journey is not to arrive at the grave safely in a well preserved body, but rather to skid in sideways, totally worn out, shouting: "Wow what a ride!"
I still play Cowboys and Bad Guys but now I use real bullets. CAS
I still play Cowboys and Bad Guys but now I use real bullets. CAS
Re: change pressure on remstar choice and tranquility quest
6PtStar wrote:If I remember right there is little door on the bottom right of the machine. If you open it there is a screwdriver slot that says pressure. You have to turn it and set it with a water manometer. Does not have pre programed pressure settings.
Jerry
Exactly
DSM
(Tranquility Quest that is )
Last edited by dsm on Sat Nov 28, 2009 12:11 am, edited 1 time in total.
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)
Re: change pressure on remstar choice and tranquility quest
either this one?

the top comes off the tower to set it but it still needs a manometer - two dials inside - one is pressure - one is ramp

the top comes off the tower to set it but it still needs a manometer - two dials inside - one is pressure - one is ramp
Re: change pressure on remstar choice and tranquility quest
Hello, this is my first post.
I have an old RemStar Choice and would like to change the pressure settings from the current 6 cm H2O. I tried taking off the case of the unit by removing the 3 screws on the bottom but the removal was halted by the interference of 3 cables which connected to the circuit board and which were extremely difficult to get at and likely even more difficult to reconnect so I didn't go any further. The posts on this subject seemed to indicate there was an external way to get at the trim pots but I have not been able to find the openings to the pots. For those who have posted about this, can you be any more specific about exactly where the external access is located on this unit ? Thanks for any help you might be able to provide in this regard.
I have an old RemStar Choice and would like to change the pressure settings from the current 6 cm H2O. I tried taking off the case of the unit by removing the 3 screws on the bottom but the removal was halted by the interference of 3 cables which connected to the circuit board and which were extremely difficult to get at and likely even more difficult to reconnect so I didn't go any further. The posts on this subject seemed to indicate there was an external way to get at the trim pots but I have not been able to find the openings to the pots. For those who have posted about this, can you be any more specific about exactly where the external access is located on this unit ? Thanks for any help you might be able to provide in this regard.
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: change pressure on remstar choice and tranquility quest
I have the service manual ( in .pdf form ) for the "Choice" models, which clearly shows how to change it (and more).aquavion wrote: ↑Tue Mar 19, 2019 9:41 amHello, this is my first post.
I have an old RemStar Choice and would like to change the pressure settings from the current 6 cm H2O. I tried taking off the case of the unit by removing the 3 screws on the bottom but the removal was halted by the interference of 3 cables which connected to the circuit board and which were extremely difficult to get at and likely even more difficult to reconnect so I didn't go any further. The posts on this subject seemed to indicate there was an external way to get at the trim pots but I have not been able to find the openings to the pots. For those who have posted about this, can you be any more specific about exactly where the external access is located on this unit ? Thanks for any help you might be able to provide in this regard.
If you wish me to send it to you, PM (private message) me with your email address and I'll shoot it to you. It's not all that big of a file.......less than 1 MB.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: change pressure on remstar choice and tranquility quest
<r><QUOTE><s>
I have the service manual ( in .pdf form ) for the "Choice" models, which clearly shows how to change it (and more).<br/>
If you wish me to send it to you, PM (private message) me with your email address and I'll shoot it to you. It's not all that big of a file.......less than 1 MB.<br/>
<br/>
Many thanks to Wulfman for providing me a complete service manual on the RemStar "Choice" !!</r>
</e></QUOTE></s><e>
I have the service manual ( in .pdf form ) for the "Choice" models, which clearly shows how to change it (and more).<br/>
If you wish me to send it to you, PM (private message) me with your email address and I'll shoot it to you. It's not all that big of a file.......less than 1 MB.<br/>
<br/>
Many thanks to Wulfman for providing me a complete service manual on the RemStar "Choice" !!</r>
- chunkyfrog
- Posts: 34545
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: change pressure on remstar choice and tranquility quest
So that is what a dinosaur looks like . . .


_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- Okie bipap
- Posts: 3566
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: change pressure on remstar choice and tranquility quest
I always thought dinosaurs were much bigger.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
Re: change pressure on remstar choice and tranquility quest
just to give you a heads up, if you want to quote someone here on the forum, just use the quotation marks in the upper right hand side. otherwise, it looks like what i just quoted above.aquavion wrote: ↑Wed Mar 20, 2019 7:57 am<r><QUOTE><s></e></QUOTE></s><e>
I have the service manual ( in .pdf form ) for the "Choice" models, which clearly shows how to change it (and more).<br/>
If you wish me to send it to you, PM (private message) me with your email address and I'll shoot it to you. It's not all that big of a file.......less than 1 MB.<br/>
<br/>
Many thanks to Wulfman for providing me a complete service manual on the RemStar "Choice" !!</r>
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg