E-47 message on remstar pro2 machine
Re: E-47 message on remstar pro2 machine
E-47 ERR_FL12_SENSOR_RAILED •Occluded or restricted pressure tube •Defective flow sensor •Replace flow tube •Replace Main PCA
More than likely the small pressure line just inside the case that runs off of the hose connector has broken off. Some have repaired it with epoxy or some similar material. It you twist off the hose it puts pressure on the pressure line connector and sometimes causes it to break off. The line could just be plugged or restricted if you are lucky.
Jerry
More than likely the small pressure line just inside the case that runs off of the hose connector has broken off. Some have repaired it with epoxy or some similar material. It you twist off the hose it puts pressure on the pressure line connector and sometimes causes it to break off. The line could just be plugged or restricted if you are lucky.
Jerry
_________________
| Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
| Additional Comments: 11cm/H2O, Encore Pro 1.8i, Pro Analyzer, Encore Viewer1.0 - 3 Remstar Pro2's, 1 Remstar Auto |
Life's journey is not to arrive at the grave safely in a well preserved body, but rather to skid in sideways, totally worn out, shouting: "Wow what a ride!"
I still play Cowboys and Bad Guys but now I use real bullets. CAS
I still play Cowboys and Bad Guys but now I use real bullets. CAS
Re: E-47 message on remstar pro2 machine
Any chance you got some water in it?
A little searching will turn up a number of previous discussions about that error.
search.php?keywords=E47
Den
A little searching will turn up a number of previous discussions about that error.
search.php?keywords=E47
Den