Popping noise from valve

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Ms.Snuffleupagus
Posts: 281
Joined: Sun Nov 22, 2009 6:28 pm
Location: Canada

Re: Popping noise from valve

Post by Ms.Snuffleupagus » Sun Nov 29, 2009 11:25 am

Thanks all,
I lowered the humidity and put the hose under my blanket as suggested, no problems with popping/fluttering noise last night.

Tristan

Re: Popping noise from valve

Post by Tristan » Thu Aug 11, 2011 4:35 pm

It's interesting that several of you mentioned there is water in the tube, as my machine was also making really loud 'popping' noises that sounded like you were hitting the end of a hollow tube. I disconnected the mask, and the popping continued even for a while after the machine was turned off and it was 'cooling down'- but stopped when I put my finger in the tube (ie. blocked the tube) as the sound was really loud and annoying. I ended up sleeping half the night without the CPAP on, which resulted in a terrible nights sleep ! The only thing I've done differently, is that I'm staying at a friends place and I just threw in the standard slimline tube, when I usually use the heated tube- so it's entirely possible there was water within the tube, although it seems strange that water could make such a loud popping sound- I wonder what the water was impacting on to make this sound ? It's happened to me once before- but I can't remember if I was using the heated tube at the time or not- but it's definately a possibility and thank you for all your suggestions

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Popping noise from valve

Post by archangle » Thu Aug 11, 2011 11:53 pm

yobtaf wrote:This morning my wife and I woke up to the mask making a popping noise that sounded like it was coming from the valve. It sounded like when you would take your palm and hit it on the end of a pipe. very strange. I have noticed that sometimes when I lay down at first and the machine is on the ramp mode, that the mask will suck hard to my face when I breathe. I dont know if this is all related and there is a problem with the mask or valve.
Put on the mask, disconnect the hose from the mask. Plug the pipe where the hose connects with your thumb. When you inhale, the valve should open and let air in. If not, you have a problem with the valve. When you exhale, the valve may or may not close, depending if you get the pressure high enough. You may hear some noises.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

SleepyMcSleepFace
Posts: 1
Joined: Thu May 30, 2019 5:26 am

Re: Popping noise from valve

Post by SleepyMcSleepFace » Thu May 30, 2019 5:38 am

I had a similar issue with my machine: if I took a big breath in, the mask would get sucked on to my face, and then when I breathed out, the valve would go THUMP!

The problem was that not enough air was getting through the machine into my mask. In my case, this was because the little dust filter was full and needed to be replaced. I imagine that any other problem that blocks airflow through the machine or hose would have a similar effect.

User avatar
LSAT
Posts: 13319
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Popping noise from valve

Post by LSAT » Thu May 30, 2019 9:29 am

8 year old post

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Popping noise from valve

Post by Goofproof » Thu May 30, 2019 10:17 am

Cleaning better late than never, not only is your breathing hurt, the machine counts on air flow to cool it down. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Popping noise from valve

Post by chunkyfrog » Thu May 30, 2019 11:08 am

SleepyMcSleepFace wrote:
Thu May 30, 2019 5:38 am
I had a similar issue with my machine: if I took a big breath in, the mask would get sucked on to my face, and then when I breathed out, the valve would go THUMP!

The problem was that not enough air was getting through the machine into my mask. In my case, this was because the little dust filter was full and needed to be replaced. I imagine that any other problem that blocks airflow through the machine or hose would have a similar effect.
Try READING the responses on the first page.
I lurked for 4 months before joining.
Reading is FUNdamental--really!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

travistheory
Posts: 1
Joined: Sun Jun 23, 2019 2:13 am

Re: Popping noise from valve

Post by travistheory » Sun Jun 23, 2019 2:26 am

  • PLEASE STOP COMPLAINING ABOUT PEOPLE POSTING ON VERY OLD THREADS! Its 2019 and its 1am and this popping noise is getting on my freakin nerves! I googled the issue and UP YOUR SITE CAME. So just because some guy had a question and had a singular answer, doesnt mean people need to stop answering, even years later, because issues change. Especially as companies come out with inferior versions of a product, or updated versions of something but you still have an old product, needing old answers.. i am proof. I dont use water at all, so all the moisture comments would mean nothing to me. So i needed the ones about taking off the hose and seeing if you can resolve the popping noise. An answer given years later by someine else. Some sites close old posts based on soneone demanding it be such,
    And ruin google searches for answers. Also, making a new thread wont work, because those same internet board police will say, "this topic was already discussed" and have your question closed! So just be quiet, let humans help humans, and stop trying to police everyone.iits not like companies offer 24hour help on all the thing we buy...

User avatar
Pugsy
Posts: 64942
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Popping noise from valve

Post by Pugsy » Sun Jun 23, 2019 6:52 am

travistheory wrote:
Sun Jun 23, 2019 2:26 am
  • PLEASE STOP COMPLAINING ABOUT PEOPLE POSTING ON VERY OLD THREADS! Its 2019 and its 1am and this popping noise is getting on my freakin nerves! I googled the issue and UP YOUR SITE CAME. So just because some guy had a question and had a singular answer, doesnt mean people need to stop answering, even years later, because issues change. Especially as companies come out with inferior versions of a product, or updated versions of something but you still have an old product, needing old answers.. i am proof. I dont use water at all, so all the moisture comments would mean nothing to me. So i needed the ones about taking off the hose and seeing if you can resolve the popping noise. An answer given years later by someine else. Some sites close old posts based on soneone demanding it be such,
    And ruin google searches for answers. Also, making a new thread wont work, because those same internet board police will say, "this topic was already discussed" and have your question closed! So just be quiet, let humans help humans, and stop trying to police everyone.iits not like companies offer 24hour help on all the thing we buy...
People who don't use a humidifier still exhale moisture in their exhale breath and condensation can still happen if conditions are just right and still cause popping noises simply from this moisture condensing. It's very common and that is why it is mentioned so often...
That needs to be ruled out first because the fix is so easy.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Popping noise from valve

Post by palerider » Sun Jun 23, 2019 1:14 pm

travistheory wrote:
Sun Jun 23, 2019 2:26 am
  • PLEASE STOP COMPLAINING ABOUT PEOPLE POSTING ON VERY OLD THREADS! Its 2019 and its 1am and this popping noise is getting on my freakin nerves! I googled the issue and UP YOUR SITE CAME. So just because some guy had a question and had a singular answer, doesnt mean people need to stop answering, even years later, because issues change. Especially as companies come out with inferior versions of a product, or updated versions of something but you still have an old product, needing old answers.. i am proof. I dont use water at all, so all the moisture comments would mean nothing to me. So i needed the ones about taking off the hose and seeing if you can resolve the popping noise. An answer given years later by someine else. Some sites close old posts based on soneone demanding it be such,
    And ruin google searches for answers. Also, making a new thread wont work, because those same internet board police will say, "this topic was already discussed" and have your question closed! So just be quiet, let humans help humans, and stop trying to police everyone.iits not like companies offer 24hour help on all the thing we buy...
PLEASE STOP COMPLAINING ABOUT PEOPLE COMPLAINING ABOUT POSTING ON VERY OLD THREADS!
Your noise adds nothing,

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Popping noise from valve

Post by Wulfman... » Sun Jun 23, 2019 1:49 pm

travistheory wrote:
Sun Jun 23, 2019 2:26 am
  • PLEASE STOP COMPLAINING ABOUT PEOPLE POSTING ON VERY OLD THREADS! Its 2019 and its 1am and this popping noise is getting on my freakin nerves! I googled the issue and UP YOUR SITE CAME. So just because some guy had a question and had a singular answer, doesnt mean people need to stop answering, even years later, because issues change. Especially as companies come out with inferior versions of a product, or updated versions of something but you still have an old product, needing old answers.. i am proof. I dont use water at all, so all the moisture comments would mean nothing to me. So i needed the ones about taking off the hose and seeing if you can resolve the popping noise. An answer given years later by someine else. Some sites close old posts based on soneone demanding it be such,
    And ruin google searches for answers. Also, making a new thread wont work, because those same internet board police will say, "this topic was already discussed" and have your question closed! So just be quiet, let humans help humans, and stop trying to police everyone.iits not like companies offer 24hour help on all the thing we buy...
It gets on MY freakin' nerves that people post on the forum, asking questions and complaining........and don't tell us what the HELL equipment they're using!!!

Some years back, there were a bunch of complaints about the ResMed S9 making "clicking" or "popping" noises.
Do you use an S9?

search.php?keywords=%22popping+noise%22

search.php?keywords=%22clicking+noise%22

And, for what it's worth, I DON'T have a problem with people resurrecting old threads if their problems are related to that subject.

So, instead of ripping on everybody else, fill out your profile!


Den

.
Last edited by Wulfman... on Sun Jun 23, 2019 4:01 pm, edited 1 time in total.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
zonker
Posts: 11245
Joined: Fri Jun 19, 2015 4:36 pm

Re: Popping noise from valve

Post by zonker » Sun Jun 23, 2019 3:31 pm

travistheory wrote:
Sun Jun 23, 2019 2:26 am
  • PLEASE STOP COMPLAINING ABOUT PEOPLE POSTING ON VERY OLD THREADS!
meerkat.gif


bombayone, is that you?
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Popping noise from valve

Post by chunkyfrog » Sun Jun 23, 2019 9:46 pm

I wonder why it is so hard for people to LOOK.
Look at the equipment--turn the lights on--look really hard.
Most people can figure stuff out if they just take a good look at it.
It is not really that hard.
Even a frog can do it.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Popping noise from valve

Post by palerider » Mon Jun 24, 2019 12:34 am

chunkyfrog wrote:
Sun Jun 23, 2019 9:46 pm
I wonder why it is so hard for people to LOOK
...
Even a frog can do it.
Frogs are much more intelligent than most people.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.