Battery Usage Time
- BleepingBeauty
- Posts: 2454
- Joined: Thu Apr 02, 2009 5:30 pm
- Location: Aridzona ;-)
Battery Usage Time
I think it would be helpful to have a thread where battery usage times can be compared (for anyone contemplating backup power for their machine). If you have some sort of backup power device, please tell us what you use and how much time you can get from a full charge on that device (along with your pressure setting). Also let us know whether or not you're powering a heated humidifier with the backup power source.
I tested my backup battery (the C-222 from batterygeek.net) sometime last year and was able to get three nights out of it (with the heated humidifier), but I can't remember what pressure setting I had at the time. So I decided to retest the battery a couple of nights ago. Unfortunately, it occurred to me this morning that I'm now using the HC150 humidifier, and last year's test of the battery occurred while I was using the M-Series integrated humidifier; so the battery has not actually powered the HC150 for the past two nights, and the current results can't be fairly compared to my earlier experiment.
I exchanged the HC150 for my old integrated humidifier this morning, and I'm recharging my battery to full power today so I can re-start the test tonight. I'll post my results in a few days.
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
So, if you use a backup power source:
1. What do you use?
2. What's your pressure setting(s)?
3. How much time can you get from a full charge?
4. Are you powering a heated humidifier, as well?
I tested my backup battery (the C-222 from batterygeek.net) sometime last year and was able to get three nights out of it (with the heated humidifier), but I can't remember what pressure setting I had at the time. So I decided to retest the battery a couple of nights ago. Unfortunately, it occurred to me this morning that I'm now using the HC150 humidifier, and last year's test of the battery occurred while I was using the M-Series integrated humidifier; so the battery has not actually powered the HC150 for the past two nights, and the current results can't be fairly compared to my earlier experiment.
I exchanged the HC150 for my old integrated humidifier this morning, and I'm recharging my battery to full power today so I can re-start the test tonight. I'll post my results in a few days.
~~~~~~~~~~~~~~~~~~~~~~~~~~~~~
So, if you use a backup power source:
1. What do you use?
2. What's your pressure setting(s)?
3. How much time can you get from a full charge?
4. Are you powering a heated humidifier, as well?
Veni, vidi, Velcro. I came, I saw, I stuck around.
Dx 11/07: AHI 107, central apnea, Cheyne Stokes respiration, moderate-severe O2 desats. (Simple OSA would be too easy.
)
PR S1 ASV 950, DreamWear mask, F&P 150 humidifier, O2 @ 2L.
Dx 11/07: AHI 107, central apnea, Cheyne Stokes respiration, moderate-severe O2 desats. (Simple OSA would be too easy.

PR S1 ASV 950, DreamWear mask, F&P 150 humidifier, O2 @ 2L.
Re: Battery Usage Time
/JMMO/ - I think in most cases (power outages) one will be much better served to not use any power for any humidifier. Esp. if you don't know how long the power will be out. The humidifier can be used in the "Passover Mode" during these brief periods w/o commercial power. /Just My Meaningless Opinion/
_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Additional Comments: New users can't remember they can't remember YET! |
BeganCPAP31Jan2007;AHI<0.5
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
I have no doubt, how I sleep affects every waking moment.
I am making progress-NOW I remember that I can't remember

If this isn’t rocket science why are there so many spaceshots?
Be your own healthcare advocate!
Re: Battery Usage Time
Hi,
Just a couple of thoughts on what would be "useful information" to collect.
1. What Battery are you using?
2. How many nights did you use it between recharging the battery?
3. Number of Hours used per night while on Battery power?
4. What PAP device are you using?
5. What is your Pressure Setting(s).
6. What Humidifier are you using (if any)?
7. What is the Humidifier Setting that you used?
8. What Inverter (if any) did you use to convert battery power to AC power?
9. If an Inverter was used did it generate 110V or 220V AC Voltage?
Just a couple of thoughts on what would be "useful information" to collect.
1. What Battery are you using?
2. How many nights did you use it between recharging the battery?
3. Number of Hours used per night while on Battery power?
4. What PAP device are you using?
5. What is your Pressure Setting(s).
6. What Humidifier are you using (if any)?
7. What is the Humidifier Setting that you used?
8. What Inverter (if any) did you use to convert battery power to AC power?
9. If an Inverter was used did it generate 110V or 220V AC Voltage?
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Additional Comments: Encore Pro, Backup-M Series BiPAP Auto. Set as ABiPAP 15I/9E AHI 0.2. Normally operates @ 12I/9E 1987 OSA diagnosis |
If your never go off on a tangent, you are doomed to live your life going around in circles.
Re: Battery Usage Time
Mseries auto /cflex 1 12cm(h2o implied)
Optima BlueTop 55AH
no HH /passover
straight 12v w/adapter
4 nights ~8hr/night
it would have gone another night maybe 2
I'm w/"I'm GumbyCT, damn'it" on this one,
the few nights I would need this I can live without HH.
Optima BlueTop 55AH
no HH /passover
straight 12v w/adapter
4 nights ~8hr/night
it would have gone another night maybe 2
I'm w/"I'm GumbyCT, damn'it" on this one,
the few nights I would need this I can live without HH.
"If your therapy is improving your health but you're not doing anything
to see or feel those changes, you'll never know what you're capable of."
I said that.
to see or feel those changes, you'll never know what you're capable of."
I said that.
- Lou Cypher
- Posts: 347
- Joined: Mon Aug 17, 2009 3:51 pm
- Location: Pittsburgh
Re: Battery Usage Time
I for one would like to give a BIG THANK YOU for this post. I am looking into buying a battery backup and this will be most welcomed info....
Might I suggest to the powers that be, Maybe even make a post like this a sticky so anyone in the future can easily lookup fact and figures on Battery Backups.
Might I suggest to the powers that be, Maybe even make a post like this a sticky so anyone in the future can easily lookup fact and figures on Battery Backups.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: Heated tubing, Chin strap, Zero Gravity, memory foam adjustable Bed. |
Re: Battery Usage Time
I agree that this is a useful exercise. However, a confounding question is how many times are you able to get that many nights out of your battery before the battery is damaged.
Batteries may be able to be discharged to a very low voltage a handful of times (that is, they may get you a full two, three, four or whatever nights use on one charge), but such deep discharge cycles are damaging to the battery. How damaging depends on the type of battery. But in any case, you will reduce the life of the battery, potentially significantly.
Batteries may be able to be discharged to a very low voltage a handful of times (that is, they may get you a full two, three, four or whatever nights use on one charge), but such deep discharge cycles are damaging to the battery. How damaging depends on the type of battery. But in any case, you will reduce the life of the battery, potentially significantly.
I'm workin' on it.
Re: Battery Usage Time
I really think everyone should periodically give thier backup battery an "acid" test. Differences between machines, changes in pressure, and age of the battery can all effect how much time you get out of it. I recharge the battery whenever I use it, it's my understanding that if the battery doesn't have a "memory", continually draining it will shorten its life. Part of your monthly maintennce should include "topping off" your backup battery. Also, remember that some humidifiers can not be run on DC.
My machine is a RemStar M Series. During a "power emergncy" (or camping) if I use the humidifer, it is in passover mode. Last time I tested, my pressure was 10. My nights are 6 to 8 hours.
I own both a Black & Deck 400 and a Battery Geek.
The Black & Decker on DC I got 4 nights and most of the 5th on DC, not quite 2 nights using the AC. I've never pushed the Battery Geek to the limit, but I can easily get 2 nights out of it.
My machine is a RemStar M Series. During a "power emergncy" (or camping) if I use the humidifer, it is in passover mode. Last time I tested, my pressure was 10. My nights are 6 to 8 hours.
I own both a Black & Deck 400 and a Battery Geek.
The Black & Decker on DC I got 4 nights and most of the 5th on DC, not quite 2 nights using the AC. I've never pushed the Battery Geek to the limit, but I can easily get 2 nights out of it.
-
- Posts: 1
- Joined: Mon Sep 28, 2009 12:52 pm
Re: Battery Usage Time
I have the ResMed S8 Elite II without humidifier. I recently purchase what is called a gobe power system. It includes a 12V 14.5 amp battery with solar panels along with an AC power cord to charge the battery from the wall. It also comes with a DC cord with male and female adapters. Where I get the most benefit is from the gobe solar briefcase. When I camp/travel I can charge the battery during the day and get a good 10 hours of usage when I sleep at night. Pretty inexpensive when you consider that you get the whole kit for $180. From everywhere I researched the best price I found for this is at http://www.wattme.com.
BTW-the battery has an integrated inverter so you can also run devices from it that require a USB, AC, or DC connection.
BTW-the battery has an integrated inverter so you can also run devices from it that require a USB, AC, or DC connection.
- bearded_two
- Posts: 459
- Joined: Mon Aug 10, 2009 8:01 pm
Re: Battery Usage Time
The blue and yellow top Optimas are AGM batteries that are designed to withstand deep discharge with little or no damage. The red top Optimas can also withstand deep discharge with little damage, but they are intended primarily as a starting battery. I have a red top Optima as my car starting battery and yellow top Optima in my car trunk.
The Battery Geek CPAP battery uses Li ion batteries that easily withstand deep discharge.
carbonman was using a 55 Ah battery for four nights, 8 hours a night, and I assume that his CPAP is drawing about an Amp. 4 * 8 * 1 = 32 Ah. His battery is rated at 55 Ah; 55 - 32 = 23 Ah reserve. That battery will take that sort of discharge without even breathing heavy. My guess is that he could run his CPAP for six nights without damaging the battery. 6 * 8 * 1 = 48 Ah and 55 - 48 = 7 Ah reserve (over 12%)
The Battery Geek CPAP battery uses Li ion batteries that easily withstand deep discharge.
carbonman was using a 55 Ah battery for four nights, 8 hours a night, and I assume that his CPAP is drawing about an Amp. 4 * 8 * 1 = 32 Ah. His battery is rated at 55 Ah; 55 - 32 = 23 Ah reserve. That battery will take that sort of discharge without even breathing heavy. My guess is that he could run his CPAP for six nights without damaging the battery. 6 * 8 * 1 = 48 Ah and 55 - 48 = 7 Ah reserve (over 12%)
Re: Battery Usage Time
My results are in!
Using this battery: http://www.bestcpapprice.com/Owell-Indu ... 0-516.html
and this ResMed converter: http://www.bestcpapprice.com/ResMed-S8- ... 3-335.html
Machine: ResMed S8 Elite II
Pressure: 12cm, EPR=3 fulltime, no ramp
Humidifier: passover mode
I got 25.5 hours!
Using this battery: http://www.bestcpapprice.com/Owell-Indu ... 0-516.html
and this ResMed converter: http://www.bestcpapprice.com/ResMed-S8- ... 3-335.html
Machine: ResMed S8 Elite II
Pressure: 12cm, EPR=3 fulltime, no ramp
Humidifier: passover mode
I got 25.5 hours!
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: CPAP 6/5/2009, Rx: 11-16, OSCAR |
- BleepingBeauty
- Posts: 2454
- Joined: Thu Apr 02, 2009 5:30 pm
- Location: Aridzona ;-)
Re: Battery Usage Time
Good deal.twokatmew wrote:My results are in!
I got 25.5 hours!
I'm sorry to report that my test has been derailed temporarily. I think my battery charger is faulty, as the power "brick" in the middle of the cord gets hot to the touch (too hot for comfort) within about an hour of plugging it in. So I need to replace that before I can effectively recharge my battery and repeat the test.
I hope it's just the charger that's defective...
Veni, vidi, Velcro. I came, I saw, I stuck around.
Dx 11/07: AHI 107, central apnea, Cheyne Stokes respiration, moderate-severe O2 desats. (Simple OSA would be too easy.
)
PR S1 ASV 950, DreamWear mask, F&P 150 humidifier, O2 @ 2L.
Dx 11/07: AHI 107, central apnea, Cheyne Stokes respiration, moderate-severe O2 desats. (Simple OSA would be too easy.

PR S1 ASV 950, DreamWear mask, F&P 150 humidifier, O2 @ 2L.
- bearded_two
- Posts: 459
- Joined: Mon Aug 10, 2009 8:01 pm
Re: Battery Usage Time
I have a Battery Geeks 2200 and get two 8 hour nights use out of it, plus about five hours the third night -- 21 hours. I get the same with either an IntelliPAP or an Auto M set at 11 cmh2o.
Re: Battery Usage Time
This is an excellent battery solution and I have used it for 3 nights several times. It weighs about 13 lbs, will fit in my motorcycle saddlebag and allows me to go camping with the Boy Scouts. I remove the battery charger from the front pocket when I travel and my M-Series cpap fits right into the battery case. Cpap setting 8-12, no humidifier.
https://www.cpap.com/productpage/respir ... -pack.html
This is an excellent battery solution I just used for an overnight backpacking trip. Also provided excellent service and met my critical shipping date to make a Boy Scout trip. It weighs only 20 oz. and is the size of a crackerjack box. I used it for 8 hours at Cpap setting 8-12, no humidifier. I plan on this being for one night only.
http://www.batteryspace.com/polymerli-i ... arger.aspx
https://www.cpap.com/productpage/respir ... -pack.html
This is an excellent battery solution I just used for an overnight backpacking trip. Also provided excellent service and met my critical shipping date to make a Boy Scout trip. It weighs only 20 oz. and is the size of a crackerjack box. I used it for 8 hours at Cpap setting 8-12, no humidifier. I plan on this being for one night only.
http://www.batteryspace.com/polymerli-i ... arger.aspx
Last edited by Georgio on Sun Oct 11, 2009 9:15 pm, edited 2 times in total.
DreamStation 2, Oscar
Resmed AirFit P30i Nasal Mask
Resmed AirFit P30i Nasal Mask
Re: Battery Usage Time
So, at 14 cm pressure how many amp-hours would I need to get three successful nights on a ResMed S8 with no HH (passover only)?
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: 15-18 cm, EPR 1, PAPcap |
Sleep well and live better!
- bearded_two
- Posts: 459
- Joined: Mon Aug 10, 2009 8:01 pm
Re: Battery Usage Time
From the chart below, a CPAP set at 14 cmh2o should last about 12 hours with an 18.5 Ah battery and about 8 hours with a 12.5 Ah battery. You would probably be safe figuring that you need about 1.6 Ah per hour. A 12 Ah battery will be exhausted after 8 hours, I would conservatively figure that an 18 Ah battery would last one night. If using a lead-acid battery, I would figure at least 20 Ah per night. This is a starting point for testing, and your mileage may vary.
The following chart is adapted from
http://www.batterygeek.net/v/vspfiles/S ... _222Wh.asp
Estimated CPAP Lithium Ion Battery Usage Chart
CPAP Usage Time in Hours
Pressure.....Battery....Capacity
cm/H2O.....18.5Ah.....12.5Ah
................Hours......Hours
....4............25..........18
....7...........18.8.......12.7
...10...........14.3........9.7
...13...........12.1........8.2
...16...........10.5........7.1
...20............9.8........6.6
All Times Approximate, For lead acid batteries, allow at least a 25% reserve to help preserve battery life.
The following chart is adapted from
http://www.batterygeek.net/v/vspfiles/S ... _222Wh.asp
Estimated CPAP Lithium Ion Battery Usage Chart
CPAP Usage Time in Hours
Pressure.....Battery....Capacity
cm/H2O.....18.5Ah.....12.5Ah
................Hours......Hours
....4............25..........18
....7...........18.8.......12.7
...10...........14.3........9.7
...13...........12.1........8.2
...16...........10.5........7.1
...20............9.8........6.6
All Times Approximate, For lead acid batteries, allow at least a 25% reserve to help preserve battery life.