How can I view my AHI and leak data on a Remstar M? I can see my hours of use and 7- and 30-day averages, but that's it.
Thanks,
kal
View AHI settings on Remstar M
Re: View AHI settings on Remstar M
mine either. that's set by the clinic i believe. time to save up for the software.
Re: View AHI settings on Remstar M
Which model? If it's a "Plus", there is nothing to see because it doesn't record anything but the hours slept.Kalnel2 wrote:How can I view my AHI and leak data on a Remstar M? I can see my hours of use and 7- and 30-day averages, but that's it.
Thanks,
kal
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: View AHI settings on Remstar M
If your machine is the "Pro" as listed in your profile, this will work.steveeas wrote:mine either. that's set by the clinic i believe. time to save up for the software.
The "Show AHI/Leak Feature" needs to be set to "On".
Den
Menu for M Series Pro
1. Hold down the <- -> buttons while plugging in the power on the back, wait for 2 beeps, release buttons.
2. Press the + key. <- -> buttons move to next field, -/+ keys decrement/increment
3. Check the following field(s):
-Therapy Mode = CPAP
-CPAP Pressure = NN (4 - 20 cm.)
-C-Flex Setting = (options are off, 1, 2 or 3)
-Ramp Time = (options are 05 to 45 min)
-Ramp Start Pressure = cm (Min. 4cm)
-Mask Alert Feature = (On/Off)
-Auto Off Feature = Off (On/off)
-Show AHI/Leak Feature = On (On/Off)
Press On/Off button to exit.
==========Resetting LCD Display Data====================
To reset LCD Therapy data (does NOT erase any data from SmartCard):
1. Hold down the (<-) (->) buttons while plugging in the power on the back,
wait for 2 beeps or 5 seconds, release buttons.
2. Press the (->) key and navigate to Therapy screen.
3. Press and HOLD down the (-) minus key for 5 seconds and/or until therapy data is reset to Zero.
4. Press the (->) key to scroll back to main menu, press (+) key to enter Setup or
On/Off button to exit.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: View AHI settings on Remstar M
Thanks, I was able to view the data that way -- instructions were great!
I have an AHI total of 0.6 for both 7 and 30 days, and a leak rate of 20 minutes. I'm imaging much the the "leak time" is due to my fiddling around with the mask (or taking it off to get up in the night.)
I have an AHI total of 0.6 for both 7 and 30 days, and a leak rate of 20 minutes. I'm imaging much the the "leak time" is due to my fiddling around with the mask (or taking it off to get up in the night.)