Is ResMed EPR needed ??

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Spirit
Posts: 433
Joined: Fri Aug 07, 2009 9:27 pm
Location: Central Tx

Is ResMed EPR needed ??

Post by Spirit » Sat Aug 29, 2009 8:54 pm

Hi,
I was wonderfing if EPR is needed ? Here's my cpap data:
EPR: Fulltime
EPR Level: 3
ERP inhale: Med
EPR pressure: 5.8

If I am prescribed pressure 8cm, shouldn't pressure be increased to compensate for EPR ? Like 11 cm, to actually recieve 8cm ? Would a person recieve, better level of treatment straight cpap ?

Thanks for any help.

gailandartie
Posts: 87
Joined: Tue Jun 24, 2008 6:46 pm
Location: United States

Re: Is ResMed EPR needed ??

Post by gailandartie » Sat Aug 29, 2009 9:05 pm

Hi,

I have the same machine and I have turned EPR off. I found that EPR made the machine noisier . My settings are min 8 and max 10.5 . At these settings I found I was more comfortable without it, my numbers are great and the machine is so quiet.

Since this machine is an APAP, you need to know your min and max settings. Do you have any data - AHI, AI, HI? You need to know this in order to adjust pressure properly.

_________________
Mask: Swift™ FX For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: CMS Model 60-C Pulse Oximeter

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: Is ResMed EPR needed ??

Post by jdm2857 » Sat Aug 29, 2009 11:35 pm

I can tell you that ResMed does not think it is essential.

The ResMed S8 AutoSet Spirit II (the international version of the AutoSet) does not have EPR. I think that ResMed thinks that their EasyBreathe technology is fine without EPR, but to compete with Respironic's C-Flex and A-Flex in North America, they needed a similar feature with settings for their advertising.\

I've never turned EPR on.
jeff

User avatar
Spirit
Posts: 433
Joined: Fri Aug 07, 2009 9:27 pm
Location: Central Tx

Re: Is ResMed EPR needed ??

Post by Spirit » Sun Aug 30, 2009 2:39 pm

Thanks for reply. I experimented with EPR last night, turned it off. My HI & leak value improved, but AI increased.
Lk: .26 > .12
AI: 0.3 > 1.3
HI: 5.5 > 3.0

Noticed after I stopped using Afrin nasal spray, my HI's were increasing, but more pressure decreased the HI's. Something I need to discuss with sleep doc. Will need to try no EPR for a few more nights. Will try with Afrin tonight and see what happens, will post update.

User avatar
twokatmew
Posts: 606
Joined: Wed Jun 03, 2009 6:06 pm
Location: Mid-Michigan, US

Re: Is ResMed EPR needed ??

Post by twokatmew » Sun Aug 30, 2009 3:10 pm

I was using an EPR of 1, and when I turned it off, AIs increased while HIs decreased on most days. The other night I set EPR at 3, and so far both AI and HI are a bit better. Also, I stopped puffing air out my mouth, and I took a nap today w/o tape -- no mouth leaks!

Some users say they'd increase pressure by the number of the EPR setting, but ResMed claims EPR provides exhalation relief (as merely a comfort feature) and doesn't affect inhalation pressure. They also claim that CPAP with EPR is just as effective as CPAP without EPR.

When I ran my machine in Autoset mode with and without EPR (using the same min and max pressures), my 95% pressure was the same.

Everyone's different, but the couple days I've set EPR at 3, I seem to be doing a bit better with no increase in pressure.

The machine is indeed a bit louder on inhale with EPR at 3, but I have slept better regardless so far.

My doc is wanting to do a BiPap trial hoping that I might do better on it than A/CPAP. That gave me the idea to up my EPR setting in the meantime. Whether it'll actually make a difference ... or is merely showing some improvement from the power of suggestion I don't know. Time will tell ... unless I get switched to BiPap before I can gather more data!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: CPAP 6/5/2009, Rx: 11-16, OSCAR

User avatar
Spirit
Posts: 433
Joined: Fri Aug 07, 2009 9:27 pm
Location: Central Tx

Re: Is ResMed EPR needed ??

Post by Spirit » Sun Aug 30, 2009 3:56 pm

Hi twokatmew,
Thanks for reply. It really helps to do comparisons & learn experiences of others. You had the same experience I did, turning off EPR. If no EPR setting doesn't work, I'll go back to level 3,. Maybe constent pressure, throws off breathing rythem. With no EPR my mouth inflates like a balloon.

Thanks for sharing experience.

User avatar
alexmihaila
Posts: 10
Joined: Fri Jun 12, 2009 9:54 am
Location: Romania
Contact:

Re: Is ResMed EPR needed ??

Post by alexmihaila » Mon Aug 31, 2009 1:39 am

Sory jdm but your info is outdated; ResMed S8 AutoSet Spirit II (the international version of the AutoSet) does HAVE EPR ( i have this machine, have EPR printed on the machine and EPR works on both CPAP and APAP modes)

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Eson™ Nasal CPAP Mask with Headgear
Additional Comments: titrated CPAP 10; Setting APAP 10-15; EPR 1
Alex

Alive now!

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: Is ResMed EPR needed ??

Post by jdm2857 » Mon Aug 31, 2009 8:56 am

When did they add it? It wasn't listed on resmed.com.au (australian site) when I checked about 2 months ago.
jeff

User avatar
alexmihaila
Posts: 10
Joined: Fri Jun 12, 2009 9:54 am
Location: Romania
Contact:

Re: Is ResMed EPR needed ??

Post by alexmihaila » Tue Sep 01, 2009 4:34 am

I've got one just the day I joined.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Eson™ Nasal CPAP Mask with Headgear
Additional Comments: titrated CPAP 10; Setting APAP 10-15; EPR 1
Alex

Alive now!

ozij
Posts: 10515
Joined: Fri Mar 18, 2005 11:52 pm

Re: Is ResMed EPR needed ??

Post by ozij » Tue Sep 01, 2009 4:52 am

Resmed -- and Respironics too sell different models, with different names in different parts of the world.
Resmed.au will not show the equipment available in the States - you have to have to choose the country to get reliable info.


O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: Is ResMed EPR needed ??

Post by jdm2857 » Tue Sep 01, 2009 8:11 am

Ozij --

I was discussing the ResMed S8 AutoSet Spirit II. That's why the reference to the Australian site.
jeff

ozij
Posts: 10515
Joined: Fri Mar 18, 2005 11:52 pm

Re: Is ResMed EPR needed ??

Post by ozij » Tue Sep 01, 2009 11:12 pm

jdm2857 wrote:Ozij --

I was discussing the ResMed S8 AutoSet Spirit II. That's why the reference to the Australian site.
You assumed the international version of the Autoset Spirit II was described in the Ausralian site:
jdm2857 wrote:I can tell you that ResMed does not think it is essential.

The ResMed S8 AutoSet Spirit II (the international version of the AutoSet) does not have EPR. I think that ResMed thinks that their EasyBreathe technology is fine without EPR, but to compete with Respironic's C-Flex and A-Flex in North America, they needed a similar feature with settings for their advertising.\

I've never turned EPR on.


(my emphasis added to the quote)

This is the description on the international site:

http://www.resmed.com/int/products/s8_a ... s&sec=true
Easy-Breathe waveform: a truly natural breathing experience
ResMed has added the new Easy-Breathe waveform to the clinically effective EPR to create an exceptional pressure relief system in both fixed CPAP and AutoSet modes.

Throughout the night and breath by breath, the Easy-Breathe waveform softens the pressure upon exhalation and inhalation, creating a gentle breathing rhythm that adapts to the patient's natural breathing cycle.

No other system offers such synchronisation. With pressure cycles and patient breathing more closely synchronised, breathing has never felt more natural. All the advantages of EPR and AutoSet—combined in a single device.
(my emphasis added to the quote).


The international version of the Autoset Spirit II is described on Resmesd's international site and does does have EPR.

O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: Is ResMed EPR needed ??

Post by jdm2857 » Wed Sep 02, 2009 8:58 am

I didn't realize that ResMed makes multiple versions of machines with the same name. I'll make a note of the international site.

I do know that at least one owner of one version of the S8 AutoSet Spirit II posted that it did not have EPR.

Thanks, O.
jeff

ozij
Posts: 10515
Joined: Fri Mar 18, 2005 11:52 pm

Re: Is ResMed EPR needed ??

Post by ozij » Wed Sep 02, 2009 9:36 am

I think I know who you're thinking about - an Australian who was on of the first to have an Autoset II...
O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023