CPAP machines/Masks

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Dixie62
Posts: 22
Joined: Sun Jul 26, 2009 10:48 pm

CPAP machines/Masks

Post by Dixie62 » Sun Aug 02, 2009 10:23 pm

Do all masks work with all CPAP machines. If not how do you know what mask will work with what machine... Sorry for the questions I am such a newbie.. I take my machine/mask back to doctor on Tuesday and then after they download the info I will decide what I will do about getting one..
Thank you for any comments..
When Jesus died on the cross; he was thinking of you!

User avatar
Hawthorne
Posts: 3972
Joined: Tue Oct 19, 2004 4:46 am
Location: London Ontario -Canada

Re: CPAP machines/Masks

Post by Hawthorne » Mon Aug 03, 2009 4:29 am

All masks will work with all cpap machines. With Resmed machines, you need to program in the mask you are using. With Respironics machines, there is no need to program in the mask in any way.

Good luck.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments:  Backups- FX Nano masks. Backup machine- Airmini auto travel cpap

User avatar
YouMaySayImaDreamer
Posts: 27
Joined: Tue Jun 16, 2009 6:46 am
Location: Seoul, Korea

Re: CPAP machines/Masks

Post by YouMaySayImaDreamer » Mon Aug 03, 2009 5:25 am

With Resmed machines, you need to program in the mask you are using.
Hi,

I'm just curious about the above. Does that mean ResMed machines adjust for certain masks, or is it just for generating reports?

Thanks,

YM
I hope some day you'll join us
And the world will sleep as one

User avatar
Pugsy
Posts: 65295
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: CPAP machines/Masks

Post by Pugsy » Mon Aug 03, 2009 6:05 am

YouMaySayImaDreamer wrote: With Resmed machines, you need to program in the mask you are using.



Hi,

I'm just curious about the above. Does that mean ResMed machines adjust for certain masks, or is it just for generating reports?

It is just for the reporting. ResMed machines subtract the intentional leak rate (normal exhaust flow rate) for the mask chosen before they report the leak. They do not alter therapy, pressure, anything in regards to actual therapy.
It is strictly a numbers thing. Respironics machines don't do this so we have to do the math for ourselves. Not a big deal. ResMed machines just try to help out by doing the math for you. They offer ResMed brand choices though and if you use something else you have to use the plain standard choice which may not have the same flow rate as your mask does at your pressure.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: CPAP machines/Masks

Post by jdm2857 » Mon Aug 03, 2009 9:16 am

There are other things you can do for a non-ResMed mask with a ResMed machine. If you compare the vent rate of your mask at your pressure (range) with the leak rates of the various ResMed masks, you can probably find one that is pretty close and set your machine for that mask. Then you'll get good numbers.
jeff