one more newbie

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
wearysoul
Posts: 53
Joined: Wed Jun 17, 2009 1:37 pm
Location: North Carolina

one more newbie

Post by wearysoul » Fri Jun 19, 2009 8:34 am

Another newbie here—a 53-year-old male. I weigh about 152 lbs and am 5 ft 10 in tall. I have mild/moderate sleep apnea. The prescribed pressure is 6 cm.

I usually fall asleep quickly. But on my first night, I couldn’t sleep. I had trouble getting my headgear and mask adjusted properly. I don’t think I ever really got it right, for I kept hearing and feeling leaks. Seems I’ve read that some leakage is normal, though.

I also had trouble breathing. At times it felt as if the machine wanted me to breathe faster than I wanted to. We just weren’t in sync. I couldn’t seem to relax into a normal breathing pattern. I’m wondering if some of that may have been due to the 15-minute ramp that the DME set for me. Or maybe the leaks contributed to this issue.

In spite of being highly motivated, I was more anxious than I thought I’d be about connecting to “the hose.” So I wonder if my anxiety contributed to this experience. I stayed on the machine for a little over two hours and decided to quit. I wasn’t really sleeping—just dozing a bit—and I didn’t want to go on like that all night.

I felt worse than usual the next day. So I decided to take the second night off. I thought that maybe I would do better to just ease into this. I wondered if I was putting a lot of pressure on myself to perform, and if this was making it even harder for my mind/body to adjust to this new contraption.

Last night, I tried it again and I think I made a bit of progress. I took an Epsom salt bath before bed to help me relax. In spite of the bath, I had a headache when I went to bed so I took a Tylenol. I got the headgear and mask adjusted better and it felt like a better fit. No leaks, except at the joint where the hose connects (and maybe those are normal). And after my ramp ended, I slowly slid into a more normal breathing pattern. I’m now thinking I may not need the ramp. I even slept a little, but kept waking up. The headgear/mask felt a little too tight and the headache seemed to come and go. After about 2 ½ hours, I decided to quit for the night. But I feel more encouraged now. Maybe I’ll make it 3-4 hours tonight.

Maybe some of you can offer some practical advice or words of wisdom? I’ve already read a bunch of stuff on the forum, but feel I’m still at the bottom of a pretty steep learning curve. Thanks!

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: started CPAP 6-16-09; have used ASV, oral appliance, and Winx; currently on APAP

ozij
Posts: 10514
Joined: Fri Mar 18, 2005 11:52 pm

Re: one more newbie

Post by ozij » Fri Jun 19, 2009 11:28 am

)Welcome, oh wearysoul!
First of all - remember you are not alone. Many of us - I'm one of them - went through similar experience in the beginning.

Second - masks don't have to be tight, certainly not to the point of hurting. They have to be snug - and gauging the difference is an art you acquire with time.

Third: The way you route your hose effects the masks stability. It boils down to what feels good for you - for me the hose has to dangle from above my head. A girl's hair elastic around the hose, a (clean, new ) shoelace tying that to a hook somewhere above my head made a big difference for me.

Fourth: The pillow you lay you head on also makes a difference.

Be patient - it takes time to learn this new way of sleeping --

Good luck!
O

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
jdm2857
Posts: 2982
Joined: Wed Aug 20, 2008 8:29 pm
Location: South Jersey

Re: one more newbie

Post by jdm2857 » Fri Jun 19, 2009 12:38 pm

Every mask is designed to lead a certain amount of air to allow exhaled CO2 to escape. Look at the mask where you felt the leak. You'll probably find small holes there.

I have not experience this, but many forum users complain that if a ramp is set to too low of a pressure they feel that they cannot get enough air. So if your ramp is set to start at 4, you may want to bump it up to 5 or even 6. Many users find that they can turn the ramp off altogether after a few days, but if it makes you more comfortable it really does no harm.

If you don't know how to adjust your machine, post back, and someone will give you the secret incantation.
jeff

computerguy2202
Posts: 13
Joined: Mon May 11, 2009 2:00 pm
Location: Hamilton, Ontario, Canada

Re: one more newbie

Post by computerguy2202 » Fri Jun 19, 2009 12:45 pm

Hi and welcome;

It does take time to get use to the equipment...after all how many years have you been sleeping without anything on your face? Everyday gets better and you will get use to where the mask goes, the best place for the hose, ramping, etc.. I have my ramp starting at 7 and goes to 17 in 5 minutes. Within the 5 minutes I can adjust the mask, if needed, and get settled in before the machine blows my face off Don't give up, give it time, and use it every time you sleep.

_________________
Mask
Additional Comments: Pressure 17cm

User avatar
twokatmew
Posts: 606
Joined: Wed Jun 03, 2009 6:06 pm
Location: Mid-Michigan, US

Re: one more newbie

Post by twokatmew » Fri Jun 19, 2009 12:57 pm

I started at 7cm two weeks ago, and I tried the ramp at first. I ended up turning it off, as I felt like I wasn't getting enough air. I'm still waking with headaches and a sore right eye, but I manage to keep the mask on all night. (I do spend a bit of time lying there waiting for sleep to whisk me away.) I'm now at 9cm and still have the ramp off. I think it might be helpful with higher pressures, but I found at such a low pressure, the ramp is more hassle than help.

Welcome!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: CPAP 6/5/2009, Rx: 11-16, OSCAR

User avatar
LinkC
Posts: 3154
Joined: Wed Oct 29, 2008 1:06 pm
Location: Amelia Island, FL

Re: one more newbie

Post by LinkC » Fri Jun 19, 2009 1:07 pm

Welcome, brother hosehead!

Know that mask leakage is rarely (if ever!) solved in one night. Some of us have been thru SEVERAL masks and almost infinite adjustments before conquering them. But it will come.

If the out of sync breathing issue remains, it's probably the result of a "breath relief" feature (called C-flex on Respironics machines) which lowers the pressure during exhale. The problem is, it senses the slowdown in your exhalation and starts restoring pressure before you finish exhaling. The idea is to provide full pressure for the entire inhalation. Some people have trouble finishing their exhale because the pressure is already being reapplied. If that sounds like your situation, you can turn the feature down...or off.

Your leak "at the joint" may actually be that...OR it may be the normal operation of the mask exhaust port, which is often in that area. (Dunno what mask you have...) It allows the incoming air to flush out the CO2 so you don't breathe it right back in. But there ARE certain masks that are prone to leaks at the swivel fitting. (The fix is fairly easy, involving teflon plumbing tape, but we won't get into that unless it's an actual leak there.)

As ozij suggested, it's common for "rookies" to adjust their masks too tight at first. Most masks are designed so that the "skirt" inflates and internal air pressure makes the seal. If you can't feel that happening when you turn the air on, then you may be too tight. Many times, LOOSENING the straps will stop a leak (I know that's counter-intuitive, but it happens.) Experiment...

Good luck with your new "adventure"!

_________________
MachineMaskHumidifier
Additional Comments: 11-14 cmH2O
The OSA patient died quietly in his sleep.
Unlike his passengers who died screaming as the car went over the cliff...

User avatar
twasbrillig
Posts: 76
Joined: Fri May 30, 2008 1:39 pm

Re: one more newbie

Post by twasbrillig » Fri Jun 19, 2009 1:29 pm

Weary soul, good luck! I just wanted to put in a word for good sleep hygiene as you are getting mask and other issues adjusted. Even though you may not have needed to do this before, the extra stress of sleeping with a mask on might require this approach:

1) cut back on or eliminate caffeine
2) exercise briskly for a little while almost every day
3) avoid a nap unless you are so tired you can't work, drive, function, etc
4) do something relaxing before bedtime (I recommend some do-it-yourself yoga - you can learn some basic stuff from a book)
5) a warm bath is a good idea (Your body temperature has to drop for you to sleep. If you raise your temperature with a hot bath or shower, it makes it easier for your temperature to drop as you fall asleep...)

Hope this helps. You are right about the learning curve, and I think you have come to the right place to learn. CPAPtalk is a place where people are focused on educating and helping each other. I have been to other forums, and this one is THE BEST!
Why does everyone else post cutesy avatars, and I'm the only one who posts a picture of myself?

User avatar
timbalionguy
Posts: 888
Joined: Mon Apr 27, 2009 8:31 pm
Location: Reno, NV

Re: one more newbie

Post by timbalionguy » Fri Jun 19, 2009 2:59 pm

IMHO, your pressure might be too low to be comfortable long-term. Many people here have had problems with pressures below about 7. A little extra pressure won't hurt most people, so ask your doctor about that.

My doctor recommended that the machine should come on at full minimum pressure, even though an autoadjust range was specified. This was good wisdom. My lower pressure was 10, and it did not take me long at all to get used to 10. It also helped me diagnose leak problems early on before moving on to higher pressures.
Lions can and do snore....

RipVW
Posts: 1684
Joined: Tue Oct 09, 2007 7:29 pm
Location: USA
Contact:

Re: one more newbie

Post by RipVW » Fri Jun 19, 2009 5:32 pm

Image
Hello & Welcome, wearysoul! Glad you found this forum—SO much useful info and everyone here is so willing to help!

You're lucky to have found this forum as you begin CPAP--took me almost year before I found this place and, with the help of folks here, got my CPAP therapy working as it should. It does take time to get everything tweaked and working at its best, so patience is key as one begins CPAP.

Image
"You are getting sleepy . . ."
Check out my chinstrap--> http://cpapchinstraps.com
Image

wearysoul
Posts: 53
Joined: Wed Jun 17, 2009 1:37 pm
Location: North Carolina

Re: one more newbie

Post by wearysoul » Sat Jun 20, 2009 8:16 am

ozij wrote:)Welcome, oh wearysoul!
First of all - remember you are not alone. Many of us - I'm one of them - went through similar experience in the beginning.
Second - masks don't have to be tight, certainly not to the point of hurting. They have to be snug - and gauging the difference is an art you acquire with time.

Third: The way you route your hose effects the masks stability. It boils down to what feels good for you - for me the hose has to dangle from above my head. A girl's hair elastic around the hose, a (clean, new ) shoelace tying that to a hook somewhere above my head made a big difference for me.

Fourth: The pillow you lay you head on also makes a difference.
Be patient - it takes time to learn this new way of sleeping --

Good luck!
O
Thanks for the tips, O. I fiddled around with the hose last night, and I can see how it would work better dangling from above my head. That's definitely on my "to do" list.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: started CPAP 6-16-09; have used ASV, oral appliance, and Winx; currently on APAP

wearysoul
Posts: 53
Joined: Wed Jun 17, 2009 1:37 pm
Location: North Carolina

Re: one more newbie

Post by wearysoul » Sat Jun 20, 2009 8:22 am

Thank you for your reply. I turned the ramp off last night and it didn't seem to have hurt anything. It may have even helped. I'll continue without it and see how it goes. wearysoul

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: started CPAP 6-16-09; have used ASV, oral appliance, and Winx; currently on APAP

wearysoul
Posts: 53
Joined: Wed Jun 17, 2009 1:37 pm
Location: North Carolina

Re: one more newbie

Post by wearysoul » Sat Jun 20, 2009 8:29 am

Thanks, computerguy, for your reply. And thank you to all the others who also replied. I'm blown away by the warm welcome I've received. And the comments have been very useful. I'll keep you posted on how it goes. wearysoul

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: started CPAP 6-16-09; have used ASV, oral appliance, and Winx; currently on APAP

User avatar
rested gal
Posts: 12880
Joined: Thu Sep 09, 2004 10:14 pm
Location: Tennessee

Re: one more newbie

Post by rested gal » Sat Jun 20, 2009 10:06 am

wearysoul wrote:Thanks for the tips, O. I fiddled around with the hose last night, and I can see how it would work better dangling from above my head. That's definitely on my "to do" list.
Welcome to the message board, wearysoul. Cool nickname, btw.

Here's a link where you can get lots of ideas from simple to elaborate, for hanging the air hose out of the way:

LINKS to Hose hangers and methods of managing the air hose
viewtopic.php?t=10640
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435

User avatar
DoriC
Posts: 5214
Joined: Sat Sep 13, 2008 9:28 pm
Location: NJ

Re: one more newbie

Post by DoriC » Sat Jun 20, 2009 10:36 am

Welcome. Glad you found your way here. As the spouse and "in-house cpap specialist" , I was able to observe that my hubby's breathing pattern did not match the Cflex mode(your machine has EPR). and the pressure was increasing before he had completed his exhale. At first we lowered it to the minimum (1) because it did help a bit adjusting to the mask, pressure,etc, but we very soon turned it completely off and he and his machine are now in sync. Pressure=12cms. Good luck.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: 14/8.4,PS=4, UMFF, 02@2L,
"Do or Do Not-There Is No Try"-"Yoda"
"We are what we repeatedly do,so excellence
is not an act but a habit"-"Aristotle"
DEAR HUBBY BEGAN CPAP 9/2/08