Can someone please explain pressure?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
ractar28
Posts: 222
Joined: Wed Oct 01, 2008 3:06 pm

Can someone please explain pressure?

Post by ractar28 » Sat Oct 18, 2008 4:32 pm

I'm looking at "Patient Circuit Accessory Pressure / Flow Characteristics in section 4.3.2 of the Remstar LX CPAP Systems Service Manual.

It lists different items (masks, tubing, valves, etc) with a "Pressure Drop @ 30 lpm (cm H2O)"

The 6ft Reusuable Flexible Tubing has 0.11 listed.

Is that saying that at 30cm or pressure (which the machine can only do 20 anyway, so I'm confused) that the pressure will drop .11 cm or .11 (11%)?

If it means neither, what does it mean?

User avatar
Craig-Tx
Posts: 232
Joined: Fri Dec 22, 2006 6:56 am
Location: Fort Worth, TX

Re: Can someone please explain pressure?

Post by Craig-Tx » Sat Oct 18, 2008 5:13 pm

ractar28 wrote: It lists different items (masks, tubing, valves, etc) with a "Pressure Drop @ 30 lpm (cm H2O)"

The 6ft Reusuable Flexible Tubing has 0.11 listed.
LPM is a flow rate (Liter per minute)
cm H20 is a pressure

My guess is that they are saying the tube has a .11 cm H20 pressure drop when air is flowing through it at 30 LPM. Which .11 cm H20 is basically negligible.

-craig

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Started PAP on 1/16/07. - Typical 90% pressure ~ 10 cm H20

fuzzy96
Posts: 448
Joined: Sun Sep 02, 2007 5:34 pm
Contact:

Re: Can someone please explain pressure?

Post by fuzzy96 » Sat Oct 18, 2008 5:27 pm

you are being confused . pressure by definitionis a force being applied .for there to be pressure there must be a resistance to FLOW. in your example the flow rate is 30 litres /minute. yes your machine is capable of only 20 cm-h20. at the 20 cm-h20, a pressure drop of .11 would result in a pressure reading at the end of the hose of 19.89cm-h20. pnuematics or hydraulics either one,pressure pushing creates movement of the media to overcome the resistance to flow. the pressure pushing against the tissues in your airways moves those tissues allowwing room for the air to flow- thus the theory of cpap. lung volume is meaured in litres. pressure is measured (for cpap purposes) in centimeters of water. we are all used to psi for our tires.10cm h20 basically is enough pressure to raise a column of water 10 cm higher than if no pressure was applied. and to really confuse you there is a correction depending on how high above sea level the measurement is take

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Can someone please explain pressure?

Post by Goofproof » Sat Oct 18, 2008 5:44 pm

There are three states of matter, Matter, Anti-Matter, and Don't Matter. Check, when your book was printed, it's for a very old system, if it says the manuel was hand printed by Ben Franklin, keep it it's worth money. With a modern machine nothing in that book will apply. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

CRS
Posts: 59
Joined: Wed Oct 08, 2008 12:35 am
Location: The Woodlands, TX

Re: Can someone please explain pressure?

Post by CRS » Sat Oct 18, 2008 9:03 pm

Goofproof wrote:There are three states of matter, Matter, Anti-Matter, and Don't Matter. Check, when your book was printed, it's for a very old system, if it says the manuel was hand printed by Ben Franklin, keep it it's worth money. With a modern machine nothing in that book will apply. Jim
I work for a scientific company; will have to remember that for meetings.
M Series Auto w/A-flex, M Series HH, Mirage Liberty hybrid mask, Respironics hose snuggie.

ractar28
Posts: 222
Joined: Wed Oct 01, 2008 3:06 pm

Re: Can someone please explain pressure?

Post by ractar28 » Sat Oct 18, 2008 9:12 pm

Sure it matters. If my titration was at 12cm, then was that 12cm at the machine or at the end of a "standard" circuit? I checked my old machine and it was at 12, so obviously, it was the measure of the complete circuit. But I started with no humidifier. I've moved to a heated humidifier and a bacteria filter (and may change masks). Since different masks have different leak rates, if I were on a CPAP, I'd want to do some math and make sure the "final pressure" is correct. Even on an APAP, I still want to keep the starting pressure "somewhat in line".

My Ultra Mirage 2 has a flow of 38 at a pressure of 12. If I move to an Activa, the flow is about 34 at 12. Now all I have to figure out is what the bleeping leak rate is (and if that's a percentage or what).

Yep, I try hard to make this harder than it is.

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Can someone please explain pressure?

Post by Goofproof » Sat Oct 18, 2008 10:01 pm

All machines have a pressure spec of .5cm +/_ your set pressure, nothing normally used will make much difference in your treatment pressure, Except for the "Bacterial filter", which is extra and not needed, as long as your machine isn't contaminated. If it was, I wouldn't be using it in the first place. The air the machine breaths, is the same air you breath all day, do you wear a filter all day.

What is really needed is a machine that collects data and the software to see how your treatment is doing. One of those would have auto alt comp, and leak compensation to some degree. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

User avatar
Snoredog
Posts: 6397
Joined: Sun Mar 19, 2006 5:09 pm

Re: Can someone please explain pressure?

Post by Snoredog » Sat Oct 18, 2008 10:39 pm

Can someone please explain pressure?
Pressure is like a fart....


It seeks the path of least resistance
someday science will catch up to what I'm saying...

CRS
Posts: 59
Joined: Wed Oct 08, 2008 12:35 am
Location: The Woodlands, TX

Re: Can someone please explain pressure?

Post by CRS » Sun Oct 19, 2008 2:43 am

ractar28 wrote:If it means neither, what does it mean?
It sounds like the service manual you are looking at is indeed giving the exact loss in pressure (in cm H20) for each accessory added to the patient circuit, but I don't think it's meaningful information for us. The xPAP is set to provide 12 cm H20 of pressure, meaning that if you take your mask off and the machine doesn't have anything to blow AGAINST, it will blow like crazy, trying to find some resistance to flow so it can do its job of creating 12 cm of pressure again. Put it back on, it has something to blow against, the blower slows down. So what if someone at Respironics hooked up some sensors to either end of a 6ft hose, set up a table fan blowing air through it at a rate of 30 lpm, and discovered that the hose has .11 cm H20 less pressure coming out the far end than going in? Does the table fan detect that loss of pressure and increase its speed to make up for it? No, but your xPAP does.

Same thing for the vent flow rates for masks. Vent flow rate (aka leak rate) is only useful when you have something to compare it to, for example, a leak rate reported by your xPAP. Say your machine displays last night's leak rate as 45 lpm. The user guide to your mask says that its vent flow rate, at the pressure your machine is set to, is 38 lpm. Then you know that you have a leak on the order of (45-38) or 7 lpm. That's useful information.
M Series Auto w/A-flex, M Series HH, Mirage Liberty hybrid mask, Respironics hose snuggie.

User avatar
dsm
Posts: 6996
Joined: Mon Jun 20, 2005 6:53 am
Location: Near the coast.

Re: Can someone please explain pressure?

Post by dsm » Sun Oct 19, 2008 2:58 am

Things aren't always what they seem.

I did a series of tests about 18-24 months ago that showed across a variety of machines, that on a 6ft hose whith air flowing and pressure on the LCD or at the machine end saying say 12 CMs that only 11 CMs will show on a dial manometer at the mask UNLESS, the machine is one that uses an auxiliary pressure sensing line (like the PB330 or PB42x models in which case the pressures always matched the machine end. A 10ft hose had about 1.3 CMs of loss.

Cheers

DSM

#2 viewtopic.php?f=1&t=29035&hilit=pressur ... +6%27+tube

I can't seem to locate the original thread

D
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)

CRS
Posts: 59
Joined: Wed Oct 08, 2008 12:35 am
Location: The Woodlands, TX

Re: Can someone please explain pressure?

Post by CRS » Sun Oct 19, 2008 4:39 am

Here we go again. So much for keeping it non-technical.
M Series Auto w/A-flex, M Series HH, Mirage Liberty hybrid mask, Respironics hose snuggie.

blowfish
Posts: 366
Joined: Wed Apr 04, 2007 8:53 am

.

Post by blowfish » Wed Oct 22, 2008 7:31 pm

.
Last edited by blowfish on Tue Nov 04, 2008 3:52 pm, edited 1 time in total.

blowfish
Posts: 366
Joined: Wed Apr 04, 2007 8:53 am

.

Post by blowfish » Wed Oct 22, 2008 7:34 pm

.
Last edited by blowfish on Tue Nov 04, 2008 3:52 pm, edited 1 time in total.

User avatar
roster
Posts: 8153
Joined: Wed Jan 18, 2006 8:02 pm
Location: Chapel Hill, NC

Re: Can someone please explain pressure?

Post by roster » Thu Oct 23, 2008 8:16 am

mth712 wrote:........ You know what a fart is? Clearing the way for the main event!
That's clever. Could you also say, "Clearing the room so the main event can be enjoyed in private"?
Rooster
I have a vision that we will figure out an easy way to ensure that children develop wide, deep, healthy and attractive jaws and then obstructive sleep apnea becomes an obscure bit of history.https://www.youtube.com/watch?v=0ycw4uaX ... re=related

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Can someone please explain pressure?

Post by Goofproof » Thu Oct 23, 2008 11:17 am

rooster wrote:
mth712 wrote:........ You know what a fart is? Clearing the way for the main event!
That's clever. Could you also say, "Clearing the room so the main event can be enjoyed in private"?
Nature's way of telling you The Party's Over. Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire