RESMED S7 LIGHTWEIGHT
-
baxchels
RESMED S7 LIGHTWEIGHT
Does anyone know how to adjust the prescribed pressure on the Resmed S7 lightweight?
Re: RESMED S7 LIGHTWEIGHT
For the S-7
Turn the power on.
Hold the lower arrow and left arrow down together for 5 seconds
Clinicians menue will appear.
Follow the prompts
Brenda
Turn the power on.
Hold the lower arrow and left arrow down together for 5 seconds
Clinicians menue will appear.
Follow the prompts
Brenda
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Love my papillow, Aussie heated hose and PAD-A-CHEEKS! Also use Optilife, UMFF(with PADACHEEK gasket), and Headrest masks Pressure; 10.5 |
-
baxchels
Re: RESMED S7 LIGHTWEIGHT
I tried that and it did not work. I tried most of the button in combination but no luck?? Please help????
Re: RESMED S7 LIGHTWEIGHT
These instructions are supposed to work for the S7 and S8 series machines.
Den
While pressing the Down arrow button, press and hold down the right hand button. Hold both buttons down together for several seconds.
That brings up the "Welcome" clinical menu.
Use the up/down button to navigate to the settings menu.
Use the left button to "Change" settings. A "Change" will not be saved until you hit the left key again under the word "Apply".
It's safe to walk through the changes just to see what they would be. When looking at a "Change" setting, the up/down arrow buttons will scroll through the possible changes of that setting. As long as you don't press the left button under the word "Apply", no change will actually happen. If you want to back out of the "Change" without changing the setting, press the righthand button under the word "Exit."
Use the right button under the word "Exit" to go backwards from whatever menu section you're in. Hit "Exit" several times to exit your way backwards completely out of the clinical menu.
Den
While pressing the Down arrow button, press and hold down the right hand button. Hold both buttons down together for several seconds.
That brings up the "Welcome" clinical menu.
Use the up/down button to navigate to the settings menu.
Use the left button to "Change" settings. A "Change" will not be saved until you hit the left key again under the word "Apply".
It's safe to walk through the changes just to see what they would be. When looking at a "Change" setting, the up/down arrow buttons will scroll through the possible changes of that setting. As long as you don't press the left button under the word "Apply", no change will actually happen. If you want to back out of the "Change" without changing the setting, press the righthand button under the word "Exit."
Use the right button under the word "Exit" to go backwards from whatever menu section you're in. Hit "Exit" several times to exit your way backwards completely out of the clinical menu.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
-
tundra
Re: RESMED S7 LIGHTWEIGHT
Try holding the right arrow and the down arrow at the same time for at least 5 seconds. Worked for mine.
Re: RESMED S7 LIGHTWEIGHT
Brenda,bdp522 wrote:For the S-7
Turn the power on.
Hold the lower arrow and left arrow down together for 5 seconds
Clinicians menue will appear.
Follow the prompts
Brenda
I think that is the process for accessing nightly data / user menu on the Resmed machines.
As explained in the other posts to access clin menu it is the right arrow + down key held down together for 3+ secs.
Cheers
DSM
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)
Re: RESMED S7 LIGHTWEIGHT
Thanks DSM!
Brenda
Brenda
_________________
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Love my papillow, Aussie heated hose and PAD-A-CHEEKS! Also use Optilife, UMFF(with PADACHEEK gasket), and Headrest masks Pressure; 10.5 |
-
AmandaV
Re: RESMED S7 LIGHTWEIGHT
This saved me a 2 hr drive and I'm sure a chunk of change. So they could change the cpap level which took 2minutes to change myself...once I found out how to get into the menu. I looked everywhere in the book but I guess it's a big secret. AHHH a goods night rest again! Thank you!
-
Dennis tne menace
Re: RESMED S7 LIGHTWEIGHT
will the max pressure setting go higher then 7 once i am in the clinic menu i can change the ramp up pressure setting but on the gain setting app it will not change either up or down , it will also only move in the 4-7 range on the ramp changes also
I need a higher pressure then 7 , hope someone can help me
I need a higher pressure then 7 , hope someone can help me
Re: RESMED S7 LIGHTWEIGHT
Of course. It'll go all the way up to 20 cm. (if a person needed that much)Dennis tne menace wrote:will the max pressure setting go higher then 7 once i am in the clinic menu i can change the ramp up pressure setting but on the gain setting app it will not change either up or down , it will also only move in the 4-7 range on the ramp changes also
I need a higher pressure then 7 , hope someone can help me
Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
-
chetcolumbo
Re: RESMED S7 LIGHTWEIGHT
Hold the down arrow and the red button down till the screen changes to clinical. After the machine is on, but not running.
-
Guest
Re: RESMED S7 LIGHTWEIGHT
bdp522 wrote:For the S-7
Turn the power on.
Hold the lower arrow and left arrow down together for 5 seconds
Clinicians menue will appear.
Follow the prompts
Brenda
-
Jeskatron
Re: RESMED S7 LIGHTWEIGHT
Can the setting be lower than 4 or does it not go lower than that.
- billbolton
- Posts: 2264
- Joined: Wed Jun 07, 2006 7:46 pm
- Location: Sydney, Australia
Re: RESMED S7 LIGHTWEIGHT
4cm H20 is the lowest setting for any CPAP machine.
Cheers,
Bill
Cheers,
Bill