Hi
I currently have the ResMed AirSense 10 AutoSet machine which I've had for 3 years. I also have spare hose, filters … a new tub and also the Medistrom Pilot 24 Lite battery for this machine. This CPAP device has become noisy AGAIN – a whistling noise on inhaling. It is cleaned regularly. Last time it became noisy I did lots of research and chose to replace the tub with a new one which is what is on it at the moment and this helped until very recently. This was less than 12 months ago. I’m tossing up whether to buy the same machine again (I normally get on pretty well with it) especially because of the spares I have or whether it would be wiser to buy the AirSense 11 which I believe is not compatible with spare parts mentioned above.
I’d like some advice please. What do people think is my best option. I am on an aged pension and with the cost of CPAP here is damned expensive. My private health cover (which is also damned expensive) will cover some of my costs. Any advice would be greatly appreciated.
Thanks and regards
Pam
Machine upgrade? advice please
-
mlmollenkamp
- Posts: 111
- Joined: Fri Apr 01, 2016 3:46 pm
Re: Machine upgrade? advice please
First, why not replace the water tub again? Its not uncommon for the tub seal to begin to leak which causes a whistle. Note that under insurance these are usually replaced more often than once a year. The flexible seal that is squeezed between the bottom and top can leak. Also look to make sure that the tub is fully inserted into the machine, mine often whistles if its not completely inserted. I suggest you try pushing the tub in while breathing, if the whistle disappears you know where the issue is. You can also block the end of the hose where the mask attaches, run the machine, then carefully listen close to the machine to find where the whistle is coming from. An old mechanics trick is to use a piece of tubing in your ear then use the other end like a stethoscope to find the source of the noise.
But also. If you look inside where the tub is inserted there are two flexible pieces into which the tub tubes insert. The one on the output side is removable and replaceable. You should carefully inspect and clean both of these. If a replacement water tub doesn't fix the issue I'd replace the output adapter, they aren't terribly expensive and are pretty easy to replace (there are youtube video's showing how to do it).
Its unlikely that a 3 year old machine has failed, they normally are good for 20,000 hours or more, so 6 to 8 years.
I hope this helps.
But also. If you look inside where the tub is inserted there are two flexible pieces into which the tub tubes insert. The one on the output side is removable and replaceable. You should carefully inspect and clean both of these. If a replacement water tub doesn't fix the issue I'd replace the output adapter, they aren't terribly expensive and are pretty easy to replace (there are youtube video's showing how to do it).
Its unlikely that a 3 year old machine has failed, they normally are good for 20,000 hours or more, so 6 to 8 years.
I hope this helps.
_________________
| Machine: Aircurve 11 asv |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: P30i is modified with Dreamwear headgear. |
Re: Machine upgrade? advice please
Thank you so much. I will certainly take your advice step by step. Have a great day.mlmollenkamp wrote: ↑Sun Sep 06, 2026 8:59 pmFirst, why not replace the water tub again? Its not uncommon for the tub seal to begin to leak which causes a whistle. Note that under insurance these are usually replaced more often than once a year. The flexible seal that is squeezed between the bottom and top can leak. Also look to make sure that the tub is fully inserted into the machine, mine often whistles if its not completely inserted. I suggest you try pushing the tub in while breathing, if the whistle disappears you know where the issue is. You can also block the end of the hose where the mask attaches, run the machine, then carefully listen close to the machine to find where the whistle is coming from. An old mechanics trick is to use a piece of tubing in your ear then use the other end like a stethoscope to find the source of the noise.
But also. If you look inside where the tub is inserted there are two flexible pieces into which the tub tubes insert. The one on the output side is removable and replaceable. You should carefully inspect and clean both of these. If a replacement water tub doesn't fix the issue I'd replace the output adapter, they aren't terribly expensive and are pretty easy to replace (there are youtube video's showing how to do it).
Its unlikely that a 3 year old machine has failed, they normally are good for 20,000 hours or more, so 6 to 8 years.
I hope this helps.
kind regards
Pam
-
mlmollenkamp
- Posts: 111
- Joined: Fri Apr 01, 2016 3:46 pm
Re: Machine upgrade? advice please
A final thought...
I'm very sensitive to noise when I sleep. Just like you can see many more stars at night after an hour outside in the dark, your ears increase their sensitivity tremendously in a quiet room. I absolutely can't sleep with my machine without another source of noise. When I travel I always carry a noise machine. It need not be loud, just a quiet rush. I use one at home during the winter when I'm not running the window AC. If the whistle is subtle you might consider one. I really like the Lectrofan machines, I have both the travel and bedside machines,I like both. I also find that the air intake side of the machine emits "breathing" noise that's synchronized with my breathing. I keep the noise machine right next to the air inlet for that reason. And I prefer to sleep on the side of the bed that orients the intake away from me.
Please let me know if you find a solution, or if indeed it turns out to be the machine, it will be useful to others in the future.
I'm very sensitive to noise when I sleep. Just like you can see many more stars at night after an hour outside in the dark, your ears increase their sensitivity tremendously in a quiet room. I absolutely can't sleep with my machine without another source of noise. When I travel I always carry a noise machine. It need not be loud, just a quiet rush. I use one at home during the winter when I'm not running the window AC. If the whistle is subtle you might consider one. I really like the Lectrofan machines, I have both the travel and bedside machines,I like both. I also find that the air intake side of the machine emits "breathing" noise that's synchronized with my breathing. I keep the noise machine right next to the air inlet for that reason. And I prefer to sleep on the side of the bed that orients the intake away from me.
Please let me know if you find a solution, or if indeed it turns out to be the machine, it will be useful to others in the future.
_________________
| Machine: Aircurve 11 asv |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: P30i is modified with Dreamwear headgear. |
Re: Machine upgrade? advice please
Thanks again for your wonderful advice. Since the last episode of 'whistling' we have kept the radio on all night quite low. I am now very careful with 'seating' the tub each day and it does seem less noisy. I will certainly look into the noise machine though. And yes, I agree, I am also very sensitive to noise when I sleep .... mind you, I was never kept awake by my snoring but hubby was lol lolmlmollenkamp wrote: ↑Wed Sep 09, 2026 8:34 amA final thought...
I'm very sensitive to noise when I sleep. Just like you can see many more stars at night after an hour outside in the dark, your ears increase their sensitivity tremendously in a quiet room. I absolutely can't sleep with my machine without another source of noise. When I travel I always carry a noise machine. It need not be loud, just a quiet rush. I use one at home during the winter when I'm not running the window AC. If the whistle is subtle you might consider one. I really like the Lectrofan machines, I have both the travel and bedside machines,I like both. I also find that the air intake side of the machine emits "breathing" noise that's synchronized with my breathing. I keep the noise machine right next to the air inlet for that reason. And I prefer to sleep on the side of the bed that orients the intake away from me.
Please let me know if you find a solution, or if indeed it turns out to be the machine, it will be useful to others in the future.