Therapy started in February and I have been battling Aerophagia nearly the whole time. Doctor put me on an acid inhibitor to counteract the heartburn/reflux that it was causing (though it still sometimes exists). Burping all day and stomach pressures is constant. For a week now, however, I have felt very fatigued and I can't figure out why.
I have been retrying a wedge pillow for a few nights now, no improvement.
I generally fall asleep on my left side, but more than half the time, wake up on my back.
https://sleephq.com/public/teams/share_ ... edf4f17f23
Constant Aerophagia
Re: Constant Aerophagia
You may need to reduce your pressure settings in small increments to see if you can get to a point you don't have the aerophagia bothering you so much...like reducing the max setting by .2 cm increments. This option will take some time.
How did you or your medical team come up with the settings you are currently using?
They sure aren't very high but you aren't the first person to have significant aerophagia with rather low pressures.
Another option is to back way up with the pressures and work upwards monitoring data to evaluate the results as far as therapy effectiveness and aerophagia relief.
Like start at 4 minimum and 6 maximum and EPR at 2..or 4 min and 7 max with EPR at 3.
This might be a faster option.
How did you or your medical team come up with the settings you are currently using?
They sure aren't very high but you aren't the first person to have significant aerophagia with rather low pressures.
Another option is to back way up with the pressures and work upwards monitoring data to evaluate the results as far as therapy effectiveness and aerophagia relief.
Like start at 4 minimum and 6 maximum and EPR at 2..or 4 min and 7 max with EPR at 3.
This might be a faster option.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Constant Aerophagia
I was initially started on the standard 4-20 (through Lofta). After my first follow up (complaining about Aerophagia), the tech lowered me to 4-7.6.
based on advice from the board here and other sources, I have been raising my minimum to avoid larger spikes during reactionary events.
it's kind of a weird toss up between getting the lowest pressure versus stability and good sleep.
based on advice from the board here and other sources, I have been raising my minimum to avoid larger spikes during reactionary events.
it's kind of a weird toss up between getting the lowest pressure versus stability and good sleep.
Re: Constant Aerophagia
did you try pugsys suggestion in your other post? she said set epr at 3. did you try that yest?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg