Terrified I will be forced to stop BiPAP therapy

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
eoram
Posts: 14
Joined: Sat Apr 18, 2026 6:08 pm

Terrified I will be forced to stop BiPAP therapy

Post by eoram » Sun Jul 12, 2026 4:35 pm

I am on Day 24 of BiPAP and I am getting awful stomach pains that wake me up from aerophagia. My pressures are low: EPAP 4 IPAP 7. I use a chin strap and cervical collar and HOB elevation. I get aerophagia with my Dreamwear nasal and also my F40 - I don't breathe through my mouth with the F40, at least I don't think I do because my mouth is never dry. I usually use the nasal mask. My AHI is usually 1 or <1.

I felt like I was going to die when I was suffering from sleep apnea. I can't go back to that, I just can't. I have to find a way to make BiPAP work - what else can I do, and please please offer me some hope that this just needs time to get better.

Many thanks.

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Terrified I will be forced to stop BiPAP therapy

Post by Pugsy » Sun Jul 12, 2026 8:47 pm

Which brand and model bipap are you currently using now?
Will it allow you to perhaps reduce the 4 cm EPAP pressure to 3?
The ResMed AirCurve 10 or the latest model 11 I think allows for 3 EPAP when in S or fixed mode. I don't know about other brands.
Then you could try 3/6 or 3/7.....the change from 3 to 4 pressure support (the difference between inhale and exhale) and that little change might be enough to not be getting that air into your stomach.
Might work and of course might not but I would think it would be worth a try if you are able.

What does your doctor say?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

eoram
Posts: 14
Joined: Sat Apr 18, 2026 6:08 pm

Re: Terrified I will be forced to stop BiPAP therapy

Post by eoram » Thu Jul 16, 2026 6:26 am

Thank you so much for responding. I use an AirCurve 11 in S Mode. I did wonder if my PS was too low but didn't know a number below 4 was an option. I have made some progress on the aerophagia, it will wake me up with the burping but for 3 days now I haven't had the sharp abdominal pains. I have been on BiPAP for a month and still feel exhausted all day so was thinking of posting my OSCAR and Sleep HQ to the Forum for advice. As lousy as I feel I want to keep going because apnea is worse. (my doctor says dental device and referral for Modafanil, he's washing his hands of me)