AirSense 10 not connecting

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
fred0506
Posts: 1
Joined: Fri Feb 06, 2026 5:33 pm

AirSense 10 not connecting

Post by fred0506 » Fri Feb 06, 2026 5:48 pm

Hello,

I have an AirSense 10 and according to the myAir portal stopped sending data on Nov 19 2025. When I look at the machine I see the wireless icon with a green circle crossed out. I have tried restarting the machine and checked to make sure airplane mode is turned off. Any ideas on how to resolve the issue?

Note: I do realize I can get the data off the SD card still but since it isn't uploaded my doctor requires you to go in person instead of virtual appointments.

Thanks!
10000100023.jpg
10000100023.jpg (642.97 KiB) Viewed 5722 times
1000010003.jpg
1000010003.jpg (861.8 KiB) Viewed 5722 times
1000010004.jpg
1000010004.jpg (373.91 KiB) Viewed 5722 times

User avatar
tlohse
Posts: 267
Joined: Fri Jun 27, 2014 5:32 pm
Location: West Bend, Wisconsin
Contact:

Re: AirSense 10 not connecting

Post by tlohse » Fri Feb 06, 2026 10:22 pm

I had that problem once. What fixed it is unplugging it waiting for a few minutes and Plugging it in again. It’s been working since then. Also I think the airsense 10 is discontinued now there is an 11. So if you had your unit at least 5 years I think you might be able to get a new one. I know that’s how it is with some providers and insurances.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack
Additional Comments: Machine is Autoset.
Thomas Lohse
Machine ResMed AirSense 11 Auto Set
Mask: ResMed AirFit N30i Standard Frame with Medium cushion
CPAP USER SINCE JUNE 2013

Grumpy48
Posts: 462
Joined: Tue May 12, 2020 7:57 am

Re: AirSense 10 not connecting

Post by Grumpy48 » Sat Feb 07, 2026 7:11 am

It may not help where you presently use the AS10, but try placing the unit by a window and see if it can show a received signal there. On my cell phone the signal is weak inside the house, but if I try it near a window I can generally get a few more bars to show. One of the numbers on your AS10 indicates it uses a 4G cellular system so that is current enough that it should work if signal is getting to it.

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: AirSense 10 not connecting

Post by Dog Slobber » Sat Feb 07, 2026 7:32 am

It's either not broadcasting or there is no cell tower close enough to connect to.

Take it on a road trip, (a city centre, a friends house, etc.) where you know there should be strong cell signals. Plug it in.

If you see bars, in your non-home location you know it's your location. If you don't see bars in any of your road trip locations, your 4g modem is toast. See if you can buy a cheap (used)Air 10 (4g) device and swap the modem. They're easy to swap out.

I believe in the US, Air 10s use(d) T-Mobile and T-Mobile has started to phase out 4G.
Battery Backup: EcoFlow Delta 2