How to test a Climateline heated hose

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
mlmollenkamp
Posts: 95
Joined: Fri Apr 01, 2016 3:46 pm

How to test a Climateline heated hose

Post by mlmollenkamp » Mon Jan 19, 2026 3:28 pm

Hose failures have come up off and on. I just went through a batch of about 10 hoses to check them electrically and thought I'd pass on my procedure. Note that I've never found a test procedure from Resmed, this is purely my invention.
Note that I was testing S10 hoses, but the S11 hoses should be the same.

You need a digital multimeter. It doesn't need to be expensive or fancy, although one with a temperature probe is useful and isn't a lot a lot more money.

First I tested using the temperature probe on my multimeter.
I have a felt cover on the hose and slipped the temperature probe between the felt and the hose, pushing it up from the machine end about a foot or so.
I put on the mask and ran the machine normally, breathing normally.
With my machine hose temperature set to 86 degrees I read 78 degrees after about 5 minutes, the starting temperature was the same as the room, 68 degrees.
The preheat feature on the machine does not seem to preheat the hose, only the humidifier.

Hose electrical testing:
First, Inspect the pins on the hose where it plugs into the machine. They should be solid gold color without any green or other discoloration. If there is any discoloration, discard the hose.

There are 3 pins on the connector that plugs into the machine.

With the connector facing up:
Test between the pin on the left and the center pin with the Ohm meter set to read a low resistance, usually 200 ohm range.
The resistance should be 9.7 ohms, although the exact resistance may depend on the calibration of your meter.
When I tested 10 tubes they were all exactly 9.7 on my meter.
A bad tube will probably read open (very high resistance), but its conceivable that they could short given the way the wires are wound on the hose. If its less than 9 or more than 10.5 on most meters I think it's bad, but I've never found a bad hose to know for sure.

Now set the meter to read a higher resistance (usually the 20K range) Test between the center pin and the right most pin. It should read about 12 k Ohms. The actual reading depends on temperature.
Reach inside the mask end of the tube and place your finger against the pin that sticks into the center of the connector. Hold it there for a minute or two, then read the resistance again. It should go down.
Mine went from 12.5 k at room temperature to 10.7 k with my finger on the sensor.
This is the thermistor that tells the machine what the air temperature is. If its way off the machine will think the temperature is higher or lower than it actually is.

The machine connector can become dirty. Especially if you touch the connector on the hose with dirty hands, or hands with cream on them etc. before connecting it.
You can clean it with contact cleaner that is available online. I like Deoxit D5 by Caig, but any cleaner that's safe for plastics should be OK.
If a known good hose isn't working try cleaning the connector on the machine.
First cover the air opening so no cleaner can get inside. A good way to do this is to connect an old non Climateline tube to cover the hole and outside of the pipe.
Use a small burst of spray directly into the connector. Now attach and detach a Climateline hose a few times to wipe the pins inside. Wipe off the pins on the hose with a lint free cloth after connecting it each time.
Cleanup any spray that might have gotten on the machine.

Finally, there is a document online, "airsense10-aircurve10_disinfection-guide_amer_eng.pdf" that gives the Resmed official instructions for cleaning machines and hoses for reuse in medical facilities.



I hope this is helpful.

_________________
Machine: Aircurve 11 asv
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: P30i is modified with Dreamwear headgear.

User avatar
LSAT
Posts: 13399
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: How to test a Climateline heated hose

Post by LSAT » Mon Jan 19, 2026 4:42 pm

Why do you have 10 heated hoses and why do they need testing? I don't see the point.

User avatar
jimbud
Posts: 1484
Joined: Sat Dec 01, 2018 8:19 pm

Re: How to test a Climateline heated hose

Post by jimbud » Mon Jan 19, 2026 5:43 pm

LSAT wrote:
Mon Jan 19, 2026 4:42 pm
Why do you have 10 heated hoses and why do they need testing? I don't see the point.
Hose fetish and nothing better to do? :lol:

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

Sleepzilla
Posts: 41
Joined: Fri Nov 28, 2025 3:39 pm

Re: How to test a Climateline heated hose

Post by Sleepzilla » Tue Jan 20, 2026 10:05 am

.
Last edited by Sleepzilla on Tue Feb 10, 2026 8:51 am, edited 1 time in total.

User avatar
Nocibur
Posts: 454
Joined: Sat Jan 25, 2025 6:08 am

Re: How to test a Climateline heated hose

Post by Nocibur » Tue Jan 20, 2026 10:54 am

The definitive manual for troubleshooting ResMed Climateline tubing and similar devices is covered extensively in Electrical Testing And Maintenance Handbook, Vol. 11, AKA "The Electrician's Bible". Referring to Chapter 5, page 86, it is noted that evaluation of heated tubing can be classified into one of two categories:

"Worky"; and

"No Worky".

Grumpy48
Posts: 455
Joined: Tue May 12, 2020 7:57 am

Re: How to test a Climateline heated hose

Post by Grumpy48 » Tue Jan 20, 2026 2:21 pm

Nocibur wrote:
Tue Jan 20, 2026 10:54 am
The definitive manual for troubleshooting ResMed Climateline tubing and similar devices is covered extensively in Electrical Testing And Maintenance Handbook, Vol. 11, AKA "The Electrician's Bible". Referring to Chapter 5, page 86, it is noted that evaluation of heated tubing can be classified into one of two categories:

"Worky"; and

"No Worky".
When I worked in engineering we had another word for 'No Worky'....... 'FUBAR'. There are some slight variants on it's exact meaning, but they all suggest the same condition of a piece of equipment.

mlmollenkamp
Posts: 95
Joined: Fri Apr 01, 2016 3:46 pm

Re: How to test a Climateline heated hose

Post by mlmollenkamp » Tue Jan 20, 2026 3:32 pm

1. I didn't say this was intended to diagnose or test the machine. Only to verify that the hose sensor and heater are still connected and not shorted. There aren't many failure modes for these things and if the resistances are correct it's almost certain the hose is OK. It's just a go/no go hose test, not a test of the entire system. I'm trying to help anyone who wants to verify that their hose isn't defective. Without measuring a few hoses it's difficult to know what the typical resistances are, I've not seen them published. It's better to rule out the hose before you decide the machine is defective.
People sometimes post that they don't think their hose heat is working, or that they've had a lot of hose failures. This is step one of a logical trouble shooting process.

2. I have so many hoses because I test them, clean them thoroughly then give them to people who's insurance doesn't cover them. I was once in that position and it would have been nice to have a spare hose, or a working one for that matter. Its a shame to throw away a perfectly good $40 hose twice a year. I used to pay for them and used them at least 2 or 3 years.

3. I like to change to a clean hose fairly regularly. Cleaning only one hose at a time is wasted effort, I can clean 3 or 4 just as quickly so I always have one to swap in.

I didn't intend start a fight, just hoping to help anyone who isn't sure if their hose is OK.

Feel free to do better.

_________________
Machine: Aircurve 11 asv
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack
Additional Comments: P30i is modified with Dreamwear headgear.

User avatar
Nocibur
Posts: 454
Joined: Sat Jan 25, 2025 6:08 am

Re: How to test a Climateline heated hose

Post by Nocibur » Tue Jan 20, 2026 4:45 pm

mlmollenkamp wrote:
Tue Jan 20, 2026 3:32 pm
Feel free to do better.
OK, let's take everything apart!!

A11 Service Manual:

https://www.manualslib.com/manual/34521 ... tml#manual

User avatar
zonker
Posts: 11436
Joined: Fri Jun 19, 2015 4:36 pm

Re: How to test a Climateline heated hose

Post by zonker » Tue Jan 20, 2026 6:36 pm

mlmollenkamp wrote:
Tue Jan 20, 2026 3:32 pm


Feel free to do better.
LOVE it!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Nocibur
Posts: 454
Joined: Sat Jan 25, 2025 6:08 am

Re: How to test a Climateline heated hose

Post by Nocibur » Wed Jan 21, 2026 2:47 am

zonker wrote:
Tue Jan 20, 2026 6:36 pm
mlmollenkamp wrote:
Tue Jan 20, 2026 3:32 pm


Feel free to do better.
LOVE it!
You love it that mollincramp's panties are in a bunch?

That's awful!!

User avatar
Nocibur
Posts: 454
Joined: Sat Jan 25, 2025 6:08 am

Re: How to test a Climateline heated hose

Post by Nocibur » Wed Jan 21, 2026 3:06 am

Sleepzilla wrote:
Tue Jan 20, 2026 10:05 am
The S10 Climate Line hoses have 3 pins at the electrical connector, and the newer S11 hoses have 4 pins.
I'd like to know how many users even own a DVM, much less how to use it.

That said, on A11 the outer 2 pins are the thermistor contacts, inner 2 the heater (they now have separate grounds).

User avatar
Conrad
Posts: 700
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: How to test a Climateline heated hose

Post by Conrad » Wed Jan 21, 2026 7:44 am

Nocibur wrote:
Wed Jan 21, 2026 3:06 am
Sleepzilla wrote:
Tue Jan 20, 2026 10:05 am
The S10 Climate Line hoses have 3 pins at the electrical connector, and the newer S11 hoses have 4 pins.
I'd like to know how many users even own a DVM, much less how to use it.

That said, on A11 the outer 2 pins are the thermistor contacts, inner 2 the heater (they now have separate grounds).
*Waves hand*

I have five of them, one bench model and four handhelds, all Flukes, and I know how to use them, mostly. lol
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.

User avatar
Conrad
Posts: 700
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: How to test a Climateline heated hose

Post by Conrad » Wed Jan 21, 2026 7:47 am

Conrad wrote:
Wed Jan 21, 2026 7:44 am
Nocibur wrote:
Wed Jan 21, 2026 3:06 am
Sleepzilla wrote:
Tue Jan 20, 2026 10:05 am
The S10 Climate Line hoses have 3 pins at the electrical connector, and the newer S11 hoses have 4 pins.
I'd like to know how many users even own a DVM, much less how to use it.

That said, on A11 the outer 2 pins are the thermistor contacts, inner 2 the heater (they now have separate grounds).

*Waves hand*

I have five of them, one bench model and four handhelds, all Flukes, and I know how to use them, mostly. lol




To the OP

"There are 3 pins on the connector that plugs into the machine. With the connector facing up:"

In the description of your testing, is the opening of the heated tube facing you or facing away from you when you ID'ed the connectors? This makes a big difference, ya know?
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.

Sleepzilla
Posts: 41
Joined: Fri Nov 28, 2025 3:39 pm

Re: How to test a Climateline heated hose

Post by Sleepzilla » Wed Jan 21, 2026 8:25 am

.
Last edited by Sleepzilla on Tue Feb 10, 2026 8:48 am, edited 1 time in total.

Grumpy48
Posts: 455
Joined: Tue May 12, 2020 7:57 am

Re: How to test a Climateline heated hose

Post by Grumpy48 » Wed Jan 21, 2026 9:04 am

Nocibur wrote:
Wed Jan 21, 2026 3:06 am
Sleepzilla wrote:
Tue Jan 20, 2026 10:05 am
The S10 Climate Line hoses have 3 pins at the electrical connector, and the newer S11 hoses have 4 pins.
I'd like to know how many users even own a DVM, much less how to use it.

That said, on A11 the outer 2 pins are the thermistor contacts, inner 2 the heater (they now have separate grounds).
I think about 7 DVM's in different locations about the house and the workshops and 3 analog VOM's.

I confess that I also measured the AS10 heated hoses when I thought I was having a problem with one. My readings compare about the same as what mlmollenkamp measured. At low ohm measurements take into account the leads of the meter resistance and subtract that out of the hose measurement.