I need some help setting up my AirCurve 10 Vauto please

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
WHB74
Posts: 7
Joined: Sat Jul 31, 2021 5:00 am

I need some help setting up my AirCurve 10 Vauto please

Post by WHB74 » Mon Jun 09, 2025 2:34 am

This is probably been asked before I would think.

I bought a ResMed AirCurve 10 and I want to set it up in VAuto mode. I have been using an AirSense 11 in APAP mode for 2 years. My apnea level is very low but my Flow limitation is higher than it should be.

I am enclosing a screenshot from OSCAR so that if you want to help, you can see my Device Settings. Thank you in advance for any help you will provide to me.


Screen Shot 2025-06-07 at 04.09.40.png
Screen Shot 2025-06-07 at 04.09.40.png (279.89 KiB) Viewed 2548 times

User avatar
Pugsy
Posts: 65030
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I need some help setting up my AirCurve 10 Vauto please

Post by Pugsy » Mon Jun 09, 2025 6:10 am

Are you having much, if any, nasal congestion?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

WHB74
Posts: 7
Joined: Sat Jul 31, 2021 5:00 am

Re: I need some help setting up my AirCurve 10 Vauto please

Post by WHB74 » Mon Jun 09, 2025 1:40 pm

No. Every night before I go to sleep I use nasal spray to clear my nose.

Do you have any suggestions on what to set my AirCurve 10 VAuto settings to?

Thanks.

User avatar
Pugsy
Posts: 65030
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I need some help setting up my AirCurve 10 Vauto please

Post by Pugsy » Mon Jun 09, 2025 2:19 pm

Are you wanting to sort of mimic the apap settings with the VAuto but use a little more PS than you can get using EPR?

Your AHI is nothing so other than the flow limitations your report is better than mine. :lol:

Try this
Mode....VAuto
EPAP.... 8
PS..... 4
Max IPAP.... 14
Leave all the other comfort features and settings to the factory defaults. The defaults work very well for most people.

I have no idea if this will help or not but gotta start somewhere.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

WHB74
Posts: 7
Joined: Sat Jul 31, 2021 5:00 am

Re: I need some help setting up my AirCurve 10 Vauto please

Post by WHB74 » Tue Jun 10, 2025 12:08 pm

Thank you.

WHB74
Posts: 7
Joined: Sat Jul 31, 2021 5:00 am

Re: I need some help setting up my AirCurve 10 Vauto please

Post by WHB74 » Tue Jun 10, 2025 12:10 pm

Yes my AHI is good. My main concern is my Flow Limitation is too high and I want to get that problem solved.