Machine Servicing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Zaiforce
Posts: 98
Joined: Thu Dec 14, 2023 3:53 pm

Machine Servicing

Post by Zaiforce » Wed May 21, 2025 5:24 pm

Hi there,

Just wondering if I could have some advice if possible.

My mother is using a ResMed Lumis 150 VPAP-ST bi-level machine which was prescribed to her after multiple trips to ICU for type 2 respiratory failure with Co2 build-up.

She received her machine in March 2024 and so far it has been doing well for her, no ICU visits since using it :).

Her pressure settings are 20 IPAP / 10 EPAP if that is relevant info.

We clean her tubing/mask/humidifier tank etc. regularly however I have recently read that ResMed recommends the machines be professionally serviced every 6 months to keep them functioning properly, in which they clean and service parts within the machine that the user cannot maintain themselves.

We were unaware of this and therefore we haven't had this done yet and her machine has now been used every night for 14 months and unfortunately our house has a lot of dust which might not help (I replace the filter about every 2 months for the dust reason).

I have read online that if you don't have the machines serviced their pressures can lower unintendedly over time (despite the machine still displaying the intended pressure on the screen).

Does anybody have any advice on this? Is this machine servicing important and how do you go about organising it, just contact ResMed?

We live in a pretty rural area, it could be difficult for us to organise the machine servicing without her having to go without the machine for a period of time, but I am not sure.

Thank you for any advice :).

_________________
Mask: Evora Full Face Mask - Fitpack

User avatar
lazarus
Posts: 1425
Joined: Sat Mar 12, 2022 3:21 pm

Re: Machine Servicing

Post by lazarus » Wed May 21, 2025 6:04 pm

Is this supposed 6-month service/maintenance protocol written in print in the ResMed manual that came with the machine for home use?

User avatar
Pugsy
Posts: 65285
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Machine Servicing

Post by Pugsy » Wed May 21, 2025 6:18 pm

Zaiforce wrote:
Wed May 21, 2025 5:24 pm
I have recently read that ResMed recommends the machines be professionally serviced every 6 months to keep them functioning properly,
Where did you read this?
This is the first I have ever heard of this.
I have been using ResMed machines for over 10 years now and I have never had any of them "serviced"...nor have I ever seen documentation as to the need.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Zaiforce
Posts: 98
Joined: Thu Dec 14, 2023 3:53 pm

Re: Machine Servicing

Post by Zaiforce » Wed May 21, 2025 6:23 pm

lazarus wrote:
Wed May 21, 2025 6:04 pm
Is this supposed 6-month service/maintenance protocol written in print in the ResMed manual that came with the machine for home use?
Hi there,

Thank you for the reply, I haven't checked the user manual yet, but the location I seen the recommendation on was the ResMed website on this page:

https://www.resmed.com.au/blog/the-esse ... ap-machine

_________________
Mask: Evora Full Face Mask - Fitpack

Zaiforce
Posts: 98
Joined: Thu Dec 14, 2023 3:53 pm

Re: Machine Servicing

Post by Zaiforce » Wed May 21, 2025 6:25 pm

Pugsy wrote:
Wed May 21, 2025 6:18 pm
Zaiforce wrote:
Wed May 21, 2025 5:24 pm
I have recently read that ResMed recommends the machines be professionally serviced every 6 months to keep them functioning properly,
Where did you read this?
This is the first I have ever heard of this.
I have been using ResMed machines for over 10 years now and I have never had any of them "serviced"...nor have I ever seen documentation as to the need.
Thank you for the reply also.

Your comment is a relief, as I was worried we were missing something very important!

The recommendation I found was on the ResMed website here:
https://www.resmed.com.au/blog/the-esse ... ap-machine

I did some Googling and found mention of it on other sites as well, but I wonder how important it actually is.

The comment on one of the sites about pressures dropping over time (despite displaying as if they are correct on the screen) if the machines aren't serviced was what worried me most.

_________________
Mask: Evora Full Face Mask - Fitpack

User avatar
Pugsy
Posts: 65285
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Machine Servicing

Post by Pugsy » Wed May 21, 2025 7:29 pm

Thanks for providing the link.
Won't change anything with me though....just another ResMed "recommendation" that I will opt to continue to ignore. :lol:
I have been on cpap over 15 years now and this is the first time I have heard of this "recommendation".

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Zaiforce
Posts: 98
Joined: Thu Dec 14, 2023 3:53 pm

Re: Machine Servicing

Post by Zaiforce » Wed May 21, 2025 7:40 pm

Pugsy wrote:
Wed May 21, 2025 7:29 pm
Thanks for providing the link.
Won't change anything with me though....just another ResMed "recommendation" that I will opt to continue to ignore. :lol:
I have been on cpap over 15 years now and this is the first time I have heard of this "recommendation".
Thank you, I figured the vast majority of people probably wouldn't do this, just wanted to check the opinions as this forum is always very helpful with advice :)

I did reach out to ResMed via an enquiry on their website to ask how important it is and how I would even go about it, it will be interesting to see what they say.

Thanks again.

_________________
Mask: Evora Full Face Mask - Fitpack

User avatar
lazarus
Posts: 1425
Joined: Sat Mar 12, 2022 3:21 pm

Re: Machine Servicing

Post by lazarus » Wed May 21, 2025 8:12 pm

As a general rule, home PAP machines are only serviced when a problem presents, which that link's wording hints at, despite the later "recommendation." There is no oil to change or anything. It is designed to work perfectly for at least 5 years.

User avatar
Respirator99
Posts: 403
Joined: Mon Jul 05, 2021 12:39 am
Location: Australia

Re: Machine Servicing

Post by Respirator99 » Wed May 21, 2025 9:29 pm

It seems Resmed have just invented a new revenue stream. There's really nothing in these machines that need "servicing". I suspect they just blow out any dust. If you are in a dusty area, you might want to increase the frequency of filter changes - I do mine monthly.

The machines come with a five year warranty in Australia so long as you sign up for the SleepVantage program. https://www.sleepvantage.com.au/ This tells me the machines will last much longer than five years, and I've never heard of anybody getting them serviced.
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ

I have no medical training or qualifications. Take my advice for what it's worth.

Zaiforce
Posts: 98
Joined: Thu Dec 14, 2023 3:53 pm

Re: Machine Servicing

Post by Zaiforce » Wed May 21, 2025 10:17 pm

Thank you everyone, seems like it's not much to worry about, I kind of guessed that might be the case but just wanted to check the advice of others here.

Thank you as always :)

_________________
Mask: Evora Full Face Mask - Fitpack

Zaiforce
Posts: 98
Joined: Thu Dec 14, 2023 3:53 pm

Re: Machine Servicing

Post by Zaiforce » Wed May 21, 2025 10:18 pm

Respirator99 wrote:
Wed May 21, 2025 9:29 pm
The machines come with a five year warranty in Australia so long as you sign up for the SleepVantage program. https://www.sleepvantage.com.au/ This tells me the machines will last much longer than five years, and I've never heard of anybody getting them serviced.
I had not heard of the SleepVantage program, I wonder if it is too late for us to sign up for that, thank you I will look into it :).

_________________
Mask: Evora Full Face Mask - Fitpack

ozij
Posts: 10527
Joined: Fri Mar 18, 2005 11:52 pm

Re: Machine Servicing

Post by ozij » Thu May 22, 2025 7:33 pm

Zaiforce wrote:
Wed May 21, 2025 5:24 pm
She received her machine in March 2024 and so far it has been doing well for her, no ICU visits since using it .
Great news!
Zaiforce wrote:
Wed May 21, 2025 7:40 pm
I did reach out to ResMed via an enquiry on their website to ask how important it is and how I would even go about it, it will be interesting to see what they say.
Pugsy wrote:
Wed May 21, 2025 7:29 pm
Thanks for providing the link.
Won't change anything with me though....just another ResMed "recommendation" that I will opt to continue to ignore. :lol:
I have been on cpap over 15 years now and this is the first time I have heard of this "recommendation".
Ditto. More like 20 years for me.
Zaiforce wrote:
Wed May 21, 2025 7:40 pm
I did reach out to ResMed via an enquiry on their website to ask how important it is and how I would even go about it, it will be interesting to see what they say.
Please share their response.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

Zaiforce
Posts: 98
Joined: Thu Dec 14, 2023 3:53 pm

Re: Machine Servicing

Post by Zaiforce » Fri May 23, 2025 12:17 am

ozij wrote:
Thu May 22, 2025 7:33 pm
Zaiforce wrote:
Wed May 21, 2025 5:24 pm
She received her machine in March 2024 and so far it has been doing well for her, no ICU visits since using it .
Great news!
Zaiforce wrote:
Wed May 21, 2025 7:40 pm
I did reach out to ResMed via an enquiry on their website to ask how important it is and how I would even go about it, it will be interesting to see what they say.
Pugsy wrote:
Wed May 21, 2025 7:29 pm
Thanks for providing the link.
Won't change anything with me though....just another ResMed "recommendation" that I will opt to continue to ignore. :lol:
I have been on cpap over 15 years now and this is the first time I have heard of this "recommendation".
Ditto. More like 20 years for me.
Zaiforce wrote:
Wed May 21, 2025 7:40 pm
I did reach out to ResMed via an enquiry on their website to ask how important it is and how I would even go about it, it will be interesting to see what they say.
Please share their response.
Thank you ozij, I will make sure to post their response if they send me one.

Cheers.

_________________
Mask: Evora Full Face Mask - Fitpack