Leak vs events

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
desertguy
Posts: 1
Joined: Wed Nov 27, 2024 12:01 pm

Leak vs events

Post by desertguy » Thu Dec 12, 2024 5:52 pm

Long time user...just got a new Airsense 10 (old one had over 30,000 hrs) with an N30i mask. Feels pretty good, as before. However, I am using MyAir to monitor on this machine, and over the past 2 weeks have had 25-35 L/min leak nightly. However, only 1-2 events per hour and I feel fine in the morning.

Question: Is it okay to have some leakage if my events are well controlled?

(Before you ask, I have tried a few other masks this week, and yes nasal pillows do improve leakage but they don't fit as well. Despite less leakage (20-22 L/min), the events per hour are still 1-2)

User avatar
zonker
Posts: 11399
Joined: Fri Jun 19, 2015 4:36 pm

Re: Leak vs events

Post by zonker » Fri Dec 13, 2024 12:06 am

desertguy wrote:
Thu Dec 12, 2024 5:52 pm


Question: Is it okay to have some leakage if my events are well controlled?

yes. if the leaks don't wake you, I think you're fine. if you want a more detailed explanation, we'd need to see some of your charts. myair doesn't supply them. Oscar will. for the latest iteration of Oscar software, see this-

viewtopic/t189097/StickyOSCAR-160--Is-N ... lable.html
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
LSAT
Posts: 13374
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Leak vs events

Post by LSAT » Fri Dec 13, 2024 7:19 pm

Youre MyAir report gives you an explanation on how the score is reported. On your report card, tap the leak area and you will get the monthly scores. In the upper RH corner there is an (?). Tap on it to see how the score is calculated. Your machine can compensate for leaks up to about 24L. To much above that may affect your AHI.