ResMed Air Sense 10 Air Blowing Cold
ResMed Air Sense 10 Air Blowing Cold
Awoke this morning at usual time with air blowing cold and nasal passages feeling very irritated. I noticed that the humidifier water tub was empty, used up. My question is does this machine sense water is gone and shut off the heat for the water tub? Or does the air just blow a lot colder without water in the tub? I decided to test it and added water to the tub and see what happened. Much to my relief the machine started blowing warmed air and may not be broke as I assumed. Sure would appreciate your take and any information on this.
- ChicagoGranny
- Posts: 15419
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: ResMed Air Sense 10 Air Blowing Cold
1. Humid air is warmer than dry air.
2. ResMed has no heat shutoff for an empty humidifier tank.
https://ap.resmed.com/knowledge/what-ha ... t-of-water
2. ResMed has no heat shutoff for an empty humidifier tank.
https://ap.resmed.com/knowledge/what-ha ... t-of-water
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Re: ResMed Air Sense 10 Air Blowing Cold
The air the machine blows is your room air. It would be the same temp. as the air you are breathing in your room. It is the moist air from the humidifier that warms the air you breathe. If you run out of water, the machine cannot produce humidity and you are back to the temp. of your room.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: ResMed Air Sense 10 Air Blowing Cold
Humid air is warmer than dry air? How's that work? Can they not both be the same temp?ChicagoGranny wrote: ↑Fri Nov 29, 2024 2:16 pm1. Humid air is warmer than dry air.
2. ResMed has no heat shutoff for an empty humidifier tank.
https://ap.resmed.com/knowledge/what-ha ... t-of-water
Sorry, that makes no sense at all.
Humid air is more humid than dry air, that's the only fact here. They can both be the same temp, or different. Humid air isn't inherently warmer than dry air.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
It takes less time to do a thing right than to explain why you did it wrong.
Do not regret growing older. It is a privilege denied to many...
It takes less time to do a thing right than to explain why you did it wrong.
- ChicagoGranny
- Posts: 15419
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: ResMed Air Sense 10 Air Blowing Cold
Don't overlook the context. The water vapor carries the heat to the nasal passages more effectively than room air.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
- Dog Slobber
- Posts: 4269
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: ResMed Air Sense 10 Air Blowing Cold
Yes.ChicagoGranny wrote: ↑Sat Nov 30, 2024 11:36 amDon't overlook the context. The water vapor carries the heat to the nasal passages more effectively than room air.
And probably more importantly (and in context), humans do a poor job of judging temperature, what we feel is heat transfer, the ability to transfer heat from our bodies, or to our bodies. The difference may seem subtle, but it's not.
This becomes obvious when comparing two objects that are identical in temperature but have two very different heat conductivity rates.
Put a block of wood and a metal item on your counter long enough for them to reach room temperature. They both will be identical temperature. But, touching the metal item will be much cooler than the wooden one. We don't feel the temperature of the items, but the heat transfer, and the metal item draws much more heat from us.
Less humid air may be at the same temperature as humid air, but it feels a lot cooler because the drier air draws more heat from our passages.
We also see this phenomenon when people believe their heated hose isn't working because their hose doesn't "feel" warm. Their hoses don't feel very warm because the plastic tube is not a very good heat conductor.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: ResMed Air Sense 10 Air Blowing Cold
Dog Slobber wrote: ↑Sat Nov 30, 2024 12:02 pm
And probably more importantly (and in context), humans do a poor job of judging temperature, what we feel is heat transfer, the ability to transfer heat from our bodies, or to our bodies.

thank you for this wonderful explanation.
have you ever considered teaching?
(ducks,runs)
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: ResMed Air Sense 10 Air Blowing Cold
Well! Thanks to all here as this is clarifies what I experienced. The air blowing out of the my cpap hose was probably around 68 degrees as we have our heat set down cooler for better sleeping. I increased the water in the humidifier tank and the effect was comfortable once more. Nothing wrong with my aging cpap that adding a little more water didn’t fix.
Re: ResMed Air Sense 10 Air Blowing Cold
Yes, humid air FEELS warmer, subjectively, than dry air at the same temperature.
Thus
Damn it's so hot! True, but it's a dry heat...
lol
Thus
Damn it's so hot! True, but it's a dry heat...
lol
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
It takes less time to do a thing right than to explain why you did it wrong.
Do not regret growing older. It is a privilege denied to many...
It takes less time to do a thing right than to explain why you did it wrong.