Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
antinatalism
Posts: 21
Joined: Sun Oct 28, 2018 11:43 pm

Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by antinatalism » Sat Nov 23, 2024 5:26 pm

1. What’s the official term for the blue Left Air Outlet (left photo below)?

Monthly, I replace filter — it remains clean and white. Daily, I clean water chamber with vinegar, then refill with distilled water.

When I woke up, I saw white flakes in the water chamber (right photo). I dumped the water, again cleaned chamber, refilled with distilled water, and pushed in humidifier ― but when I turned on AirSense, I again saw these flakes in the water chamber!

So I pulled, and left, out the water chamber. When I turned on AirSense, I saw the Left Blue Outlet spurt out white flakes! But nothing in my bedroom changed ― last week, there weren’t these flakes!

2. Why’s AirSense spouting out these flakes? Where are these flakes from — inside the machine? What are these flakes — Buildup?

Image

User avatar
Dog Slobber
Posts: 4232
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by Dog Slobber » Sat Nov 23, 2024 9:35 pm

Are you using a room humidifier?

They generate white particles (mineral) from the water. Your ResMed is not generating it, it's blowing it through the system.
Battery Backup: EcoFlow Delta 2

antinatalism
Posts: 21
Joined: Sun Oct 28, 2018 11:43 pm

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by antinatalism » Sun Nov 24, 2024 12:41 am

No - no room humidifier. But even if I use a room humidifier that generates mineral from water, wouldn't AirSense's hypoallergenic filter block those minerals from invading AirSense?

User avatar
Conrad
Posts: 666
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by Conrad » Sun Nov 24, 2024 8:26 am

Dog Slobber wrote:
Sat Nov 23, 2024 9:35 pm
Are you using a room humidifier?

They generate white particles (mineral) from the water. Your ResMed is not generating it, it's blowing it through the system.
Every humidifier that I've ever seen states that the use of distilled, or similar, water is required for use.

No minerals do disperse into your air in that case.
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.

User avatar
Conrad
Posts: 666
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by Conrad » Sun Nov 24, 2024 8:31 am

antinatalism wrote:
Sat Nov 23, 2024 5:26 pm
1. What’s the official term for the blue Left Air Outlet (left photo below)?

Monthly, I replace filter — it remains clean and white. Daily, I clean water chamber with vinegar, then refill with distilled water.

When I woke up, I saw white flakes in the water chamber (right photo). I dumped the water, again cleaned chamber, refilled with distilled water, and pushed in humidifier ― but when I turned on AirSense, I again saw these flakes in the water chamber!

So I pulled, and left, out the water chamber. When I turned on AirSense, I saw the Left Blue Outlet spurt out white flakes! But nothing in my bedroom changed ― last week, there weren’t these flakes!

2. Why’s AirSense spouting out these flakes? Where are these flakes from — inside the machine? What are these flakes — Buildup?

Image


How's the intake on the other side look? Under the filter?

I periodically take a moistened Qtip and clean the crap that builds up on the machine underneath the edges of the filter. Where some air seems to get past the edges a bit and leaves a bit of dust or something.
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.

User avatar
Dog Slobber
Posts: 4232
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by Dog Slobber » Sun Nov 24, 2024 11:16 am

Conrad wrote:
Sun Nov 24, 2024 8:26 am
Dog Slobber wrote:
Sat Nov 23, 2024 9:35 pm
Are you using a room humidifier?

They generate white particles (mineral) from the water. Your ResMed is not generating it, it's blowing it through the system.
Every humidifier that I've ever seen states that the use of distilled, or similar, water is required for use.

No minerals do disperse into your air in that case.
Humidifiers sold in areas where distilled water is an abundant consumer product do not state distilled only. So basically the bulk of non-north american markets.

And people selectively read instructions, and of those who do read instructions, they selectively follow instructions.

This problem has been reported a few times, and in most, room humidifiers have been the cause.
Battery Backup: EcoFlow Delta 2

antinatalism
Posts: 21
Joined: Sun Oct 28, 2018 11:43 pm

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by antinatalism » Sun Nov 24, 2024 3:35 pm

Conrad wrote:
Sun Nov 24, 2024 8:31 am
How's the intake on the other side look? Under the filter?

Squeaky clean! To everyone else — please don't go off topic. Pls forgo any discussion of a room humidifier.

User avatar
Dog Slobber
Posts: 4232
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by Dog Slobber » Sun Nov 24, 2024 4:10 pm

antinatalism wrote:
Sun Nov 24, 2024 3:35 pm
Conrad wrote:
Sun Nov 24, 2024 8:31 am
How's the intake on the other side look? Under the filter?

Squeaky clean! To everyone else — please don't go off topic. Pls forgo any discussion of a room humidifier.
So by everyone else, you mean me.

Humidifiers may no longer be *your* solution, but they are still relevant to the topic. Even though you created the topic, topics are for all members to potentially benefit from.

Pushy controlling people tend to shutdown suggestions. Good luck solving your problem
Battery Backup: EcoFlow Delta 2

Jdougc74
Posts: 49
Joined: Thu Oct 17, 2024 6:39 am

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by Jdougc74 » Sun Nov 24, 2024 6:28 pm

I would be concerned enough to consider replacing this unit, or at least looking into the motor chamber. What does your DME say?
https://www.youtube.com/watch?v=vXoSxPP7shw
What if smaller particles were getting through? Maybe use an inline bacteria filter and see if any get caught? Don't use the inline filter for long, just a test to see if particles get trough to the tubing.
I had to replace my Phillips DreamStation foam air chamber with a foamless one.

_________________
MachineMask
Additional Comments: No Humidity during summer months, Oscar, I use a larger P10

User avatar
Conrad
Posts: 666
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by Conrad » Mon Nov 25, 2024 7:36 am

Dog Slobber wrote:
Sun Nov 24, 2024 11:16 am
Conrad wrote:
Sun Nov 24, 2024 8:26 am
Dog Slobber wrote:
Sat Nov 23, 2024 9:35 pm
Are you using a room humidifier?

They generate white particles (mineral) from the water. Your ResMed is not generating it, it's blowing it through the system.
Every humidifier that I've ever seen states that the use of distilled, or similar, water is required for use.

No minerals do disperse into your air in that case.
Humidifiers sold in areas where distilled water is an abundant consumer product do not state distilled only. So basically the bulk of non-north american markets.

And people selectively read instructions, and of those who do read instructions, they selectively follow instructions.

This problem has been reported a few times, and in most, room humidifiers have been the cause.
Did you read what I said? If folks don't read the instructions, who's fault is that? If they read but don't follow, who's fault is that?
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.

User avatar
Conrad
Posts: 666
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by Conrad » Mon Nov 25, 2024 7:38 am

antinatalism wrote:
Sun Nov 24, 2024 3:35 pm
Conrad wrote:
Sun Nov 24, 2024 8:31 am
How's the intake on the other side look? Under the filter?

Squeaky clean! To everyone else — please don't go off topic. Pls forgo any discussion of a room humidifier.
Attempting to control/mandate the replies here, or anywhere, usually doesn't end well. Best to just filter out what pertains to you.
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.

lynninnj
Posts: 1324
Joined: Mon Jul 25, 2022 8:56 am

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by lynninnj » Sat Dec 07, 2024 11:59 am

I didn't notice if you are using distilled water. If you are NOT and if you have hard water, the vinegar is possibly loosening it up but still leaving bits behind (like shower scale for example.)

Are you using only distilled?
If not, you should probably be doing so now.

HTH

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: Newbie who loves her machine!
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.

antinatalism
Posts: 21
Joined: Sun Oct 28, 2018 11:43 pm

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by antinatalism » Sat Dec 07, 2024 2:38 pm

lynninnj wrote:
Sat Dec 07, 2024 11:59 am
I didn't notice if you are using distilled water.
Why didn't you carefully peruse? I wrote twice "refill with distilled water" and "refilled with distilled water".

User avatar
Dog Slobber
Posts: 4232
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by Dog Slobber » Sun Dec 08, 2024 12:35 pm

antinatalism wrote:
Sat Dec 07, 2024 2:38 pm
lynninnj wrote:
Sat Dec 07, 2024 11:59 am
I didn't notice if you are using distilled water.
Why didn't you carefully peruse? I wrote twice "refill with distilled water" and "refilled with distilled water".
You don't understand how people use forms.

Many don't "carefully fully", nor should they be expected to.

They see topics, maybe scan them, offer suggestions. You are the one asking for help, and your pushy, controlling crap, jsut takes away any incentive to try suggestions.

Enjoy your morning flakes.
Battery Backup: EcoFlow Delta 2

User avatar
Conrad
Posts: 666
Joined: Mon Jul 30, 2012 9:10 am
Location: Northern Illinois

Re: Why’s AirSense 10 spewing white flakes, out of the blue left air exhaust?

Post by Conrad » Mon Dec 09, 2024 7:02 am

Dog Slobber wrote:
Sun Dec 08, 2024 12:35 pm
antinatalism wrote:
Sat Dec 07, 2024 2:38 pm
lynninnj wrote:
Sat Dec 07, 2024 11:59 am
I didn't notice if you are using distilled water.
Why didn't you carefully peruse? I wrote twice "refill with distilled water" and "refilled with distilled water".
You don't understand how people use forms.

Many don't "carefully fully", nor should they be expected to.

They see topics, maybe scan them, offer suggestions. You are the one asking for help, and your pushy, controlling crap, jsut takes away any incentive to try suggestions.

Enjoy your morning flakes.
So, when a person posts something and someone comments on said post, we shouldn't expect that the person commenting on the post has read the post fully? Why comment if you haven't read and understood the post? Yeah, good question I know...
ResMed AirCurve 10 Vauto Swift FX

Do not regret growing older. It is a privilege denied to many...

It takes less time to do a thing right than to explain why you did it wrong.