3 months in - still high clear airway - suggestions?
3 months in - still high clear airway - suggestions?
Hi all:
I am about three months in and I can tell a difference between before and after the CPAP. Some nights my score is great and other nights my clear airways are out of control. I have tried to search, but I can not find a graph like mine in a post.
I posted two examples, including a breakdown of a clear airway. If there is a clear airway, they are typically bunched together like the second attachment.
I did an at home test and use AirTouch F20. I tried to use nasal pillows and what not, but the pressure would bust out of my mouth when I would sleep and wake me up.
I am about three months in and I can tell a difference between before and after the CPAP. Some nights my score is great and other nights my clear airways are out of control. I have tried to search, but I can not find a graph like mine in a post.
I posted two examples, including a breakdown of a clear airway. If there is a clear airway, they are typically bunched together like the second attachment.
I did an at home test and use AirTouch F20. I tried to use nasal pillows and what not, but the pressure would bust out of my mouth when I would sleep and wake me up.
- Attachments
-
- 11.8.png (217.56 KiB) Viewed 16408 times
-
- 11.12.png (220.36 KiB) Viewed 16408 times
-
- 11.12 CA.png (219.06 KiB) Viewed 16408 times
Re: 3 months in - still high clear airway - suggestions?
Same issue last night. I am definitely more rested than I was pre-CPAP. I know the clear airway could be caused by the cpap, but the nights are mixed. I either have clear airways like this, or it is completely fine.
- Attachments
-
- 11.13.png (227.15 KiB) Viewed 16296 times
- Miss Emerita
- Posts: 3732
- Joined: Sun Nov 04, 2018 8:07 pm
Re: 3 months in - still high clear airway - suggestions?
Welcome! Could you let us know the breakdown of your AHI when you had your sleep test? You could scan and post the sleep report, with your identifying information blanked out.
Do you live at a high altitude? Do you take medications or drugs that are opioids or like opioids?
(When you post your next chart, could you include the bit on the left that shows the indexes for each component of the AHI?)
Do you live at a high altitude? Do you take medications or drugs that are opioids or like opioids?
(When you post your next chart, could you include the bit on the left that shows the indexes for each component of the AHI?)
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: 3 months in - still high clear airway - suggestions?
Hello,
1) I did an at-home sleep study, so unfortunately it did not break down thee type. But, I did attach it. The diagnosis said this: Findings are consistent with Severe Obstructive Sleep Apnea (G47.33). High heart rate variability noted which may be suggestive of a cardiac arrhythmia. Note that heart rate measurement on home sleep apnea tests can
be subject to artifact and is not confirmed by EKG.
2) I am definitely high altitude - 9300 ft.
3) No opioids, but I do live in CO so I do take a small amount of things to help me fall asleep.
4) 34 y/o male, not overweight, no other health complications.
5) Reattached today's report. My tablet doesn't allow me to shrink the left bar down so I hid that part for now.
1) I did an at-home sleep study, so unfortunately it did not break down thee type. But, I did attach it. The diagnosis said this: Findings are consistent with Severe Obstructive Sleep Apnea (G47.33). High heart rate variability noted which may be suggestive of a cardiac arrhythmia. Note that heart rate measurement on home sleep apnea tests can
be subject to artifact and is not confirmed by EKG.
2) I am definitely high altitude - 9300 ft.
3) No opioids, but I do live in CO so I do take a small amount of things to help me fall asleep.
4) 34 y/o male, not overweight, no other health complications.
5) Reattached today's report. My tablet doesn't allow me to shrink the left bar down so I hid that part for now.
- Attachments
-
- 11.13v2.png (265.06 KiB) Viewed 16276 times
-
- Sleep study.png (235.14 KiB) Viewed 16276 times
- Miss Emerita
- Posts: 3732
- Joined: Sun Nov 04, 2018 8:07 pm
Re: 3 months in - still high clear airway - suggestions?
Your report says all your events were unclassified apneas, so that isn’t helpful at all. I think you need to ask your sleep doctor for an in-lab sleep test so you can really figure out what is going on. I’d assume your sleep doctor has a lot of experience with high-altitude sleep stuff too.
Please let the sleep doctor know what your indexes last night were for CA, OA, and H. That should get his or her attention.
Please let the sleep doctor know what your indexes last night were for CA, OA, and H. That should get his or her attention.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: 3 months in - still high clear airway - suggestions?
I have the appointment on Friday. I just wonder if it is altitude driven as my AHI was less than 5 three days in a row when I was in Texas
- ChicagoGranny
- Posts: 15081
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: 3 months in - still high clear airway - suggestions?
A 2011 study found that people with moderate-to-severe OSA living at higher elevation were more likely to develop central sleep apnea as well. But if they used a properly calibrated PAP machine, those who had developed central sleep apnea at higher altitudes did show clear improvements. A little bit more info:
There are solutions. Your doctor might be able to help, even without an in-lab sleep study. I would rely on her judgment if you have faith in her.Altitude associated central apnea has a significant negative effect on the quality of OSA treatment obtained during PAP titration for patients living at the altitudes addressed in this study.
https://pmc.ncbi.nlm.nih.gov/articles/PMC3227706/
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
- Miss Emerita
- Posts: 3732
- Joined: Sun Nov 04, 2018 8:07 pm
Re: 3 months in - still high clear airway - suggestions?
Many thanks for that link, ChicagoGranny. I'm especially struck by this:
The higher frequency of central apnea at altitudes above 6000 ft. indicates that such altitudes may be a relative contraindication for the routine use of limited channel OSA screeners and automatic technologies that are unable to clearly differentiate between central and obstructive apneas. The difficulty of treating OSA due to central apneas occurring on treatment with PAP at altitude is likely to lead to a higher level of untreated OSA among these patients.
So an in-lab test might be valuable after all.
To the OP: yes, high altitudes can definitely increase CA indexes. Since you are having an OA index over 5 with treatment, it's complicated. Higher pressure might help with the OAs but make the CAs worse.
I'll be quite interested to hear what your sleep doctor says.
The higher frequency of central apnea at altitudes above 6000 ft. indicates that such altitudes may be a relative contraindication for the routine use of limited channel OSA screeners and automatic technologies that are unable to clearly differentiate between central and obstructive apneas. The difficulty of treating OSA due to central apneas occurring on treatment with PAP at altitude is likely to lead to a higher level of untreated OSA among these patients.
So an in-lab test might be valuable after all.
To the OP: yes, high altitudes can definitely increase CA indexes. Since you are having an OA index over 5 with treatment, it's complicated. Higher pressure might help with the OAs but make the CAs worse.
I'll be quite interested to hear what your sleep doctor says.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: 3 months in - still high clear airway - suggestions?
Thank you all. I have been reading studies about the altitude and centrals. I am going out of town next week at a much lower elevation. The last time I did, my AHI was less than 5 each night. I will report back after next week.
Last night, AHI was 12 - comprised of 4 for OSA and the rest for centrals, but around when I was waking up. No change in what I did from previous night to last night.
My only concern is that if I go down to the city for an in-lab test, it will be almost half of the altitude as where I live, so I fear they won't see anything since I will be at a lower altitude.
Last night, AHI was 12 - comprised of 4 for OSA and the rest for centrals, but around when I was waking up. No change in what I did from previous night to last night.
My only concern is that if I go down to the city for an in-lab test, it will be almost half of the altitude as where I live, so I fear they won't see anything since I will be at a lower altitude.
- Miss Emerita
- Posts: 3732
- Joined: Sun Nov 04, 2018 8:07 pm
Re: 3 months in - still high clear airway - suggestions?
Oh, yes, getting a test at a lower altitude would not tell you much. Fingers crossed your body somehow works this out on its own, or with your sleep doctor's help. Keep us posted, would you?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: 3 months in - still high clear airway - suggestions?
Doc mentioned they want to have a steady but higher pressure and to check back in two months from now.
They acknowledged that I may need to get a bipap or ASV because of my altitude and severity of OSA, but they want to give it some more time.
At least I feel better than I did without CPAP - will update how I am doing over time.
FYI - I was reading other individuals that were at higher altitude and supplemented with oxygen and their centrals went away. But, my oxygen level has been at 95% or higher, so I would bet it is the altitude, but I am no doctor.
They acknowledged that I may need to get a bipap or ASV because of my altitude and severity of OSA, but they want to give it some more time.
At least I feel better than I did without CPAP - will update how I am doing over time.
FYI - I was reading other individuals that were at higher altitude and supplemented with oxygen and their centrals went away. But, my oxygen level has been at 95% or higher, so I would bet it is the altitude, but I am no doctor.
Re: 3 months in - still high clear airway - suggestions?
Figured I would close this thread out:
I had an in-lab sleep study. They put me in an altitude tent to mimic the altitude I am at while on an ASV machine. The ASV machine was able to control my mixed apnea and I was lower than a 5 AHI. I got approved for the ASV and hoping to get it once insurance has approved it.
From what I generally understand, (I am not a doc) is that I need a higher pressure to clear OSA, but the higher pressure leads to hyperventiliation which caused my central apnea.
https://pmc.ncbi.nlm.nih.gov/articles/PMC3227706/
I had an in-lab sleep study. They put me in an altitude tent to mimic the altitude I am at while on an ASV machine. The ASV machine was able to control my mixed apnea and I was lower than a 5 AHI. I got approved for the ASV and hoping to get it once insurance has approved it.
From what I generally understand, (I am not a doc) is that I need a higher pressure to clear OSA, but the higher pressure leads to hyperventiliation which caused my central apnea.
https://pmc.ncbi.nlm.nih.gov/articles/PMC3227706/
Re: 3 months in - still high clear airway - suggestions?
Thanks for coming up an updating us.
That helps all of us.
I for one would like to hear how you're doing with the ASV when you get it.
That helps all of us.
I for one would like to hear how you're doing with the ASV when you get it.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023