First day using Oscar

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
yribaar
Posts: 6
Joined: Mon Sep 23, 2024 8:23 am

First day using Oscar

Post by yribaar » Mon Sep 23, 2024 8:55 am

Hi,
It's been 30 years since I had a sleep study. I never thought about my settings too much using APAP. ( huge mistake i know)
some background: 63,6'1,240, bearded--- heart attack in 2020, congenital heart disease, diabetic

So i downloaded Oscar and was surprised at how awful my graphs looked.

https://imgur.com/a/6C2SZ8y

User avatar
Julie
Posts: 20042
Joined: Tue Feb 28, 2006 12:58 pm

Re: First day using Oscar

Post by Julie » Mon Sep 23, 2024 9:41 am

Do you have a question?

yribaar
Posts: 6
Joined: Mon Sep 23, 2024 8:23 am

Re: First day using Oscar

Post by yribaar » Mon Sep 23, 2024 9:50 am

Julie wrote:
Mon Sep 23, 2024 9:41 am
Do you have a question?
Why the flow rate is so jagged in relation to the hyponeas?

User avatar
Miss Emerita
Posts: 3732
Joined: Sun Nov 04, 2018 8:07 pm

Re: First day using Oscar

Post by Miss Emerita » Mon Sep 23, 2024 11:11 am

A couple of questions. Has your machine been reconditioned after the PR recall? And could you zoom in starting at about 1:20, posting a snippet of about 15 minutes? Finally, what is happening during the breaks in your usage of the machine?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

yribaar
Posts: 6
Joined: Mon Sep 23, 2024 8:23 am

Re: First day using Oscar

Post by yribaar » Mon Sep 23, 2024 11:52 am

Miss Emerita wrote:
Mon Sep 23, 2024 11:11 am
A couple of questions. Has your machine been reconditioned after the PR recall? And could you zoom in starting at about 1:20, posting a snippet of about 15 minutes? Finally, what is happening during the breaks in your usage of the machine?
This is the new machine they sent after the recall.
https://imgur.com/a/aQAXDo7
the breaks are peeing and going out with my dog. I turn off the machine because blue heelers love to cause trouble and mine thinks it's a leaf blower.

User avatar
Miss Emerita
Posts: 3732
Joined: Sun Nov 04, 2018 8:07 pm

Re: First day using Oscar

Post by Miss Emerita » Tue Sep 24, 2024 11:58 am

I had to look up blue heelers. What beautiful dogs! I'm sorry you have to get up during the night for yours, though.

Pee breaks: If you're a man, you might want to be sure you have your prostate checked. You might also want to restrict fluids for 3-4 hours ahead of bedtime. And if you wake up and think, "Well, as long as I'm awake, maybe I'll get up and pee," then instead try just rolling over and going back to sleep.

Thanks for the snippet. It's hard to know whether these hypopneas are obstructive or central (meaning your breathing dwindles without any obstruction). Some though not all of your spates of hypopneas are associated with snores, so maybe they're obstructive.

Do you sleep on your sides? Back? Some of each?

Given the OAs, flow limitations, snores, and hypopneas, I'd recommend raising your minimum pressure to 9. You don't have to do that all at once; you could start by changing it to 6 and then going up from there.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

yribaar
Posts: 6
Joined: Mon Sep 23, 2024 8:23 am

Re: First day using Oscar

Post by yribaar » Tue Sep 24, 2024 6:30 pm

Miss Emerita wrote:
Tue Sep 24, 2024 11:58 am
I had to look up blue heelers. What beautiful dogs! I'm sorry you have to get up during the night for yours, though.

Pee breaks: If you're a man, you might want to be sure you have your prostate checked. You might also want to restrict fluids for 3-4 hours ahead of bedtime. And if you wake up and think, "Well, as long as I'm awake, maybe I'll get up and pee," then instead try just rolling over and going back to sleep.

Thanks for the snippet. It's hard to know whether these hypopneas are obstructive or central (meaning your breathing dwindles without any obstruction). Some though not all of your spates of hypopneas are associated with snores, so maybe they're obstructive.

Do you sleep on your sides? Back? Some of each?

Given the OAs, flow limitations, snores, and hypopneas, I'd recommend raising your minimum pressure to 9. You don't have to do that all at once; you could start by changing it to 6 and then going up from there.
Can't resist linking a photo of Woody. He's 100% Honey Badger :)
https://imgur.com/a/hWsisYN
I'm a dude...or I was until I turned into a grumpy old man.
I always sleep on my side. ( 90% right side )

Thanks for your time. I'll come back after I make the changes.

User avatar
Miss Emerita
Posts: 3732
Joined: Sun Nov 04, 2018 8:07 pm

Re: First day using Oscar

Post by Miss Emerita » Tue Sep 24, 2024 7:19 pm

What a handsome pup! And I love the eye-patch. Keep us posted.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Julie
Posts: 20042
Joined: Tue Feb 28, 2006 12:58 pm

Re: First day using Oscar

Post by Julie » Tue Oct 01, 2024 11:07 am

Gorgeous dog!

yribaar
Posts: 6
Joined: Mon Sep 23, 2024 8:23 am

Re: First day using Oscar

Post by yribaar » Sat Nov 09, 2024 7:26 am

Miss Emerita wrote:
Tue Sep 24, 2024 7:19 pm
What a handsome pup! And I love the eye-patch. Keep us posted.
After setting the minimum pressure to 9 per your suggestion my AHI is below 2.
Thank you so much!

yribaar
Posts: 6
Joined: Mon Sep 23, 2024 8:23 am

Re: First day using Oscar

Post by yribaar » Sat Nov 09, 2024 7:30 am

Julie wrote:
Tue Oct 01, 2024 11:07 am
Gorgeous dog!
Woody says thanks.

User avatar
Miss Emerita
Posts: 3732
Joined: Sun Nov 04, 2018 8:07 pm

Re: First day using Oscar

Post by Miss Emerita » Sat Nov 09, 2024 12:57 pm

yribaar wrote:
Sat Nov 09, 2024 7:26 am
After setting the minimum pressure to 9 per your suggestion my AHI is below 2. Thank you so much!
Delighted to hear the news -- many thanks for coming back to share it. Feel free to post a chart if you think you'd might want to fine-tune, though my motto is "Don't mess with success."
Oscar software is available at https://www.sleepfiles.com/OSCAR/