Evora Full Face Mask Hose Joint Leakage multiple Units.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jrh9803
Posts: 3
Joined: Sat Aug 10, 2024 11:20 am

Evora Full Face Mask Hose Joint Leakage multiple Units.

Post by jrh9803 » Sat Aug 10, 2024 11:49 am

ResMed AirSense 10 AutoSet CPAP Machine With HumidAir (Card-to-Cloud Version)
I have found the mask for me and have tried many over the last 5 years. Is anyone else reporting hose joint leakage with this mask: Evora Full Face Mask RT-1005.
I have had 3 bad units this year.

User avatar
Okie bipap
Posts: 3566
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Evora Full Face Mask Hose Joint Leakage multiple Units.

Post by Okie bipap » Sat Aug 10, 2024 7:24 pm

My wife and I have both been using this mask since it first came out and we have had no problems with them. Are you getting leakage whee the hose snaps onto the mask frame? Could it be a tear in the hose caused by rough handling?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

jrh9803
Posts: 3
Joined: Sat Aug 10, 2024 11:20 am

Re: Evora Full Face Mask Hose Joint Leakage multiple Units.

Post by jrh9803 » Sun Aug 11, 2024 5:10 pm

No, this is an ongoing issue with the last 3 masks that my local supplier has sent home with me right out of the doctor's office.

My RRT sent me home with a new one that was leaking as I needed a mask to sleep. I will be returning it again this week. Are others reporting this issue?


---Snip-----

Reply to:

ResMed AirSense 10 AutoSet CPAP Machine With HumidAir (Card-to-Cloud Version)
I have found the mask for me and have tried many over the last 5 years. Is anyone else reporting hose joint leakage with this mask: Evora Full Face Mask RT-1005.
I have had 3 bad units this year.

DaveCLT
Posts: 4
Joined: Wed May 09, 2012 9:16 am

Re: Evora Full Face Mask Hose Joint Leakage multiple Units.

Post by DaveCLT » Wed Sep 25, 2024 1:51 pm

Hi, I'm pretty sure what you call the hose joint leakage point is the same issue I've had...

I've repeatedly had the same connector swivel joint leak issue with this mask across a half dozen masks and have created a fix using heat shrink tubing (pictures below). It prohibits the swivel part from swiveling, but I haven't found that to be much of an issue.

The 1 inch tubing we had in the garage fit, took care not to overheat in any one area to make the cpap tube too hot. The little black sharpie mark I made to note how far up the resmed climateline connector came up the tube so the shrink tubing didn’t interfere with that connection.
IMG_4503 (1).jpeg
IMG_4503 (1).jpeg (69.35 KiB) Viewed 8902 times
IMG_4504.jpeg
IMG_4504.jpeg (114.88 KiB) Viewed 8902 times
I called and spoke to a US rep, plus submitted a product design defect resolution request on their website

I would encourage others who use and like this mask but also encounter the leak issue at the swivel to do the same. Squeaky wheel works sometimes.

If you have the Evora mask with a swivel connector and experience the same leak from the connector tube, please contact them and ask for the design flaw to be fixed. Love the mask other than this!

https://www.fphcare.com/us/corporate/con...enquiries/

Contact numbers https://www.fphcare.com/us/corporate/contact-us/
Machine: Remed Airsense 11
Mask Type: Full face mask
Mask Make & Model: fisher paykel evora
Humidifier: Resmed
CPAP Pressure: 14
CPAP Software: SleepyHead myAir

User avatar
beachwhale1994
Posts: 1
Joined: Sun Feb 02, 2025 1:16 pm

Re: Evora Full Face Mask Hose Joint Leakage multiple Units.

Post by beachwhale1994 » Sun Mar 09, 2025 12:31 pm

Thank you for the tip with the heat shrink, it works perfectly. I am really missing the swivel, is there any kind of add on swivel or maybe quick disconnect swivel to connect the mask hose to the hose coming from the machine? I was looking at the Resmed 60926 hose I think it would do the trick, but I am open to suggestions.
Thank you all.

_________________
MachineMask