Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
I got a used AirCurve 10 ASV from someone who got a new machine from their insurance. It has 2811 hours on it and seems to function fine but I'm having trouble getting it set up like my AirSense 10 Autoset. I have been using the same settings for more than 5 years and I think I'm happy with my results. My doctor certainly is. I tried setting the AirCurve to CPAP mode Set pressure of 16, start pressure of 4.0, ramp time of 5 minutes. For some reason it seemed like the machine was more aggressive than my Airsense and it even seemed louder. Is this normal? If I want to try it in another mode, like ASVAuto, what shoud the Min and Max settings be for EPAP and PS? Start EPAP? Or should I just sell the machine to someone who needs these functions? Since the cost even with insurance paying half is very high, I thought it would be awesome if I had this machine as a backup and maybe even replace my current machine which is almost at the end of it's "life expectancy." I appreciate any insight and advice. Thanks.
- Attachments
-
- 20240724_205513.jpg (527.08 KiB) Viewed 13750 times
-
- 20240724_205458.jpg (320.94 KiB) Viewed 13750 times
-
- 20240724_205438.jpg (346.89 KiB) Viewed 13750 times
Re: Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
You might get close with the settings in terms of setting the ASV machine to function like the AutoSet.
I am not sure that the PS setting in AutoSV mode will do what you want or need though.
Tell us exactly what settings you were using on the AutoSet....mode and all settings.
You will need for PS minimum and PS maximum to equal the same setting and that's where there might be a problem.
I don't know if it will let you set PS min and max to equal the same setting. It may want a range.
I am not sure that the PS setting in AutoSV mode will do what you want or need though.
Tell us exactly what settings you were using on the AutoSet....mode and all settings.
You will need for PS minimum and PS maximum to equal the same setting and that's where there might be a problem.
I don't know if it will let you set PS min and max to equal the same setting. It may want a range.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
I took photos of the AirSense 10 settings from the technician menu and put them in my post. Are there more that I am not aware of? Mode is AutoSet Max Pressure is 16.0 Min Pressure is 4.2 Full Face Mask Ramp Time 5 Min Start Pressure is 4.2 EPR is On, Full Time at level 3 Humidity level is 4 Standard tube no AB filter Essentials is On SmartStart is On
Re: Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
So these are your AutoSet settings?
Pressure 4.2 minimum and max is 16?
EPR at 3 full time
To use AutoSV mode and have it function like the AutoSet
Minimum EPAP 4.2
Maximum IPAP 16
PS minimum 3 and PS maximum 3 (but I don't know if your machine will let you do that or not)
Are you aware that with the minimum starting at 4.2 (and when using ramp) that the EPR does not do much since the machine won't go below 4 cm. So that EPR at 3 doesn't really do a damn thing until your pressure goes from 4.2 to 7?
It went from 4.2 to 16 in one fell swoop....and yes at that pressure any machine will be a bit louder.
Pressure 4.2 minimum and max is 16?
EPR at 3 full time
To use AutoSV mode and have it function like the AutoSet
Minimum EPAP 4.2
Maximum IPAP 16
PS minimum 3 and PS maximum 3 (but I don't know if your machine will let you do that or not)
Are you aware that with the minimum starting at 4.2 (and when using ramp) that the EPR does not do much since the machine won't go below 4 cm. So that EPR at 3 doesn't really do a damn thing until your pressure goes from 4.2 to 7?
After 5 minutes of ramp the machine went straight to 16 cm when in fixed cpap mode. No wonder it felt a lot more "aggressive"...it was a LOT more pressure than you were using with the AutoSet.
It went from 4.2 to 16 in one fell swoop....and yes at that pressure any machine will be a bit louder.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- Respirator99
- Posts: 383
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
From the Aircurve 10 ASV clinical manual:
In ASV or ASVAuto mode, PSmin can be 0–6 cm H2O; PSmax 5–20 cm H2O
So I'd try the settings Pugsy suggested, but you can't have PSmin = PSmax = 0. My feeling is that the PS will stay low unless the machine sees a need to increase it, eg when you have a central apnea event.
In ASV or ASVAuto mode, PSmin can be 0–6 cm H2O; PSmax 5–20 cm H2O
So I'd try the settings Pugsy suggested, but you can't have PSmin = PSmax = 0. My feeling is that the PS will stay low unless the machine sees a need to increase it, eg when you have a central apnea event.
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
Re: Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
Okay, I'm trying ASVAuto but Max EPAP only goes to 15 not 16. I set Min a PS to 3 and Max PS defaulted to 8 and won't go lower so I left it there. If I understand what Pugsy said the settings on my Airsense 10 Autoset of EPR = 3 was not really doing anything? I'll let you know how things go with the Aircurve.
Thanks for your help!
Thanks for your help!
-
- Posts: 523
- Joined: Sat Feb 13, 2021 2:40 am
Re: Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
FYI the max pressure can be up to 23 (Max EPAP 15 + Max PS 8 )
There may or may not mean some aerophasia.
You can try to work backward and subtract min or max PS from your Max Pressure.
min EPAP 4.2
max EPAP 13 (16 - 3 min PS ; your max pressure here will be 21) or
or max EPAP 8 (16 - max PS 8 ; max pressure here is 16)
min PS 3
max PS 8
Good luck in your experiments.
ResMed Lumis Tx
Re: Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
That's not what Pugsy said.mannme1 wrote: ↑Thu Jul 25, 2024 9:14 pmOkay, I'm trying ASVAuto but Max EPAP only goes to 15 not 16. I set Min a PS to 3 and Max PS defaulted to 8 and won't go lower so I left it there. If I understand what Pugsy said the settings on my Airsense 10 Autoset of EPR = 3 was not really doing anything? I'll let you know how things go with the Aircurve.
Thanks for your help!
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
"Max PS defaulted to 8" doesn't mean you can't give it a higher number.GrumpyHere wrote: ↑Thu Jul 25, 2024 9:39 pmFYI the max pressure can be up to 23 (Max EPAP 15 + Max PS 8 )
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
-
- Posts: 523
- Joined: Sat Feb 13, 2021 2:40 am
Re: Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
My bad, I meant for his specific setting the Max Pressure/IPAP will go up to 23.ozij wrote: ↑Thu Jul 25, 2024 10:05 pm"Max PS defaulted to 8" doesn't mean you can't give it a higher number.GrumpyHere wrote: ↑Thu Jul 25, 2024 9:39 pmFYI the max pressure can be up to 23 (Max EPAP 15 + Max PS 8 )
The ASV will go up to 30 cm h2o for IPAP.
ResMed Lumis Tx
- vandownbytheriver
- Posts: 542
- Joined: Sat Feb 03, 2024 11:42 pm
Re: Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
You can use this machine in straight CPAP mode if you examine your charts and determine what your good pressure is... obviously not 16cm, right? We use Oscar (www.sleepfiles.com/oscar) to determine this... as has been noted, the Median pressure is probably a good place to start.
IMO you should sell this machine and get three or four Airsense 10 Autosets... you don't need ASV therapy and this machine is not going to give you an easy setting that mimics your Autoset. It's worth a lot more than a used Autoset. ASV is for a different kind of OSA, you don't have complex sleep apnea.
IMO you should sell this machine and get three or four Airsense 10 Autosets... you don't need ASV therapy and this machine is not going to give you an easy setting that mimics your Autoset. It's worth a lot more than a used Autoset. ASV is for a different kind of OSA, you don't have complex sleep apnea.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |
Re: Can I set an AirCurve 10 ASV like my AirSense 10 Autoset
Hi again. For two nights I used the machine with the following settings:
Min EPAP 4.2
Max EPAP 8
Min PS 3 Max PS 8
These settings that GrumpyHere had explained worked great. However, I think I’m going to take the advice of vandownbytheriver and sell this AirCurve with less than 2900 hours on it and look for an AirSense 10 Autoset like I already have. Someone who really needs all the settings available on the AirCurve should use this. Thanks for all the great advice and explanations everyone!!
Min EPAP 4.2
Max EPAP 8
Min PS 3 Max PS 8
These settings that GrumpyHere had explained worked great. However, I think I’m going to take the advice of vandownbytheriver and sell this AirCurve with less than 2900 hours on it and look for an AirSense 10 Autoset like I already have. Someone who really needs all the settings available on the AirCurve should use this. Thanks for all the great advice and explanations everyone!!