Airsense 10 Autoset driven crazy by cardioballistic artifacts

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
vandownbytheriver
Posts: 309
Joined: Sat Feb 03, 2024 11:42 pm

Airsense 10 Autoset driven crazy by cardioballistic artifacts

Post by vandownbytheriver » Thu Apr 18, 2024 9:33 pm

This was an interesting case from one of the FB groups I participate in... poor person had the damndest flow and respiratory rates I'd ever seen! Take a look, shared with permission:

https://sleephq.com/public/2b1513c0-9ed ... d748526b9a

You have to zoom in and see what the poor machine is counting... it's counting their *heartbeat*, and finding flow limitations too! Now *that's* some thumping. Took a little while to figure it out... the crazy fuzzy Resp Rate graph was actually the clue. The obvious solution was to turn off APAP (sorry Chunky) and tell the machine to shut up and pump air. Here's their result:

https://sleephq.com/public/c037b3c8-bec ... 2c4a6925b9

Almost *too* good... might have them turn down some!

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion)
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape.