Events per hour has increased

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
chzuck
Posts: 58
Joined: Tue Feb 13, 2024 12:32 pm
Location: Pennsylvania
Contact:

Events per hour has increased

Post by chzuck » Tue Feb 13, 2024 2:41 pm

I am new to this site and to OSCAR. I found this site because I was looking for help in finding why my events per hour went up on average this past month. I am looking for things I should check that would cause this to happen. I use the ResMed app and my average was running 6 and under. Now I am running 6 to 10 plus on average.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure: EPAP 15.0 IPAP 19.0 (cmH2O)

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Events per hour has increased

Post by Miss Emerita » Tue Feb 13, 2024 4:30 pm

Welcome! It isn't possible to figure out why your AHI is going up without having more information to go on. Do you know the breakdown of your AHI into three types of events: obstructive apneas, hypopneas, and central apneas (aka clear airway apneas)?

What kind of ResMed machine do you have? Depending on the model of machine, we can give you some pointers on how to get this information if you don't already have it.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

chzuck
Posts: 58
Joined: Tue Feb 13, 2024 12:32 pm
Location: Pennsylvania
Contact:

Re: Events per hour has increased

Post by chzuck » Tue Feb 13, 2024 5:49 pm

Hopefully I snipped the data you ask for from OSCAR. I am still new at this. Only found OSCAR yesterday. My machine is an AirCurve 10 vauto.

Data.jpg
Data.jpg (13.83 KiB) Viewed 10209 times

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure: EPAP 15.0 IPAP 19.0 (cmH2O)

User avatar
Pugsy
Posts: 65400
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Events per hour has increased

Post by Pugsy » Tue Feb 13, 2024 5:52 pm


_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Events per hour has increased

Post by Miss Emerita » Wed Feb 14, 2024 12:08 am

Oh, you’re using Oscar—that’s great! Follow the instructions in Pugsy’s link and post a recent chart.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

chzuck
Posts: 58
Joined: Tue Feb 13, 2024 12:32 pm
Location: Pennsylvania
Contact:

Re: Events per hour has increased

Post by chzuck » Wed Feb 14, 2024 9:31 am

This is my first try at this, so I am not sure if I have all that you wanted.
screenshot-20240214-103022.png
screenshot-20240214-103022.png (85.66 KiB) Viewed 10132 times

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure: EPAP 15.0 IPAP 19.0 (cmH2O)

User avatar
Pugsy
Posts: 65400
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Events per hour has increased

Post by Pugsy » Wed Feb 14, 2024 9:49 am

So how come you are using fixed pressure?

Those little clusters....you need more pressure during those times.

Related to sleeping position maybe???? On your back????

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

chzuck
Posts: 58
Joined: Tue Feb 13, 2024 12:32 pm
Location: Pennsylvania
Contact:

Re: Events per hour has increased

Post by chzuck » Wed Feb 14, 2024 9:56 am

The pressures were set by my doctor. I sleep on my left side mostly and some on my back.
I have slept this way since I got the machine almost 2 years ago. My average events only went up around the start of the new year and I am trying to find out what has changed.
Stats.png
Stats.png (40.17 KiB) Viewed 10121 times

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure: EPAP 15.0 IPAP 19.0 (cmH2O)

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Events per hour has increased

Post by Miss Emerita » Wed Feb 14, 2024 12:00 pm

You're right that your AHI has gone up fairly recently, but it was never that great. Are you open to experimenting with your settings?

As for what might have changed and made a difference: Weight gain? Pain that sends you to back-sleeping more than you used to? Different set-up with your pillows? Sore throat? Jaw issues? Tooth issues? Any changes in how you side-sleep (e.g., curling up more)?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

chzuck
Posts: 58
Joined: Tue Feb 13, 2024 12:32 pm
Location: Pennsylvania
Contact:

Re: Events per hour has increased

Post by chzuck » Wed Feb 14, 2024 12:08 pm

Nothing that I can think of has changed in the things you mentioned. Not sure if I want to change pressures. I have issues now with mask sealing at times and I have no idea how to do that.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure: EPAP 15.0 IPAP 19.0 (cmH2O)

User avatar
Pugsy
Posts: 65400
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Events per hour has increased

Post by Pugsy » Wed Feb 14, 2024 12:20 pm

You know you could maybe try an experiment to see if what happens if you just go to extra measures to stay off your back and see if that helps or not.

It is easier said than done though...the staying off one's back thing. Been there and done that myself but for other reasons than OSA related stuff.

Might at least give you some sort of answer and then you can decide what to do about it at that time if needed.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

chzuck
Posts: 58
Joined: Tue Feb 13, 2024 12:32 pm
Location: Pennsylvania
Contact:

Re: Events per hour has increased

Post by chzuck » Wed Feb 14, 2024 1:12 pm

I do start out on my left side, but don't know if I stay there. This is so strange. I think back in December I was below 1 for events per hour. Periodically I would have a night over 6, but usually always below 10. I will try and monitor my sleep position.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure: EPAP 15.0 IPAP 19.0 (cmH2O)

chzuck
Posts: 58
Joined: Tue Feb 13, 2024 12:32 pm
Location: Pennsylvania
Contact:

Re: Events per hour has increased

Post by chzuck » Sat Mar 23, 2024 7:30 am

It is hard to know your sleep position while sleeping. I do know when I awakened I was on my back or partially on my right side. My events average has come down some. I am pretty new at this, so I am wondering if over time I may expect to experience periods of higher events per hour?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure: EPAP 15.0 IPAP 19.0 (cmH2O)

User avatar
zonker
Posts: 11490
Joined: Fri Jun 19, 2015 4:36 pm

Re: Events per hour has increased

Post by zonker » Sat Mar 23, 2024 11:56 am

chzuck wrote:
Sat Mar 23, 2024 7:30 am
I am pretty new at this, so I am wondering if over time I may expect to experience periods of higher events per hour?
it's been a month between your posts. what is your ahi now? the whole point of therapy is to get to the lowest ahi you can and still get restful and restorative sleep. sleep doctors will tell you an ahi of 5 is good. we here on the forum strive for an ahi of around 2.

this may require you to change the settings on your machine.

but first, can you post us some recent charts?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

chzuck
Posts: 58
Joined: Tue Feb 13, 2024 12:32 pm
Location: Pennsylvania
Contact:

Re: Events per hour has increased

Post by chzuck » Sat Mar 23, 2024 8:01 pm

I don't know how to change the air settings and did not know I was allowed. What would my doctor say. I thought the reason I had the sleep test was to determine the settings. Just asking because I am new at this.
I will post some results when I get a chance.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Pressure: EPAP 15.0 IPAP 19.0 (cmH2O)