Need help!
Need help!
I have started cpap over a month ago having so much trouble .. I am having all central Apenas they lower my pressure from 6-15 to 7-10. Still not doing good still having centrals .. can someone please help me figure this out .. what’s going on
Re: Need help!
Most people get some centrals from time to time. It's no big deal......Do you know what a Central Apnea is? How many are you getting? What is your AHI? I have been on CPAP for 15 years and still get several centrals during the night. My average AHI is 2+.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: Need help!
I been getting like 15 a night last 2 nights before it was 6 one night it was 1 .. just concern I am getting these so many a night
- ChicagoGranny
- Posts: 15462
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Need help!
15 centrals over 8 hours is not so many. Which model machine are you using? Are you using OSCAR? Have you looked to see how many seconds the centrals last?
How are you feeling during the day? Energetic? No excess sleepiness? Fatigued at bedtime but not stressed?
Can you post an OSCAR chart in this thread? Instructions are in my signature.
How are you feeling during the day? Energetic? No excess sleepiness? Fatigued at bedtime but not stressed?
Can you post an OSCAR chart in this thread? Instructions are in my signature.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Re: Need help!
How was your sleep quality? If you weren't asleep when those centrals were flagged they only mean you weren't asleep.
If you aren't asleep they don't really mean much.
See all these centrals on this image. I know for a fact I wasn't asleep when they were flagged because I had a bought of really bad insomnia that night.

_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Need help!
I had like 15 centrals in 2 hours .. I feel very tired and awful pretty much all day .. I try to post it.. if I can figure it out
Re: Need help!
I would bet that the centrals are like mine and related to not being asleep and the feeling crappy is related to poor sleep or not enough sleep.
You can use SleepHQ and that would allow us to zoom in on those centrals to see if you were asleep or not.
Or you can do a lot of zoomed in images using OSCAR.
The big overview OSCAR image doesn't let us see the flow rate zoomed in enough to see the breathing in detail enough to determine asleep or awake.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Need help!
So being awake can cause this type of reading? It kinda makes sense cause I wake up all night long .. I try and use sleep HQ .. where can I find that?