A gross question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Leatherneck232
Posts: 14
Joined: Thu Dec 21, 2023 8:10 am

A gross question

Post by Leatherneck232 » Sun Dec 24, 2023 7:45 am

Ok...here is a gross question I've not seen addressed on here yet.
So I'm not sick yet every morning I wake up with my throat full of snot. It takes an hour of coughing to clear it. But that makes my throat sore for the morning and sometimes causes a headache.
I do have somewhat of a runny nose due to the winter season so I wonder if the CPAP is forcing the snot down my throat?

This is a great forum. Thanks admins for making it work.

User avatar
lazarus
Posts: 1425
Joined: Sat Mar 12, 2022 3:21 pm

Re: A gross question

Post by lazarus » Sun Dec 24, 2023 10:31 am

Leatherneck232 wrote:
Sun Dec 24, 2023 7:45 am
CPAP is forcing
My understanding is that the only "force" CPAP normally exerts is that of keeping the airway open. There is a secondary effect of some slight expansion of the lower part of the lungs because of that pressure, but there is no force inherent in PAP that actually moves things around. It just keeps things open (that otherwise might be closed) as it slightly positively pressurizes the airway.

The powerful force that moves things around in your airway is your breathing, and that is obviously not a force you would want to attempt to limit.
Leatherneck232 wrote:
Sun Dec 24, 2023 7:45 am
somewhat of a runny nose due to the winter season
Seasons themselves do not directly cause a runny nose. Viruses may do that. And changes in air temperature and humidity may cause irritation that can cause similar results. So (1) taking precautions to avoid viruses, (2) adjusting indoor humidity to prevent irritating dryness of the air there, and (3) mitigating temperature changes by covering the nose with fabric or a mask when outdoors in cold, dry air may help.

Relatively new users of CPAP often have a little trouble at first getting the machine humidifier adjusted just right at night during use. That issue usually solves itself with a little time and experience, especially as the body gets used to the 'change in climate' in the upper airway from the constant venting from the mask.

https://www.resmed.com/en-us/sleep-apne ... ongestion/

I hope those suggestions/observations may be helpful even to those whose necks are said to be made of leather, Leatherneck232. :wink: :)

* quietly hopes Marines are still good at taking jokes *
Last edited by lazarus on Sun Dec 24, 2023 10:58 am, edited 1 time in total.

User avatar
Leatherneck232
Posts: 14
Joined: Thu Dec 21, 2023 8:10 am

Re: A gross question

Post by Leatherneck232 » Sun Dec 24, 2023 10:58 am

Thx for the great reply Lazarus!
The thing is I gotta believe many others have a similar experience.
So of course doing the things you mention do others have strategies besides drugs to cope with it?

User avatar
lazarus
Posts: 1425
Joined: Sat Mar 12, 2022 3:21 pm

Re: A gross question

Post by lazarus » Sun Dec 24, 2023 11:06 am

Since you haven't been on CPAP too very long, I would probably start by reading this blog entry from Nate Devore at CPAP.com:

https://www.cpap.com/blog/best-humidity ... p-machine/

User avatar
Pugsy
Posts: 65369
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: A gross question

Post by Pugsy » Sun Dec 24, 2023 11:11 am

Leatherneck232 wrote:
Sun Dec 24, 2023 10:58 am
The thing is I gotta believe many others have a similar experience.
If you mean cpap pushing snot down into your airway and lungs....no you are the first I have heard mention this.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Leatherneck232
Posts: 14
Joined: Thu Dec 21, 2023 8:10 am

Re: A gross question

Post by Leatherneck232 » Sun Dec 24, 2023 11:47 am

Yes. Thx Pugsy

So I've played with the humidity and tube temp. I run a humidifier also and keep the house humidity near 50 percent.
Right now I've settled on my machines humidity level at 3 and tube temp of 80. I get rainout above 3.

User avatar
LoneStar
Posts: 189
Joined: Wed Mar 11, 2009 1:35 pm
Location: south central Texas

Re: A gross question

Post by LoneStar » Sun Dec 24, 2023 12:13 pm

When I first started, I went with the recommendation that most people have - use a humidifier.

I have sinus problems.

The humidifier seriously exacerbated all the mucus. Like you, I woke up drowning it it.

So I changed my humidity to pass over. I don't turn on the humidifier, but the air "passes over" the water and I get a little moisture that is enough for me. No more drowning in mucus.

I had the Fisher Paykel H150 Humidifier. Now I just have the tank (HC325) velcroed to my CPAP table.

Oh, and I use a heated hose to prevent rainout. YMMV.

_________________
Machine: Airsense 10 Card to Cloud
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: APAP, humidifier on passover, heated hose
With xPAP therapy, do or do not; there is no try.

User avatar
Leatherneck232
Posts: 14
Joined: Thu Dec 21, 2023 8:10 am

Re: A gross question

Post by Leatherneck232 » Sun Dec 24, 2023 12:30 pm

Thx Lonestar
I will try the passive humidity tonight

User avatar
ChicagoGranny
Posts: 15520
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: A gross question

Post by ChicagoGranny » Sun Dec 24, 2023 1:22 pm

Leatherneck232 wrote:
Sun Dec 24, 2023 12:30 pm
Thx Lonestar
I will try the passive humidity tonight
Some of us never use the humidifier. Not even in cold, dry months when the central furnace is running a lot.

The purpose of the humidifier is to keep the nasal passages from becoming dry. Since my passages don't dry, I figure why mess with the humidifier?

lynninnj
Posts: 1324
Joined: Mon Jul 25, 2022 8:56 am

Re: A gross question

Post by lynninnj » Sun Dec 24, 2023 2:45 pm

Pugsy wrote:
Sun Dec 24, 2023 11:11 am
Leatherneck232 wrote:
Sun Dec 24, 2023 10:58 am
The thing is I gotta believe many others have a similar experience.
If you mean cpap pushing snot down into your airway and lungs....no you are the first I have heard mention this.
am I the only one reading this with Jethro Tull Aqualung playing in thier head?

I find the heat and humidity work much like a hot shower. If a hot shower creates the runny nose and snot movement then maybe cpap contribute similarly.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: Newbie who loves her machine!
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.

User avatar
Pugsy
Posts: 65369
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: A gross question

Post by Pugsy » Sun Dec 24, 2023 3:20 pm

lynninnj wrote:
Sun Dec 24, 2023 2:45 pm
Pugsy wrote:
Sun Dec 24, 2023 11:11 am
Leatherneck232 wrote:
Sun Dec 24, 2023 10:58 am
The thing is I gotta believe many others have a similar experience.
If you mean cpap pushing snot down into your airway and lungs....no you are the first I have heard mention this.
am I the only one reading this with Jethro Tull Aqualung playing in thier head?

I find the heat and humidity work much like a hot shower. If a hot shower creates the runny nose and snot movement then maybe cpap contribute similarly.
I am not disputing the fact that the humidity selection can impact nasal congestion and/or drainage.

I was only addressing this part
Leatherneck232 wrote:
Sun Dec 24, 2023 7:45 am
if the CPAP is forcing the snot down my throat?
And while cpap use can impact nasal congestion or drainage..... it isn't forcing snot down someone's throat.
The air pressure isn't great enough to force anything down someone's throat.
Now normal drainage....that's a whole different thing and has to be dealt with by the normal means (whatever works).

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: A gross question

Post by chunkyfrog » Sun Dec 24, 2023 3:30 pm

Mucus does its job by moving "stuff" out of you.
Thick mucus results from dehydration.
Drink enough water and your mucus thins out and works better.
The tricky part is reaching a balance between your heart, lungs, and bladder.
:lol:

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Leatherneck232
Posts: 14
Joined: Thu Dec 21, 2023 8:10 am

Re: A gross question

Post by Leatherneck232 » Sun Dec 24, 2023 3:55 pm

Balance. The $64,000 question lol.
Do I take more bladder meds or just drink less??

Will be interesting to see how I feel after a night of pass over humidity.

User avatar
bwexler
Posts: 1591
Joined: Mon Nov 08, 2010 11:52 pm
Location: San Marcos, Ca. USA

Re: A gross question

Post by bwexler » Sun Dec 24, 2023 4:03 pm

Leatherneck232

I have recently noticed a similar problem. I simply move on with my day and deal with it.
Until reading your post I had not given it much thought.

I suspect it may be more of an issue when I sleep on my back instead of my left side.

_________________
Mask: SleepWeaver 3D Soft Cloth Nasal CPAP Mask with Headgear
Additional Comments: AurCurve 10 ASV Also using Sleaplyhead 1.1, ResScan 6 and CMS50i

User avatar
Leatherneck232
Posts: 14
Joined: Thu Dec 21, 2023 8:10 am

Re: A gross question

Post by Leatherneck232 » Sun Dec 24, 2023 5:33 pm

Hi Bwexler

I sleep mostly on my sides so gravity may not be the issue for me.
The really annoying thing is that I sound sick all the time.