AIRVIEW & MyAir communications
AIRVIEW & MyAir communications
Does my Airsense II call into Resmed Airview every day, and if so, is there a schedule?
Or does The Airview system call my cpap every day (at their convenience)?
In order for Airview system to work acquire information properly, do I have to have an SD card installed?
( I am thinking about the detailed data for each day, flow rate data files etc, rather than just summary data)
Or does The Airview system call my cpap every day (at their convenience)?
In order for Airview system to work acquire information properly, do I have to have an SD card installed?
( I am thinking about the detailed data for each day, flow rate data files etc, rather than just summary data)
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
Last edited by dataq1 on Sat Nov 25, 2023 10:11 am, edited 1 time in total.
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
Re: AIRVIEW communications
You're always acting like you know everything, answer this dumb question yourself.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: AIRVIEW communications
I been told that Resmed’s day begins and ands at noon, so I’d think that the data is uploaded after that time. But how/who the upload is initiated, I’ve no info.
- Dog Slobber
- Posts: 4333
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: AIRVIEW communications
ResMed's day begins and ends at noon to reduce the complications caused by single sleep sessions spanning midnight. Nothing to do with uploading data. The ResMed noon-to-noon day preceeded the ability to upload data.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: AIRVIEW communications
But it wouldn’t make much sense to upload prior to the “end of Resmed’s day”.
- Dog Slobber
- Posts: 4333
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: AIRVIEW communications
Why?
* Edit *
My data for today, uploaded at least an hour ago at 8:13am, almost four hours before the end of a Resmed day.

_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: AIRVIEW communications
The only thing you need to do is to leave the machine plugged in and not connected to a socket that goes out with a light switch or a timer (or powers off in any other manner). If you're traveling, any missed data will get caught up when you return home.dataq1 wrote: ↑Wed Nov 22, 2023 1:35 pmDoes my Airsense II call into Resmed Airview every day, and if so, is there a schedule?
Or does The Airview system call my cpap every day (at their convenience)?
In order for Airview system to work acquire information properly, do I have to have an SD card installed?
( I am thinking about the detailed data for each day, flow rate data files etc, rather than just summary data)
Re: AIRVIEW communications
I suppose that it really doesn't make that much difference (timing of uploading to some Airview server), I've been lead to believe that each sleep session is tagged with the date and time, so that it's reasonable that the server assembling the data for a "day" might recognize that a divided sleep night.
But you've incidentally raised an associated question that the OP didn't ask.... Is the MyAir app communication (from the Cpap) via cellular or Bluetooth? - I just don't know !
Does the MyAir app get it's info directly from the cpap machine or does the app get it's info from some sort of Resmed server (in the cloud) that also supplies AirView reports to providers?
In any event, (in your example), if you had started another sleep session at 9:00am (after sync to MyAir), would that additional session show up as an add-on of Nov 22 data?
But you've incidentally raised an associated question that the OP didn't ask.... Is the MyAir app communication (from the Cpap) via cellular or Bluetooth? - I just don't know !
Does the MyAir app get it's info directly from the cpap machine or does the app get it's info from some sort of Resmed server (in the cloud) that also supplies AirView reports to providers?
In any event, (in your example), if you had started another sleep session at 9:00am (after sync to MyAir), would that additional session show up as an add-on of Nov 22 data?
Re: AIRVIEW communications
Cellular...only cellular which requires an internal modem within the machine to be able to "call out" and send the data.
Bluetooth is local to your residence only...like pairing with phones, other computers or the TV via your own personal WiFi connections from whatever.
Answered above....the machine uses the cellular modem to phone home to ResMed website of some sort. Now exactly how "home" does its interchange with MyAir or AirView...who knows. Or in my case....who cares.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: AIRVIEW communications
Digging futher, here is what Resmed says (on their website)
1) what/who initiates the upload to Resmed's server: the cpap
2) what is the schedule: one hour after pt stops using
3) Airsense11 seems to communicate with MyAir via Bluetooth.
Whether you need to have a SD card installed for AirView to collect flow rate data or mask pressure data streams is still up-in-the air.
Why care? if out of range for bluetooth or in airplane mode/ out of coverage .
So that would seem to answer some of the OP's question:All Air series devices contain internal SIM cards that transmit data automatically into AirView (device must be powered on, not in "Airplane mode", and in a cellular coverage area). There is no effort required by the patient, as the device automatically uploads data 1 hour after the patient stops using the device.
However, the AirSense 11 integrates Bluetooth connectivity, so you can see your sleep stats right in your myAir app and get an accurate picture of your sleep therapy.Apr 14, 2023
1) what/who initiates the upload to Resmed's server: the cpap
2) what is the schedule: one hour after pt stops using
3) Airsense11 seems to communicate with MyAir via Bluetooth.
Whether you need to have a SD card installed for AirView to collect flow rate data or mask pressure data streams is still up-in-the air.
Why care? if out of range for bluetooth or in airplane mode/ out of coverage .
- Dog Slobber
- Posts: 4333
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: AIRVIEW communications
No it does not. Therapy data is only sent to the cloud via cellular. Air 11's use bluetooth to:
- provide device info to the app
- control features such as testdrive
- communicate with bluetooth oxi-devices
No it isn't.
The ability to transmit any of the data is not dependant on the presence of an SD Card.
High Resolution data, 25Hz or greater (like most wave forms), are *never* transmitted from the device to ResMed. They can only be sent to the cloud, by bringing an SD Card to ResScan station, reading the card and then submitting to ResMed.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: AIRVIEW communications
Thanks, That is very useful to know. My SD card or it contents have never left my home. So, when my doctor or DME tells me that they are looking at a plot of my leak readings for a particular night, that is entirely plausible. On the other hand, if the doctor tells me that he is looking at plot of my pressure or flow readings for a particular night.... - he is probably fibbing.Dog Slobber wrote: ↑Thu Nov 23, 2023 10:59 amHigh Resolution data, 25Hz or greater (like most wave forms), are *never* transmitted from the device to ResMed. They can only be sent to the cloud, by bringing an SD Card to ResScan station, reading the card and then submitting to ResMed.
Now, you said that 'Airsense 11 does not communicate with with the MyAir app via bluetooth, but then went on to say that AS11 Bluetooth provides device info to the app (and I think you are meaning the MyAir app). That leaves me confused about what information is passed between the device and the MyAir app on my phone. Is there a distinction that I'm misunderstanding?
Re: AIRVIEW communications
To all,
Thanks for the input while I was OOP.
What I gathered from the various comments were in answer to my question:
1) Device to Resmed server (to provide data to Resmed Airview) is initiated by the device and about 1 hour after session ends.
This transfer of data does not include hi-resolution data (millisecond pressure and flow rate data for example)
2) Device data (except for hi-res ) to Myair app is accomplished via my Bluetooth enabled phone.
I now know that either taking my SD card or OSCAR printouts to my doctor provides a boatload of therapy information that my doctor does not have.
Thanks for the input while I was OOP.
What I gathered from the various comments were in answer to my question:
1) Device to Resmed server (to provide data to Resmed Airview) is initiated by the device and about 1 hour after session ends.
This transfer of data does not include hi-resolution data (millisecond pressure and flow rate data for example)
2) Device data (except for hi-res ) to Myair app is accomplished via my Bluetooth enabled phone.
I now know that either taking my SD card or OSCAR printouts to my doctor provides a boatload of therapy information that my doctor does not have.
_________________
| Machine: AirSense 11 Autoset |
| Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."
- Dog Slobber
- Posts: 4333
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: AIRVIEW communications
He might be looking at low resolution plots.
It does communicate, what it doesn't do is send any therapy data to the MyAir App.
Device information, such as the unique serial number on your device.RobbyM684 wrote: ↑Thu Nov 23, 2023 9:57 pmNow, you said that 'Airsense 11 does not communicate with with the MyAir app via bluetooth, but then went on to say that AS11 Bluetooth provides device info to the app (and I think you are meaning the MyAir app). That leaves me confused about what information is passed between the device and the MyAir app on my phone. Is there a distinction that I'm misunderstanding?
No, it's not.
Therapy data is never transferred to via Bluetooth.
It's uploaded to ResMed via cellular.
Hi-res data is never uploaded to ResMed (from the App or Device). It can only be read from the SD Card.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: AIRVIEW communications
I'm only looking for clarity, not an argument.
Are you saying that there is a low-res version of the millisecond-by-millisecond mask pressure or flow rates that is sent via the automatic daily upload?
On an other point of clarity, regarding Bluetooth comms to the MyAir app. I've always considered AHI data and usage hours as being (part of) therapy data. MyAir app clearly gets that therapy data via Bluetooth. So I guess we can say that MyAir (via Bluetooth) gets some therapy data.Dog Slobber wrote: ↑Sat Nov 25, 2023 7:38 amTherapy data is never transferred to via Bluetooth.
It's uploaded to ResMed via cellular.