AirSense 11 doesn't record to SD card once in a while

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
chipschap
Posts: 6
Joined: Tue Nov 16, 2010 10:55 pm
Location: Honolulu, HI

AirSense 11 doesn't record to SD card once in a while

Post by chipschap » Sat Sep 30, 2023 10:24 am

I've been using the Resmed Airsense 11 (auto) for a little over two weeks now. On two separate occasions, there was no detailed data on the SD card; Oscar told me "nothing there." All other days are fine.

I thought perhaps the SD card was not properly inserted, but examining the SD card data structure with a file browser I noticed there is a directory entry for the missing days, just no data in the directory.

Has anyone else run into this problem, or have any insights? Mahalo in advance!

User avatar
ChicagoGranny
Posts: 15081
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AirSense 11 doesn't record to SD card once in a while

Post by ChicagoGranny » Sat Sep 30, 2023 11:36 am

chipschap wrote:
Sat Sep 30, 2023 10:24 am
no detailed data
Does that mean the summary data was there?

Kind of an is-it-plugged-in question: Was the card perhaps locked on those days? If it was, the next time it was used unlocked, the summary data would download, but the detailed data is gone forever.

chipschap
Posts: 6
Joined: Tue Nov 16, 2010 10:55 pm
Location: Honolulu, HI

Re: AirSense 11 doesn't record to SD card once in a while

Post by chipschap » Sat Sep 30, 2023 2:18 pm

There was actually NO data on the SD card for the dates in question. The card was not locked. Actually something DID get written to the card: the directory entry for the day(s) on which the data was missing.

The machine was definitely on! I used it all night on those nights, and the very high level data (AHI and hours) was available on the touch screen.

I'm completely baffled. Why would it work most days but just write an empty directory on a couple of day?

Mahalo

User avatar
Pugsy
Posts: 64932
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AirSense 11 doesn't record to SD card once in a while

Post by Pugsy » Sat Sep 30, 2023 2:35 pm

My best guess is a corrupt file of some sort for some unknown reason.

I know I have had that happen a handful of times over the years and I know for a fact the SD card was inserted correctly and not locked (I never lock the card) and I know for a fact that I slept using the machine all night despite a big goose egg for usage or even summary data. Not the AirSense 11 model but the AirCurve 10 model....and I have seen it happen myself.

Never could figure it out even when I verified using ResScan software that it wasn't just an OSCAR software problem.
ResScan gave me big goose eggs for the same handful of nights.

Crap happens....as they say.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4159
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: AirSense 11 doesn't record to SD card once in a while

Post by Dog Slobber » Sun Oct 01, 2023 3:23 pm

It probably:
  • wasn't seated properly
  • was write protected
  • left in the computer over night
You can't make the assumption that because you see a directory entry for the date with missing detail data, that the SD was inserted sand data was being written.

Summary and compliance data will be written to the SD Card for every date, upon SD insertion. This data is retained in the device memory and written to the SD Card, if it's not present.

ie:
Say you don't insert the SD Card into your ResMed tonight, and then use it. No detail data will be written. But upon insertion of the SD Card, the Summary and Compliance data will get written to the SD Card into a directory named after it's respective data.

BTW - a common problem reported for the Air 11, has been missing detail data attributed to an improperly seated SD Card.

lynninnj
Posts: 1320
Joined: Mon Jul 25, 2022 8:56 am

Re: AirSense 11 doesn't record to SD card once in a while

Post by lynninnj » Sun Oct 01, 2023 4:51 pm

chipschap wrote:
Sat Sep 30, 2023 2:18 pm
There was actually NO data on the SD card for the dates in question. The card was not locked. Actually something DID get written to the card: the directory entry for the day(s) on which the data was missing.

The machine was definitely on! I used it all night on those nights, and the very high level data (AHI and hours) was available on the touch screen.

I'm completely baffled. Why would it work most days but just write an empty directory on a couple of day?

Mahalo
What dog slobber said below is spot on.

Additionally the blue lights blink rapidly when you insert the SD card. It helps if you wake up the screen on the front and then insert the sd. the blue light blinks rapidly near where you insert it. if there is no error message on the screen then it is probably OK.

it will show the basic AHI and usage and not much else if you don’t have it seated properly.

I have been bummed when I open up the SD card and find I didn’t have any day recorded. Keep trying!

hth

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: Newbie who loves her machine!
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.

User avatar
SleepGeek
Posts: 1055
Joined: Mon Feb 09, 2015 2:56 pm

Re: AirSense 11 doesn't record to SD card once in a while

Post by SleepGeek » Sun Oct 01, 2023 6:14 pm

chipschap wrote:
Sat Sep 30, 2023 10:24 am
I've been using the Resmed Airsense 11 (auto) for a little over two weeks now. On two separate occasions, there was no detailed data on the SD card; Oscar told me "nothing there." All other days are fine.

I thought perhaps the SD card was not properly inserted, but examining the SD card data structure with a file browser I noticed there is a directory entry for the missing days, just no data in the directory.

Has anyone else run into this problem, or have any insights? Mahalo in advance!
Had you seen this data at an earlier point?
If so, is it possible when you removed the card from the computer you didn't use the *eject* function?

Keep tryin. Mahalo

User avatar
chunkyfrog
Posts: 34545
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: AirSense 11 doesn't record to SD card once in a while

Post by chunkyfrog » Sun Oct 01, 2023 6:56 pm

Little green men?

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her