Debris in the Water Tank
-
RANGERYDER2023
- Posts: 5
- Joined: Sun Sep 17, 2023 9:43 pm
Debris in the Water Tank
After reviewing the post about contamination, here is my story.
I have been using a Resmed CPAP machine for 6 years. I have always used "Distilled" water in the tank. I also use a Soclean unit after every use.
For about 4 1/2 years, I never had ANY particulate matter in the tank.
For the last couple of years, I have been using the "after-market" Soclean filters and carbon inserts.
After cleaning my tank, about a week to ten days later, I discovered black "hard" particles in the tank in addition to black "streaks" floating in the water.
It has happened before, that black "fines" from the carbon filter (these were Soclean filters) were installed in the Soclean Machine.
I clean the tank with a bleach cleaning solution and wipe out the black residue left on the stainless steel parts.
My question is, is this residue a mold residue?
On another note, something else changed. I have noticed the gallon bottle of distilled water (usually made by boiling water,) has on the label "water may be processed by reverse osmosis" (note this does not assure you of bateria-free water.)
Has anybody had this residue analyzed?
I have been using a Resmed CPAP machine for 6 years. I have always used "Distilled" water in the tank. I also use a Soclean unit after every use.
For about 4 1/2 years, I never had ANY particulate matter in the tank.
For the last couple of years, I have been using the "after-market" Soclean filters and carbon inserts.
After cleaning my tank, about a week to ten days later, I discovered black "hard" particles in the tank in addition to black "streaks" floating in the water.
It has happened before, that black "fines" from the carbon filter (these were Soclean filters) were installed in the Soclean Machine.
I clean the tank with a bleach cleaning solution and wipe out the black residue left on the stainless steel parts.
My question is, is this residue a mold residue?
On another note, something else changed. I have noticed the gallon bottle of distilled water (usually made by boiling water,) has on the label "water may be processed by reverse osmosis" (note this does not assure you of bateria-free water.)
Has anybody had this residue analyzed?
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Debris in the Water Tank
That is your problem.RANGERYDER2023 wrote: ↑Sun Sep 17, 2023 10:08 pm
. . .
I also use a Soclean unit after every use.
. . .
Ditch the SoClean. It only "cleans" out your wallet.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: Debris in the Water Tank
SC is garbage and you don't want ozone fumes in your lungs! Google the residue from their filters and see if you get any hits.
- ChicagoGranny
- Posts: 15419
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Debris in the Water Tank
It's a SoClean residue. Join the class-action lawsuit.
Sorry that you fell for this worse-than-a-scam. Here is some dated, but still relevant, information - viewtopic.php?f=1&t=176205&p=1311194&hi ... n#p1311194
-
RANGERYDER2023
- Posts: 5
- Joined: Sun Sep 17, 2023 9:43 pm
Re: Debris in the Water Tank
I am amazed, I had not seen any "legal action" against SoClean as I do receive three "class action lawsuits" newsletters. I do know that SoClean has been selling SoClean 2, presumably an improvement on SoClean 1 (original) - however, it too appears to have the same problem. I would like to thank everyone's replies. I will be following this more closely.
On another note, if this is a product (carbon) residue - then maybe the "ozone" does kill the mold and bacteria....
On another note, if this is a product (carbon) residue - then maybe the "ozone" does kill the mold and bacteria....
Re: Debris in the Water Tank
Do you want to make bets re who'll die first - the bacteria or you?
- ChicagoGranny
- Posts: 15419
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Debris in the Water Tank
Julie, macabre
Re: Debris in the Water Tank
While the foam that is used in the Resmed machine is likely formulated to withstand a normal ambient environment for a long time, a daily exposure to concentrated ozone may have the effect of breaking it down and causing pieces of it to flake off, which may possibly be what you are seeing.
_________________
| Machine: DreamStation 2 Auto CPAP Advanced with Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Resmed AirSense 10 Auto with Humidifier |
-
RANGERYDER2023
- Posts: 5
- Joined: Sun Sep 17, 2023 9:43 pm
Re: Debris in the Water Tank
Okay, I'm game. In my Resmed CPAP machine, where would I find black foam? The "air filter is white," and the tank interfaces appear to be a rubber seals.
- Dog Slobber
- Posts: 4270
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Debris in the Water Tank
Here is a teardown video of an Air 10RANGERYDER2023 wrote: ↑Mon Sep 18, 2023 3:06 pmOkay, I'm game. In my Resmed CPAP machine, where would I find black foam? The "air filter is white," and the tank interfaces appear to be a rubber seals.
https://www.youtube.com/watch?v=M9jdc9_WmwM
At about the 6 minute 30 second mark you will see a black foam disk used for sound dammpening.
I'm guessing the ozone has deteriorated it.
If you're handy take it apart inspect the foam disk and remove it.
Throw your SoClean in the garbage, they're shit.
Google "ozone" and "compatibility" you will find charts describing how ozone is incompatible with many materials. Some of which are in your ResMed.
https://spartanwatertreatment.com/ozone ... ity-chart/
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
-
RANGERYDER2023
- Posts: 5
- Joined: Sun Sep 17, 2023 9:43 pm
Re: Debris in the Water Tank
Well, I know what I have to do. The CPAP machine has helped me a lot. The SoClean has worked okay. I do not have a problem with the Ozone it admits, and as long as the litigation is active, I'm going to keep using it.
But, I can replace my Resmed Machine, it is covered under my insurance and it is more than five years old. When it is replaced, I will report back regarding the contamination of the water tank.
But, I can replace my Resmed Machine, it is covered under my insurance and it is more than five years old. When it is replaced, I will report back regarding the contamination of the water tank.
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Debris in the Water Tank
SoClean voids/has voided your Resmed warranty.
Look it up.
Your current machine is TOAST.
Get rid of the damn NOclean before it damages your LUNGS!
Definitely before you get a new machine.
Look it up.
Your current machine is TOAST.
Get rid of the damn NOclean before it damages your LUNGS!
Definitely before you get a new machine.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: Debris in the Water Tank
"emits"RANGERYDER2023 wrote: ↑Mon Sep 18, 2023 8:38 pmWell, I know what I have to do. The CPAP machine has helped me a lot. The SoClean has worked okay. I do not have a problem with the Ozone it admits, and as long as the litigation is active, I'm going to keep using it.
But, I can replace my Resmed Machine, it is covered under my insurance and it is more than five years old. When it is replaced, I will report back regarding the contamination of the water tank.
please, please, please re-think your attitude regarding soclean. it's not "worked okay". it's causing harm and you really need to stop using it.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Debris in the Water Tank
Please be reasonable.
Those people are very clever, but they FOOLED you, and are fooling you still.
Drop a rock on that scam device.
Those people are very clever, but they FOOLED you, and are fooling you still.
Drop a rock on that scam device.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: Debris in the Water Tank
Apart from the CPAP machine which can be replaced when it is damaged from an ozone cleaner, consider your own health.....
https://www.fda.gov/news-events/press-a ... one-gas-or
https://www.fda.gov/news-events/press-a ... one-gas-or
_________________
| Machine: DreamStation 2 Auto CPAP Advanced with Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Resmed AirSense 10 Auto with Humidifier |

