Resmed AutoSet pressure reading

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
arij
Posts: 5
Joined: Thu Oct 11, 2018 9:59 am

Resmed AutoSet pressure reading

Post by arij » Sun Jul 23, 2023 7:36 pm

I recently changed my settings on my Resmed 11 to autoset whereas before I had it in the CPAP mode at a constant pressure. In the morning I look at my summary and read the pressure. Is this pressure the average pressure for the night or the maximum pressure my CPAP reached during my sleep? Thanks in advance.

User avatar
Pugsy
Posts: 65334
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Resmed AutoSet pressure reading

Post by Pugsy » Sun Jul 23, 2023 8:29 pm

The pressure number shown on the LCD screen of the machine is what we call a 95% number.
A 95% number is just a number that reflects that the pressure was AT OR BELOW that number for 95% of the night.
The "or below" part is real important but often not understood completely.

It is NOT an average of the entire night.
It is NOT where you spent 95% of the night either.

If you want averages you have to use OSCAR and get those numbers off the report and you MUST have a SD card in the machine during the night when you are using the machine to get the reports from OSCAR.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
arij
Posts: 5
Joined: Thu Oct 11, 2018 9:59 am

Re: Resmed AutoSet pressure reading

Post by arij » Sun Jul 23, 2023 9:28 pm

Thanks for the explanation. I will get an SD card and download OSCAR.