AirCurve 10 factory default settings?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
BlueDragon
Posts: 563
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

AirCurve 10 factory default settings?

Post by BlueDragon » Sat Jul 15, 2023 1:35 pm

Does anyone know what the factory default settings of the ResMed AirCurve 10 VAuto are for EPAP, IPAP, and PS in VAuto mode?

I'm looking at my SIL's machine and wonder if it is still set to factory defaults.

Thanks!

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

robert-tim
Posts: 31
Joined: Thu Jun 29, 2023 6:29 pm
Location: Portland OR

Re: AirCurve 10 factory default settings?

Post by robert-tim » Sat Jul 15, 2023 4:26 pm

I believe you can revert it to those settings in the menu. I don't know what they are, but it would also interest me.

_________________
MachineMask
Resmed Aircurve 10 Vauto
Dreamwear full-face mask
OSCAR

User avatar
palerider
Posts: 32304
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AirCurve 10 factory default settings?

Post by palerider » Sun Jul 16, 2023 2:42 am

BlueDragon wrote:
Sat Jul 15, 2023 1:35 pm
Does anyone know what the factory default settings of the ResMed AirCurve 10 VAuto are for EPAP, IPAP, and PS in VAuto mode?

I'm looking at my SIL's machine and wonder if it is still set to factory defaults.

Thanks!
Yup, it's the same as the VPAP Auto, in VAuto mode, it's MinEPAP 4, MaxIPAP 25, PS 4. TiMax 2.0 Ti Min 0.3, Trigger and Cycle medium. Ie, 'wide open'.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
BlueDragon
Posts: 563
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: AirCurve 10 factory default settings?

Post by BlueDragon » Sun Jul 16, 2023 3:28 am

Thanks, PR. Looks like my SIL's machine has been adjusted. Seems odd to me with a minEPAP of 8 and PS of 6. I haven't received any data to look at, but those numbers were higher than I expected for an AHI of 22 in the sleep study. Hopefully I will get some data soon and see what is going on.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
palerider
Posts: 32304
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AirCurve 10 factory default settings?

Post by palerider » Sun Jul 16, 2023 2:45 pm

BlueDragon wrote:
Sun Jul 16, 2023 3:28 am
Thanks, PR. Looks like my SIL's machine has been adjusted. Seems odd to me with a minEPAP of 8 and PS of 6. I haven't received any data to look at, but those numbers were higher than I expected for an AHI of 22 in the sleep study. Hopefully I will get some data soon and see what is going on.
Well, bear in mind that "severity" (frequency) of apnea doesn't correlate with pressure needed to alleviate it.

PS of 6 is pretty high for most people, unless they're combating OHS.