Aircurve 10 vauto: adjusting pressure settings

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
robert-tim
Posts: 31
Joined: Thu Jun 29, 2023 6:29 pm
Location: Portland OR

Aircurve 10 vauto: adjusting pressure settings

Post by robert-tim » Thu Jul 13, 2023 10:58 am

I recently purchased an Aircurve 10 used, so no support came with it. In the first two nights of its use, including last night, AHI was 5.7 and 9.6 respectively. The first night was acceptable but the second is a little high.
The only settings that I changed were to make IPAP 15 and EPAP 10. I had reviewed some messages on this site and thought that might be about right.
According to OSCAR, AHI for 4:15 to 4:50 last night was 0 - 1, so it should be possible to achieve better performance. I'm using a mask that I have not used before, so perhaps I'm not wearing it properly. In any case, if this information is enough to indicate that the pressure settings are considerably off, I'll try that as soon as I see the recommendations.

_________________
MachineMask
Resmed Aircurve 10 Vauto
Dreamwear full-face mask
OSCAR

User avatar
zonker
Posts: 11355
Joined: Fri Jun 19, 2015 4:36 pm

Re: Aircurve 10 vauto: adjusting pressure settings

Post by zonker » Thu Jul 13, 2023 11:05 am

robert-tim wrote:
Thu Jul 13, 2023 10:58 am

According to OSCAR,
would behoove you to post those oscar charts here so the forum can look them over.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Miss Emerita
Posts: 3765
Joined: Sun Nov 04, 2018 8:07 pm

Re: Aircurve 10 vauto: adjusting pressure settings

Post by Miss Emerita » Thu Jul 13, 2023 11:46 am

Please do post them! You'll need to use an image-hosting site like Imgur, because the forum is at its capacity for attachments.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

robert-tim
Posts: 31
Joined: Thu Jun 29, 2023 6:29 pm
Location: Portland OR

Re: Aircurve 10 vauto: adjusting pressure settings

Post by robert-tim » Thu Jul 13, 2023 5:38 pm

is this how you do it?
https://imgur.com/XpXihMP

_________________
MachineMask
Resmed Aircurve 10 Vauto
Dreamwear full-face mask
OSCAR

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Aircurve 10 vauto: adjusting pressure settings

Post by palerider » Fri Jul 14, 2023 1:52 am

robert-tim wrote:
Thu Jul 13, 2023 10:58 am
AHI was 5.7 and 9.6 respectively. The first night was acceptable but the second is a little high.[
So was the first night..
robert-tim wrote:
Thu Jul 13, 2023 10:58 am
The only settings that I changed were to make IPAP 15 and EPAP 10. I had reviewed some messages on this site and thought that might be about right.
Not unless you're obese,

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Fri Jul 14, 2023 1:55 am, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Aircurve 10 vauto: adjusting pressure settings

Post by palerider » Fri Jul 14, 2023 1:54 am

robert-tim wrote:
Thu Jul 13, 2023 5:38 pm
is this how you do it?
https://imgur.com/XpXihMP
Close, but not quite. Read through this.

Put your machine in VAuto mode, PS of 3, MinEPAP 10, MaxIPAP 25, and report back.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Miss Emerita
Posts: 3765
Joined: Sun Nov 04, 2018 8:07 pm

Re: Aircurve 10 vauto: adjusting pressure settings

Post by Miss Emerita » Fri Jul 14, 2023 11:12 am

To follow up on palerider's comment, when you post your next chart, please include all and only these graphs:

Events
Flow rate
Pressure (not mask pressure)
Leaks
Flow limitations
Snores.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

robert-tim
Posts: 31
Joined: Thu Jun 29, 2023 6:29 pm
Location: Portland OR

Re: Aircurve 10 vauto: adjusting pressure settings

Post by robert-tim » Sat Jul 15, 2023 9:11 am

I need to find out how to post data from OSCAR that includes only this. Will do.
Last night with these settings AHI was 8.4, but near zero until a potty trip at 2:20 and then high for a couple of hours before settling down again. Something to do with the mask? I hope you can provide some insight after you see the graph.

Image

_________________
MachineMask
Resmed Aircurve 10 Vauto
Dreamwear full-face mask
OSCAR

User avatar
Pugsy
Posts: 65131
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Aircurve 10 vauto: adjusting pressure settings

Post by Pugsy » Sat Jul 15, 2023 10:03 am


_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
ChicagoGranny
Posts: 15389
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Aircurve 10 vauto: adjusting pressure settings

Post by ChicagoGranny » Sat Jul 15, 2023 10:51 am

robert-tim wrote:
Sat Jul 15, 2023 9:11 am
data from OSCAR that includes only this
Please follow the format in Pugsy's link.

robert-tim
Posts: 31
Joined: Thu Jun 29, 2023 6:29 pm
Location: Portland OR

Re: Aircurve 10 vauto: adjusting pressure settings

Post by robert-tim » Sat Jul 15, 2023 5:41 pm

I think I'll see if these anomalous results are repeated before wasting my time and everyone else's. Perhaps I rolled over on the hose or something.

_________________
MachineMask
Resmed Aircurve 10 Vauto
Dreamwear full-face mask
OSCAR

robert-tim
Posts: 31
Joined: Thu Jun 29, 2023 6:29 pm
Location: Portland OR

Re: Aircurve 10 vauto: adjusting pressure settings

Post by robert-tim » Sun Jul 16, 2023 7:20 am

Last night my AHI was 4.2, so I think I'll stick with this equipment and pressure settings unless there's reason to change them in the future. This is 10% of what it would be without treatment (42).

_________________
MachineMask
Resmed Aircurve 10 Vauto
Dreamwear full-face mask
OSCAR

User avatar
Miss Emerita
Posts: 3765
Joined: Sun Nov 04, 2018 8:07 pm

Re: Aircurve 10 vauto: adjusting pressure settings

Post by Miss Emerita » Sun Jul 16, 2023 10:58 am

robert-tim wrote:
Sun Jul 16, 2023 7:20 am
Last night my AHI was 4.2, so I think I'll stick with this equipment and pressure settings unless there's reason to change them in the future. This is 10% of what it would be without treatment (42).
Don't settle for 4.2. That's almost at a level that would be diagnosed as mild apnea. Try again posting a daily chart. In an earlier post, I listed the graphs to include (and don't include any others). I should also have mentioned that you should include the left panel, which contains some potentially useful information.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

robert-tim
Posts: 31
Joined: Thu Jun 29, 2023 6:29 pm
Location: Portland OR

Re: Aircurve 10 vauto: adjusting pressure settings

Post by robert-tim » Mon Jul 17, 2023 3:48 pm

Using the same settings as the previous night, I obtained an AHI of 1.62. Perhaps the mask was not adjusted properly the previous night? I don't know. Anyhow I think this image meets the requirements--in any case it has the values it's supposed to have, and includes the left panel.

Image

_________________
MachineMask
Resmed Aircurve 10 Vauto
Dreamwear full-face mask
OSCAR