Aircurve 10 ASV. Caught CA events, but OSCAR?
Aircurve 10 ASV. Caught CA events, but OSCAR?
Hi there,
I've long story but will cut it short. Second night with ASV. I believe that OSCAR doesn't show CA events, even in UA or LL row. I noted my every arousal time, and check it with OSCAR. One of the example is in photo. My classical CA. Between 10-15 sec. Airsense 10 Autoset can't solve these CAs but at least show them on OSCAR.
If there is anything I miss, pls inform me. It's typically like my CA event, as the IPAP is same EPAP at event time. But OSCAR says, you are perfect and you haven't got CA as it's not showed. Thanks in advance.
Evidence: https://ibb.co/vPRZVCL
I hope I miss something.
I've long story but will cut it short. Second night with ASV. I believe that OSCAR doesn't show CA events, even in UA or LL row. I noted my every arousal time, and check it with OSCAR. One of the example is in photo. My classical CA. Between 10-15 sec. Airsense 10 Autoset can't solve these CAs but at least show them on OSCAR.
If there is anything I miss, pls inform me. It's typically like my CA event, as the IPAP is same EPAP at event time. But OSCAR says, you are perfect and you haven't got CA as it's not showed. Thanks in advance.
Evidence: https://ibb.co/vPRZVCL
I hope I miss something.
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
OSCAR is limited in reporting to data gathered by the machine.
Are you aware that your ResMed AirCurve 10 ASV machine doesn't even flag centrals/ClearAirways.
OSCAR can't report what the machine doesn't do.
Are you aware that your ResMed AirCurve 10 ASV machine doesn't even flag centrals/ClearAirways.
OSCAR can't report what the machine doesn't do.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
I'm aware of it. But lots of body says, if CA occurs, it is in UA or LL. But it seems completely wrong. So what to do? As a result, Aircurve 10 ASV may be one of the best machine for CA but not %100 perfect when it come to solve it? So as I understood, also RESCAN don't flag CAs.
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
No software can report something if the machine doesn't record/flag it.
Now Philips Respironics ASV machine do flag/report CAs if they happen.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- Dog Slobber
- Posts: 4172
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
No, you're obviously not aware of it.
It's not OSCAR that is failing to report, it is the ResMed ASV.
Neither OSCAR, nor Rescan can report what the machine isn't reporting.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
- Respirator99
- Posts: 376
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
Resmed ASV machines do not discriminate between central or obstructive apneas, but flag all recorded apneas as UA (unclassified apnea). They do not classify any apnea as LL.
Oscar, SleepHQ, ResScan and any other similar software will not flag apneas if they are not flagged by the machine. No software is 100% perfect and neither is any machine - that is a hopeless ambition. You've shown us one incident of an apnea that was not flagged (probably due to those tiny wriggles in the flow trace) but without some context it's meaningless. What's your overall chart look like? How many of these events are there? If it's one or two, don't worry about it.But it seems completely wrong. So what to do? As a result, Aircurve 10 ASV may be one of the best machine for CA but not %100 perfect when it come to solve it? So as I understood, also RESCAN don't flag CAs.
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
Apparently not.
I have no idea what that means.
I'd agree what you typed seems wrong,
Stop complaining about things you don't understand?
You can't *HAVE* a CA with an ASV, therefore, it doesn't *ever* flag them.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Respirator99
- Posts: 376
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
I think you can if the PS isn't high enough - that's why Resmed recommend a max PS of not less than 15. So the correct statement would be "You can't *HAVE* a CA with a properly set up ASV". Resmed ASVs flag all apneas as UA.You can't *HAVE* a CA with an ASV, therefore, it doesn't *ever* flag them.
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
Im planning to try an asv in November. So a general question based on the initial post (sorry for crashing in here): Is the machine completely unable to solve short CAs that last only 10-15 seconds? Or is it only a matter of the settings?
Question is improtant to me because all of my CAs only last that long or better "short". If yes, the asv would be useless for me.
Question is improtant to me because all of my CAs only last that long or better "short". If yes, the asv would be useless for me.
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
It's just a matter of the settings.Herbert wrote: ↑Sun Jul 16, 2023 6:02 amIm planning to try an asv in November. So a general question based on the initial post (sorry for crashing in here): Is the machine completely unable to solve short CAs that last only 10-15 seconds? Or is it only a matter of the settings?
Question is important to me because all of my CAs only last that long or better "short". If yes, the asv would be useless for me.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
Are you asking about "resolve" "respond to" or "report"? It doesn't report any specific kind of apneas, they're all "unknown".Herbert wrote: ↑Sun Jul 16, 2023 6:02 amIm planning to try an asv in November. So a general question based on the initial post (sorry for crashing in here): Is the machine completely unable to solve short CAs that last only 10-15 seconds? Or is it only a matter of the settings?
Question is improtant to me because all of my CAs only last that long or better "short". If yes, the asv would be useless for me.
Respirator99 wrote: ↑Sun Jul 09, 2023 4:12 amI think you can if the PS isn't high enough - that's why Resmed recommend a max PS of not less than 15. So the correct statement would be "You can't *HAVE* a CA with a properly set up ASV". Resmed ASVs flag all apneas as UA.You can't *HAVE* a CA with an ASV, therefore, it doesn't *ever* flag them.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
don't know the difference between resolve and respond, but that's what I meant. The reporting topic is clear to me.
If it's only a matter of settings, I assume the "response time" needs to be low so it instantly starts if it doesn't detect a breath. but i assume this will also lead to being forced air in my face in case I hold my breath to turn around. is this assumption basically correct?
If it's only a matter of settings, I assume the "response time" needs to be low so it instantly starts if it doesn't detect a breath. but i assume this will also lead to being forced air in my face in case I hold my breath to turn around. is this assumption basically correct?
Re: Aircurve 10 ASV. Caught CA events, but OSCAR?
I think you are over thinking things here and worrying about facts not in evidence at this time.Herbert wrote: ↑Sun Jul 16, 2023 7:49 amdon't know the difference between resolve and respond, but that's what I meant. The reporting topic is clear to me.
If it's only a matter of settings, I assume the "response time" needs to be low so it instantly starts if it doesn't detect a breath. but i assume this will also lead to being forced air in my face in case I hold my breath to turn around. is this assumption basically correct?
The air "forced in your face" isn't nearly as big of a deal as you are thinking.
You should be asleep for one thing when the ASV forces a breath. The response is rapid and doesn't last but for one breath.
So the increase in pressure is rapid and quickly goes back down. It's one breath only....that increase isn't sustained for treating centrals as it is sustained for preventing OAs. Big difference there.
I am trying to find an example of the actual response with the push of pressure but I don't have one saved in my files on imgur and the ASV reports that I have are on another computer that has a power on problem. It won't turn on.
The push of air goes up quickly with one single breath and goes down just as quickly with that one single breath.
You may or may not even notice that push as you should be asleep. I have used ASV (long story) and the only time I even knew the machine had "treated" a central apnea was when I saw the reports. I slept through them otherwise.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.