High respiration rate?
High respiration rate?
Hello,
In the last 5 years since I started to get treated my situation has systematically and steadily got worse.
I just did a sleep study and the lady was telling me that my high respiration was bad.
The way she said it chilled me.
Check it out here: https://sleephq.com/public/2e76ff7c-dfb ... e4061de212
Or here: https://sleephq.com/public/93d7fe1f-3f5 ... 024f914abf
I have an AirSense set to 12 as base, epr at 3.
I have a mandibular advancement device "The Silencer".
I have a Zeus device (https://zeussleep.co.uk).
An ENT told me 'wow that's a long palate'.
Under endoscopy before a septoplasty I've been told that the back of my nose collapses.
I have on order an Alaxo Stent 6 inch.
My eye bags are getting worse I now have (even with low cholesterol) yellow blotches on my eyelids that are drooping over my eyes.
I wake up feeling like my heart is tired or out of breath.
This will kill me. I'm pretty sure about that. Eventually.
Anyone can share surgical options I can research or any advice at all?
Thank you kindly,
Antoine
In the last 5 years since I started to get treated my situation has systematically and steadily got worse.
I just did a sleep study and the lady was telling me that my high respiration was bad.
The way she said it chilled me.
Check it out here: https://sleephq.com/public/2e76ff7c-dfb ... e4061de212
Or here: https://sleephq.com/public/93d7fe1f-3f5 ... 024f914abf
I have an AirSense set to 12 as base, epr at 3.
I have a mandibular advancement device "The Silencer".
I have a Zeus device (https://zeussleep.co.uk).
An ENT told me 'wow that's a long palate'.
Under endoscopy before a septoplasty I've been told that the back of my nose collapses.
I have on order an Alaxo Stent 6 inch.
My eye bags are getting worse I now have (even with low cholesterol) yellow blotches on my eyelids that are drooping over my eyes.
I wake up feeling like my heart is tired or out of breath.
This will kill me. I'm pretty sure about that. Eventually.
Anyone can share surgical options I can research or any advice at all?
Thank you kindly,
Antoine
_________________
Machine: AirSense™ 10 Elite CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ N20 For Her Nasal CPAP Mask with Headgear |
Re: High respiration rate?
What makes you think your high respiratory rate is caused by your nose?
Does it only happen when you sleep?
Maybe something else that "will kill you" is the cause of that rapid breathing?
Have you had a pulmonary function test? Oxygen saturation monitored professionally for 24 hours?
Can you hide you personal data and upload the sleep study?
I'm not professional, so may be totally ignorant. Can you explain how your rapid breathing is related to the length of your palate? Does the back of your nose collapse on inhalation or on exhalation?
Edited to correct a typo
Does it only happen when you sleep?
Have you consulted a pulmonologist?I wake up feeling like my heart is tired or out of breath.
This will kill me. I'm pretty sure about that. Eventually.
Maybe something else that "will kill you" is the cause of that rapid breathing?
Have you had a pulmonary function test? Oxygen saturation monitored professionally for 24 hours?
I see no signs of obstructed breathing in the graphs you posted only very rapid breathing. Was your sleep study with or without a CPAP?I just did a sleep study and the lady was telling me that my high respiration was bad.
Can you hide you personal data and upload the sleep study?
I'm not professional, so may be totally ignorant. Can you explain how your rapid breathing is related to the length of your palate? Does the back of your nose collapse on inhalation or on exhalation?
Edited to correct a typo
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
Last edited by ozij on Wed Jun 28, 2023 9:05 pm, edited 1 time in total.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
- chunkyfrog
- Posts: 34400
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: High respiration rate?
You mentioned "the lady".
What was her TITLE?
Doctor, nurse, EMT, PA, or technician?
It matters whether or not she knew enough about the subject at hand.
Or was qualified to tell you.
What was her TITLE?
Doctor, nurse, EMT, PA, or technician?
It matters whether or not she knew enough about the subject at hand.
Or was qualified to tell you.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: High respiration rate?
What did your ENT guy suggest? That's what you pay them for.
You got palatal prolapse. See what happens with the Alaxo stents.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: High respiration rate?
Please share you knowledge, Rubicon. Is it the rate of breathing -- and if so, how it that caused by the palatl prolapse? Anything else can we see in the breathing pattern?
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: High respiration rate?
Good point this was only the technician. I'm still waiting for the doc to call.
_________________
Machine: AirSense™ 10 Elite CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ N20 For Her Nasal CPAP Mask with Headgear |
- chunkyfrog
- Posts: 34400
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: High respiration rate?
Technicians are often forbidden to make comments to the patient,
as they are not qualified or licensed to do so.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: High respiration rate?
You mean like other than his ENT told him so?
If you look at the expiratory waveforms, every sleeping waveform (and perhaps some waking waveforms) has some disturbance, ranging from a tiny blip to total obstruction, where expiratory time is cut off @<1.0 second. The rate is high because he's struggling to find someplace to exhale.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: High respiration rate?
Oh yeah?chunkyfrog wrote: ↑Thu Jun 29, 2023 3:13 pmTechnicians are often forbidden to make comments to the patient,
as they are not qualified or licensed to do so.
Watch this!
"You got palatal prolapse. You need either stents or surgery if you expect to get any kind of good sleep."
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: High respiration rate?
Besides, me and Paul told him 4 months ago what he needed on TOOF.Rubicon wrote: ↑Fri Jun 30, 2023 3:48 amYou mean like other than his ENT told him so?
If you look at the expiratory waveforms, every sleeping waveform (and perhaps some waking waveforms) has some disturbance, ranging from a tiny blip to total obstruction, where expiratory time is cut off @<1.0 second. The rate is high because he's struggling to find someplace to exhale.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: High respiration rate?
Thanks!!! I've really learned something now.Rubicon wrote: ↑Fri Jun 30, 2023 3:48 amYou mean like other than his ENT told him so?
If you look at the expiratory waveforms, every sleeping waveform (and perhaps some waking waveforms) has some disturbance, ranging from a tiny blip to total obstruction, where expiratory time is cut off @<1.0 second. The rate is high because he's struggling to find someplace to exhale.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: High respiration rate?
OP originally posted a few years back:
viewtopic.php?f=1&t=177493&p=1328726#p1328726
Browsing now, there was some PP back then. Not a lot.
viewtopic.php?f=1&t=177493&p=1328726#p1328726
Browsing now, there was some PP back then. Not a lot.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: High respiration rate?
Update...
With an ASV machine, an F1B full face mask from BMC (50$ aliexpress), I have much better sleep than CPAP. I'm looking into getting a barbed pharyngoplasty https://www.youtube.com/watch?v=s4DhudXwQbUsince it looks to be the ticket for people with CCC level 2 (complete concentric collapse) which is (I think) what palatal prolapse is.
I tried pretty much all the masks, my pressure goes up to 20 and so that mask is the best I've found for high pressure.
About the nasal stents like the Alaxo... I have the six inch and unless you don't have a gag reflex forget it... when it passes say around four inches in you start gagging and coughing and there's no getting used to that. No amount of will can void your natural reflexes. You literally have to not have a gag reflex to live with that.
I tried medical glue and tape with the mask and it'll help a little but it'll leak anyway and it's a lot of hassle.
Hoping this helps and that others can contribute what surgeries they've had or looking into.
Cheers,
Antoine
With an ASV machine, an F1B full face mask from BMC (50$ aliexpress), I have much better sleep than CPAP. I'm looking into getting a barbed pharyngoplasty https://www.youtube.com/watch?v=s4DhudXwQbUsince it looks to be the ticket for people with CCC level 2 (complete concentric collapse) which is (I think) what palatal prolapse is.
I tried pretty much all the masks, my pressure goes up to 20 and so that mask is the best I've found for high pressure.
About the nasal stents like the Alaxo... I have the six inch and unless you don't have a gag reflex forget it... when it passes say around four inches in you start gagging and coughing and there's no getting used to that. No amount of will can void your natural reflexes. You literally have to not have a gag reflex to live with that.
I tried medical glue and tape with the mask and it'll help a little but it'll leak anyway and it's a lot of hassle.
Hoping this helps and that others can contribute what surgeries they've had or looking into.
Cheers,
Antoine
_________________
Machine: AirSense™ 10 Elite CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ N20 For Her Nasal CPAP Mask with Headgear |
Re: High respiration rate?
1. You can't assume that everyone else will have the same experience that you had, especially since (some/many/most) don't;Shaare wrote: ↑Sun Mar 10, 2024 1:17 amAbout the nasal stents like the Alaxo... I have the six inch and unless you don't have a gag reflex forget it... when it passes say around four inches in you start gagging and coughing and there's no getting used to that. No amount of will can void your natural reflexes. You literally have to not have a gag reflex to live with that.
2. We can't assume that you are even using the stents correctly; and
3. However, thank you for the update.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
- vandownbytheriver
- Posts: 266
- Joined: Sat Feb 03, 2024 11:42 pm
Re: High respiration rate?
I had some palate problems... felt like I was being slapped from inside my head on exhalation. Very position dependent... supine worst, then on left, then few/no symptoms on right side. Interesting that it affected only one side.
Turned out that EPR was exacerbating the problem... the pressure change during breath transitions caused the flapping. Moving to EPR 2 lessened it... moving to EPR 1 pretty much stopped it. I'm now on EPR 0 and have not had a recurrence.
This is not necessarily for the OP but for others whom this may help... Resmed EPR algorithm has known issues with exhalation support, not sure if this is related.
Turned out that EPR was exacerbating the problem... the pressure change during breath transitions caused the flapping. Moving to EPR 2 lessened it... moving to EPR 1 pretty much stopped it. I'm now on EPR 0 and have not had a recurrence.
This is not necessarily for the OP but for others whom this may help... Resmed EPR algorithm has known issues with exhalation support, not sure if this is related.
_________________
Machine: AirCurve 10 Vauto USA C2C CO |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: I use O2Ring, Oscar, SleepHQ, and Cover Roll Stretch mouth tape. |