What would cause these flow limits?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
DaveC
Posts: 26
Joined: Mon Mar 14, 2022 7:57 pm

What would cause these flow limits?

Post by DaveC » Fri May 26, 2023 8:48 pm

I've been using pap therapy for around 16 months. I've always used one FFM or another because I can't seem to keep my big mouth shut. About 4 weeks ago I switched to mouth taping and the ResMed N30i. Life has been good...until last night. I've never had high flow limits. They've always ranged from 0.00 to maybe 0.10. Occasionally they might hit between 0.10 and 0.20 but that's a high as they go. Last night I managed 0.51 and I don't know why.
I was going to post some OSCAR screenshots but it appears the board limit has been reached.

Callen
Posts: 87
Joined: Tue Apr 25, 2023 8:52 pm

Re: What would cause these flow limits?

Post by Callen » Fri May 26, 2023 8:58 pm

Hey Dave, if you'll upload your screenshots to a site like https://imgur.com you'll be able to include the link in your post.

Here is an example https://imgur.com/a/VAm1pgj

EDIT: Pugsy informed me that my format is incorrect; however, uploading to imgur is the correct process. Just follow the sticky! :D
Last edited by Callen on Fri May 26, 2023 10:23 pm, edited 1 time in total.

User avatar
Pugsy
Posts: 65384
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: What would cause these flow limits?

Post by Pugsy » Fri May 26, 2023 9:07 pm


_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

DaveC
Posts: 26
Joined: Mon Mar 14, 2022 7:57 pm

Re: What would cause these flow limits?

Post by DaveC » Fri May 26, 2023 9:10 pm

I just saw the sticky about removing old attachments right before I came back here and saw Pugsy was leading me to it anyway. :D
I haven't been here in a while. It might take me a minute to catch up.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: What would cause these flow limits?

Post by Julie » Sat May 27, 2023 7:50 am

Why do you cut off the AHI reading at the top of the left hand panel?

User avatar
ChicagoGranny
Posts: 15525
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: What would cause these flow limits?

Post by ChicagoGranny » Sat May 27, 2023 10:52 am

DaveC wrote:
Fri May 26, 2023 9:10 pm
What would cause these flow limits?
In the second graph, notice the FLs increase and the machine raises the pressure in response. The machine reaches the set maximum pressure of 20.0 and can go no higher. At higher pressure, it might have cleared up the FLs.

Why did you need so much pressure at this time of night? It could be you had been sleeping on your sides and/or stomach and rolled to your back at this time. Typically, pressure requirements are higher on the back.