Need a new plan for BiPAP / CSA

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
SwirlySauce
Posts: 36
Joined: Sun Mar 14, 2021 5:41 pm

Re: Need a new plan for BiPAP / CSA

Post by SwirlySauce » Sat May 20, 2023 2:00 pm

Rubicon wrote:
Sat May 20, 2023 1:53 pm
Do you have the data "Changes to Machine Settings" from Statistics?
Sure thing.

Image

User avatar
Rubicon
Posts: 1717
Joined: Sun Feb 20, 2022 6:59 am

Re: Need a new plan for BiPAP / CSA

Post by Rubicon » Sat May 20, 2023 2:07 pm

SwirlySauce wrote:
Sat May 20, 2023 1:55 pm
Never took Adderall.
OK sorry-- it looked like you responded
In the beginning there was definitely a boost and pronounced effect. Now it feels like it is more subtle and some days feels like it doesn't do anything at all.
to the guy who was dropping 60 mg IR, but whatever.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

User avatar
Rubicon
Posts: 1717
Joined: Sun Feb 20, 2022 6:59 am

Re: Need a new plan for BiPAP / CSA

Post by Rubicon » Sat May 20, 2023 2:09 pm

Time to punch out anyway. SYL.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

SwirlySauce
Posts: 36
Joined: Sun Mar 14, 2021 5:41 pm

Re: Need a new plan for BiPAP / CSA

Post by SwirlySauce » Sat May 20, 2023 3:03 pm

Rubicon wrote:
Sat May 20, 2023 2:07 pm
SwirlySauce wrote:
Sat May 20, 2023 1:55 pm
Never took Adderall.
OK sorry-- it looked like you responded
In the beginning there was definitely a boost and pronounced effect. Now it feels like it is more subtle and some days feels like it doesn't do anything at all.
to the guy who was dropping 60 mg IR, but whatever.
Oh yes that was me. 60mg IR Adderall is a huge dose, I'm nowhere near that lol. Adderall and Vyvanse are both amphetamine meds, tolerance is a pain for sure
Rubicon wrote:
Sat May 20, 2023 2:09 pm
Time to punch out anyway. SYL.
Have a good one!

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Need a new plan for BiPAP / CSA

Post by palerider » Sat May 20, 2023 10:05 pm

SwirlySauce wrote:
Sat May 20, 2023 1:39 pm
Yup! I guess that's the danger of getting advice on the Internet, everyone is an expert with a different opinion. Am I better off to just go back and see a sleep therapist? Seems so hit and miss from what I've seen with other users but maybe they can get me a PSG?
Well, you could, but why don't we see how your latest settings work for a bit. You've got a much better machine than what you had with the old nightmarestation, the new machine is much more responsive to your needs, and the best thing would probably be to see if you can get good results with what you have.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Need a new plan for BiPAP / CSA

Post by palerider » Sat May 20, 2023 10:57 pm

SwirlySauce wrote:
Sat May 20, 2023 2:00 pm
Rubicon wrote:
Sat May 20, 2023 1:53 pm
Do you have the data "Changes to Machine Settings" from Statistics?
Sure thing.

Image
Let's not make any changes *at all* to your machine after that last one that I suggested, for *at least* a week.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Rubicon
Posts: 1717
Joined: Sun Feb 20, 2022 6:59 am

Re: Need a new plan for BiPAP / CSA

Post by Rubicon » Sun May 21, 2023 3:51 am

SwirlySauce wrote:
Sat May 20, 2023 1:55 pm
Never took Adderall.
SwirlySauce wrote:
Sat May 20, 2023 3:03 pm
Oh yes that was me. 60mg IR Adderall is a huge dose, I'm nowhere near that lol.
Can you clarify this? Are you on or off (that) meth?
Last edited by Rubicon on Sun May 21, 2023 8:17 am, edited 1 time in total.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

User avatar
Rubicon
Posts: 1717
Joined: Sun Feb 20, 2022 6:59 am

Re: Need a new plan for BiPAP / CSA

Post by Rubicon » Sun May 21, 2023 4:04 am

SwirlySauce wrote:
Sat May 20, 2023 3:03 pm
60mg IR Adderall is a huge dose
Yeah I know what MF meth does.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

User avatar
Rubicon
Posts: 1717
Joined: Sun Feb 20, 2022 6:59 am

Re: Need a new plan for BiPAP / CSA

Post by Rubicon » Sun May 21, 2023 4:32 am

Image

First some math. The above sample has 76 breaths. The night that it comes from has ~6630 breaths. The 6 studies you submitted (one was a duplicate) have ~39,780 breaths. Even if one looks at windows with 76 breaths at a time it takes an LFT to look through the stuff. So when those guys at TOF look at SkyLab View for 5 seconds and come up The Answer which includes arbitrarily wingin' them dials around it shouldn't come as much of a surprise that nothing works (OK if somebody posts an OAI of 70 and they're on fixed 5.0 cmH2O or a 95% Leak Rate of 90 L/M they will often stumble onto a worthwhile suggestion).

In answer to your question ResMed has no VB equivalent (one other manufacturer has a Wake Suspension mode I forgot who it's here in the forum somewhere). So if you have crap ResMed will attack it. The above is Wake Crap there's no FLs but RM merrily does algorithm-based auto-DWing. That attack also triggers some wax/wane.

As an aside SleepHQ only puts up one pressure signal so keep in mind what the settings are.

Above from 18 May.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

User avatar
Rubicon
Posts: 1717
Joined: Sun Feb 20, 2022 6:59 am

Re: Need a new plan for BiPAP / CSA

Post by Rubicon » Sun May 21, 2023 5:01 am

Image

These are probablies:

Two isolated FLs F/B pressure attacks (OK, only a few 0.0 cmH2O, but there appears to be disturbances in sleep continuity that immediately follow).

So I believe where we're going here (still need 39.5K breaths reviewed) is

1. Your sleep inherently sucks.
2. Taking crazymeds makes it worse.
3. Poking it with a stick doesn't help.
4. Is attack mode helping or hindering, especially if the events are isolated?

Consider the Jason Premise.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

User avatar
Rubicon
Posts: 1717
Joined: Sun Feb 20, 2022 6:59 am

Re: Need a new plan for BiPAP / CSA

Post by Rubicon » Sun May 21, 2023 5:19 am

Image

If you get your sleep study and it has sleep stages that would be helpful, but absent that we need a surrogate to try to figure out how much you sleep. There are other parameters we can use to fake it, as above. Red blocks are the areas of good sleep. Anything else isn't. Or not sleep at all. IMO.

Can you operate on 3 hours of sleep?

Right. That's why you're here.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

Callen
Posts: 87
Joined: Tue Apr 25, 2023 8:52 pm

Re: Need a new plan for BiPAP / CSA

Post by Callen » Sun May 21, 2023 11:19 am

Rubicon wrote:
Sun May 21, 2023 4:04 am
SwirlySauce wrote:
Sat May 20, 2023 3:03 pm
60mg IR Adderall is a huge dose
Yeah I know what MF meth does.
Stay in your lane. Pharmacology is clearly not in your wheelhouse.

Adderall and Methamphetamine are nowhere near bioequivalents.

Callen
Posts: 87
Joined: Tue Apr 25, 2023 8:52 pm

Re: Need a new plan for BiPAP / CSA

Post by Callen » Sun May 21, 2023 11:24 am

Rubicon wrote:
Sat May 20, 2023 2:07 pm
SwirlySauce wrote:
Sat May 20, 2023 1:55 pm
Never took Adderall.
OK sorry-- it looked like you responded
In the beginning there was definitely a boost and pronounced effect. Now it feels like it is more subtle and some days feels like it doesn't do anything at all.
to the guy who was dropping 60 mg IR, but whatever.
Bruv, I am taking XR (extended-release) not IR (immediate-release). It's 30mg BID (twice daily) for a total of 60mg, so I'll give you partial credit.

You're like Chat GPT 4 sometimes, confidently wrong.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Need a new plan for BiPAP / CSA

Post by palerider » Sun May 21, 2023 1:17 pm

Callen wrote:
Sun May 21, 2023 11:24 am
irrelevancy.
Please don't use your vendetta against Rubicon to derail other people's threads.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 64083
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Need a new plan for BiPAP / CSA

Post by Pugsy » Sun May 21, 2023 1:19 pm

palerider wrote:
Sun May 21, 2023 1:17 pm
Callen wrote:
Sun May 21, 2023 11:24 am
irrelevancy.
Please don't use your vendetta against Rubicon to derail other people's threads.
Amen.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.