My ahi's have been creeping up

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Marcusstephen
Posts: 122
Joined: Tue Feb 05, 2019 9:39 pm
Location: Gold Coast Queensland Australia

My ahi's have been creeping up

Post by Marcusstephen » Tue May 16, 2023 1:36 am

Hi All,
Just wondering if someone could have a quick squiz at my results of late
and tell me what could be causing my ahi's to be creeping up.
Thank you


https://sleephq.com/public/01fafe33-399 ... 3fb3d28d31
BAD BABY

User avatar
Pugsy
Posts: 64933
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: My ahi's have been creeping up

Post by Pugsy » Tue May 16, 2023 5:34 am

How's your sleep quality?
A sizable chunk of those flagged events appear to be SWJ or sleep/wake/junk to my eye.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
ChicagoGranny
Posts: 15085
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: My ahi's have been creeping up

Post by ChicagoGranny » Tue May 16, 2023 8:09 am

Marcusstephen wrote:
Tue May 16, 2023 1:36 am
My ahi's have been creeping up
1.6 AHI

User avatar
Marcusstephen
Posts: 122
Joined: Tue Feb 05, 2019 9:39 pm
Location: Gold Coast Queensland Australia

Re: My ahi's have been creeping up

Post by Marcusstephen » Tue May 16, 2023 3:29 pm

Pugsy wrote:
Tue May 16, 2023 5:34 am
How's your sleep quality?
A sizable chunk of those flagged events appear to be SWJ or sleep/wake/junk to my eye.
Hi Pugsy
you and I had a conversation recently about our “pain and suffering “ 😊 as a consequence of aging bodies specifically, back , hip and shoulders. My quality of sleep is very impacted by pain and staying in one position for too long. My AHI’s have been slowly creeping up but i was wondering if “flow limitations” or something on the chart was indicating some other problem .
Your identification of swj’s is reassuring
BAD BABY

User avatar
ozij
Posts: 10436
Joined: Fri Mar 18, 2005 11:52 pm

Re: My ahi's have been creeping up

Post by ozij » Tue May 16, 2023 7:26 pm

What I see in the chart is a minimum that is letting some snoring and some flow limitations occur.
Then the pressure goes up, you breathing gets better, the pressure drops - and, sure enough, you have flow limitation and snoring again.
Did you always have that?

If not, I'd start looking for slightly higher minimum pressure.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Marcusstephen
Posts: 122
Joined: Tue Feb 05, 2019 9:39 pm
Location: Gold Coast Queensland Australia

Re: My ahi's have been creeping up

Post by Marcusstephen » Wed May 17, 2023 11:11 am

ozij wrote:
Tue May 16, 2023 7:26 pm
What I see in the chart is a minimum that is letting some snoring and some flow limitations occur.
Then the pressure goes up, you breathing gets better, the pressure drops - and, sure enough, you have flow limitation and snoring again.
Did you always have that?

If not, I'd start looking for slightly higher minimum pressure.
Hi Ozij
It’s been getting worse as i’ve progressed in age, and become more sedentary.
I’ve dropped the minimum a bit because i found the mask pressure a bit high requiring tighter tension on the straps.
I guess the only option is to increase minimum pressure
or get fitter and drop off some kilos . Thank you for your input 🙏
BAD BABY