New to Cpap - mask questions

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Maxinelee
Posts: 6
Joined: Tue Mar 21, 2023 10:05 am

New to Cpap - mask questions

Post by Maxinelee » Tue Mar 21, 2023 10:13 am

So I got a resmed 11 a few weeks ago with a resmed f20 full face mask - i was able to sleep with right away it but it was so close to my eyes it started bothering me. yesterday I swapped it for a resmed p30i. I felt like I couldn’t breath when I first put it one and it took a while to get adjusted but then it was ok - I woke up to use the restroom and then the same thing happened - will it get easier or should I look into something else (I taped my mouth shut but chin strap should be here today)? The place I got from gives you one mask and then you can try one more within 30 days - you can’t try on any others. Any advice appreciated

User avatar
Pugsy
Posts: 65148
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New to Cpap - mask questions

Post by Pugsy » Tue Mar 21, 2023 10:22 am

Are you using the ramp feature? Is it on or off? If on, how long?

What are your current settings for pressure, exhale relief, etc.?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Maxinelee
Posts: 6
Joined: Tue Mar 21, 2023 10:05 am

Re: New to Cpap - mask questions

Post by Maxinelee » Tue Mar 21, 2023 11:05 am

Ramp off, EPR full time, epr level 3 cmH2O, APAP, 5.00-20.00

Maxinelee
Posts: 6
Joined: Tue Mar 21, 2023 10:05 am

Re: New to Cpap - mask question/Oscar report help

Post by Maxinelee » Wed Mar 22, 2023 8:31 am

Last night it felt much better - here is Oscar report - can anyone advise how it looks? no idea what I’m looking at Thanks!

sauerkraut
Posts: 51
Joined: Wed Feb 15, 2023 8:00 am
Location: Eastern Central Wisconsin

Re: New to Cpap - mask questions

Post by sauerkraut » Fri Mar 24, 2023 9:58 am

Maxinelee;
There is a mask type setting in the machine ..
Did You change it to the proper setting for Your new mask ??
Best of Luck ..
Steve

_________________
Machine: AirSense 11 Autoset
Additional Comments: F20 mask, Airsense 11 Humidfier .. Oscar ..There is no call out for these in the drop down menu ..

Maxinelee
Posts: 6
Joined: Tue Mar 21, 2023 10:05 am

Re: New to Cpap - mask questions

Post by Maxinelee » Fri Mar 24, 2023 10:14 am

I did switch it - how do I post Oscar picture?

User avatar
Miss Emerita
Posts: 3773
Joined: Sun Nov 04, 2018 8:07 pm

Re: New to Cpap - mask questions

Post by Miss Emerita » Fri Mar 24, 2023 10:49 am

Quite a few people feel air-starved when their pressure is set to 5. In addition, the machine can't go lower than 4, so you don't get your full EPR until the pressure goes up to 7.

But before recommending a change to your settings, we'd love to see that Oscar chart. Take a screenshot and share that. The forum site frequently bumps up against an attachment limit, so if it won't let you attach the screenshot, try using an image hosting service like Imgur and provide a link to your image.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

Maxinelee
Posts: 6
Joined: Tue Mar 21, 2023 10:05 am

Re: New to Cpap - mask questions

Post by Maxinelee » Fri Mar 24, 2023 11:27 am

Oscar feom 3/23 - https://imgur.com/a/euNVo7y

would any other days or screens be helpful? Thanks!!

User avatar
Miss Emerita
Posts: 3773
Joined: Sun Nov 04, 2018 8:07 pm

Re: New to Cpap - mask questions

Post by Miss Emerita » Sat Mar 25, 2023 11:14 am

There's something odd about the way Oscar is reading your SD card, and I don't have the expertise to suggest what you might do about it. I will write a PM to one or two people who might be able to help with that.

In the meantime, I would recommend that you set your minimum to 7 and keep your EPR at 3. I'm betting that will feel a little better. And turn off the ramp.

Your OAs mostly came in a cluster, which suggests you might have a positional issue. Clustered OAs sometimes occur when people who usually sleep on their sides rolls onto their back. They can also occur when people tuck their chins down toward their chests.

Do you use a high pillow, or several pillows? If so, you might benefit from using a single, low and firm pillow, which may help you maintain a good neck/head position. And do you do much sleeping on your back? If yes, no point in tormenting yourself, but if you can try doing more of your sleeping on your sides, that might make for a more restful night for you.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

Maxinelee
Posts: 6
Joined: Tue Mar 21, 2023 10:05 am

Re: New to Cpap - mask questions

Post by Maxinelee » Sat Mar 25, 2023 12:46 pm

Ok thanks! Let me know what to do when you find out 🙂

ozij
Posts: 10492
Joined: Fri Mar 18, 2005 11:52 pm

Re: New to Cpap - mask questions

Post by ozij » Sat Mar 25, 2023 10:17 pm

Miss Emerita wrote:
Sat Mar 25, 2023 11:14 am
There's something odd about the way Oscar is reading your SD card,
At this point, OSCAR reports "device settings unknown" and "PAP mode unknown". Which is almost too bizarre for words. And the other numbers (pressure, ramp time etc.) are "out of this world" as well. A total, thorough, unreliable mess.

There are number of ways or understanding /guessing what happened, e.g. did you by any chance attempt to read the files on the SD by other means?

But I don't think systematic troubleshooting is worth the time here. I'd started by trying to create a clear slate - in the machine and on the SD card - by dong the following steps:

1) Disconnect the power supply from both the wall and the machine. And let it stand like that for 10 minutes or so.
2) While the machine is disconnected, I'd creat a new OSCAR profile.
3) I'd format the SD card in the computer.
4) Connect the APAP to the electricity
5) Reset the APAP to factory settings, wait a while.
6) Make sure wireless is in flight mode, and say you don't want to change it.
7)Insert the SD card, let the machine identify it and reformat it.
8 ) Wait another 5 minutes or so
9) Take the SD card out and read it in the new OSCAR profile on my computer.

If the machine and its settings are identified the next step would be to reinstert the card in the AS11 and then reset the AS11 to your preferred settings.
Use the machine at night and check it the next day to see if the settings were identified and saved.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Miss Emerita
Posts: 3773
Joined: Sun Nov 04, 2018 8:07 pm

Re: New to Cpap - mask questions

Post by Miss Emerita » Sun Mar 26, 2023 11:31 am

Here are suggestions from two other forum members:

(1) Definitely corrupt data. I would suggest to her to delete the day's data:
Data/Advanced/Purge Current Selected Day

Then re-import to see if the odd data comes back.

Importing it into SleepHQ, to see if the strange data is also presented, is a good idea. Seeing the same unreasonable values in SleepHq would confirm the odd values are in the file and not just in this import.

If she's seeing it on other days as well, I would suggest a new SD Card.

(2) Have her load it into SleepHQ and share the link.
Oscar software is available at https://www.sleepfiles.com/OSCAR/