Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
cmpman1974
Posts: 10
Joined: Wed Feb 01, 2023 10:32 pm

Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by cmpman1974 » Wed Feb 01, 2023 10:50 pm

Hi everyone. I'm new to this group. I have done a lot of reading recrently and have been on BiPAP therapy since 2008. My sleep specialist recently upgraded my machine from a regular ResMed Air Curve 10 BiPAP to a newer ResMed Air Curve 10 ST-A unit.

I have to admit I'm very frustrated as I have been 100% compliant all these years using these machines, but still am incredibly fatigued most days and feel like I get poor sleep. I also have struggled greatly trying to get any decent mask fit w/o big leaks. I was looking at my reports with the new machine and for 7 days it has an average mask leak of 65L/hr. My wife wakes up constantly as she says she hears gurgling sounds and whistling when I'm breathing. I use a ResMed Quattro full face mask. It seems like it leaks when I open my jaw at times (from the bottom) and also I have terrible dry mouth upon waking and extremely dry eyes.

Some settings on my new machine are: EPAP: 10.0; Min PS: 10.0; Max PS: 20.0; Ti Min: 0.3s, TI Max: 2.0s,; Target VA: 5.0 L/min; Target PT Rate: 15; IVAPS Mode. A recent report says spontaneous breathing occurred 87% of the time. No Central Apneas were recorded during any sleep studies. I'm morbidly obese with pulmonary hypertension. Recent sleep report for a week showed Exp Pressure of 10.0; Vt: 800 ml; RR: 13; MV 10.8L/min, TI 1.1s; I:E: 1:3.0.

The question I've asked myself for years is am I not getting any real effect from therapy based on the mask leaks? Four times I had to have my doc pull pressure down from 21/11 to 16/8 due to not being able to handle the extreme air pressures. At every follow-up appt, I feel like I'm told new settings or a new machine will be life changing, but NEVER happens. I'm always yawning 200 times a day and can barely make it through a day. 8 hrs of sleep is unrefreshing. Doctor's office says my AHI index is well under 5 with the settings used during polysomnograms. My new machine is claiming AHI of 0 to 1 most nights, but I don't believe that for a second honestly. Makes zero sense to me.

I have struggled a bit to understand why I was moved to the ResMed ST-A machine, but I assume there was good reason as the insurance had strict requirements to quality. My doc made comments about excess CO2 retention due to obesity. All lab work always showed normal 'high range' CO2 levels on the CMP lab test (29-31) and Arterial Blood Gas testing showed normal CO2 levels recently. I understand both these indicators are not reflective of sleep levels.

Any feedback / advice would be immensely appreciated. I hope to learn a lot here.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by palerider » Wed Feb 01, 2023 11:28 pm

Please remember when you're looking for information, Resmed never has, and never will make anything called a "BiPAP", that's a registered trademark of philips.

You said you had some sort of Aircurve (which means it's one of 5 types of bilevel machine) but you didn't say *which* of the machines you had before.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

cmpman1974
Posts: 10
Joined: Wed Feb 01, 2023 10:32 pm

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by cmpman1974 » Wed Feb 01, 2023 11:41 pm

Thanks for the insight. I was not aware of the naming issue/trademark stuff. I've always been prescribed Resmed machines. I will need to look at an old machine to see if I can identify which one it was. It was silver and an Air Curve 10. I'll need to identify further labeling on it and update here. My machine was 7 years old so my physician replaced it.


palerider wrote:
Wed Feb 01, 2023 11:28 pm
Please remember when you're looking for information, Resmed never has, and never will make anything called a "BiPAP", that's a registered trademark of philips.

You said you had some sort of Aircurve (which means it's one of 5 types of bilevel machine) but you didn't say *which* of the machines you had before.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by palerider » Wed Feb 01, 2023 11:55 pm

cmpman1974 wrote:
Wed Feb 01, 2023 11:41 pm
Thanks for the insight. I was not aware of the naming issue/trademark stuff. I've always been prescribed Resmed machines. I will need to look at an old machine to see if I can identify which one it was. It was silver and an Air Curve 10. I'll need to identify further labeling on it and update here. My machine was 7 years old so my physician replaced it.


palerider wrote:
Wed Feb 01, 2023 11:28 pm
Please remember when you're looking for information, Resmed never has, and never will make anything called a "BiPAP", that's a registered trademark of philips.

You said you had some sort of Aircurve (which means it's one of 5 types of bilevel machine) but you didn't say *which* of the machines you had before.
Just look at the name on the front of the machine, underneath "Aircurve 10".

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

cmpman1974
Posts: 10
Joined: Wed Feb 01, 2023 10:32 pm

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by cmpman1974 » Thu Feb 02, 2023 10:22 am

Old machine was Resmed Air Curve 10 VAuto.

cmpman1974
Posts: 10
Joined: Wed Feb 01, 2023 10:32 pm

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by cmpman1974 » Thu Feb 02, 2023 10:38 am

I uploaded information into Oscar for the first time ever. I have a ton to learn on this stuff. From initial glance, my apneas seem pretty controlled which makes me wonder if sleep is not the true reason for all the fatigue. Total average pressures seem to be mostly around 20-21 cm IPAP over 10 cm EPAP. Extremely low AHIs most days. My leak rate last night went WAY down upon moving to a new mask: F&P Vitera. My spouse said she heard no noises unlike previous nights. I always have extremely stuff muscles/joints regardless of reported results on the machine.

User avatar
zonker
Posts: 11337
Joined: Fri Jun 19, 2015 4:36 pm

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by zonker » Thu Feb 02, 2023 12:06 pm

cmpman1974 wrote:
Thu Feb 02, 2023 10:38 am
I uploaded information into Oscar for the first time ever.
welcome to the zoo and good work, you! getting into oscar right off the bat.

now, could you post some charts?

here's some help in that regard-

viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html

if you need any further help with oscar, don't hesitate to ask.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by palerider » Thu Feb 02, 2023 2:05 pm

cmpman1974 wrote:
Thu Feb 02, 2023 10:22 am
Old machine was Resmed Air Curve 10 VAuto.
And what was the 'justification' of switching from a VAuto to a ST-A?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

cmpman1974
Posts: 10
Joined: Wed Feb 01, 2023 10:32 pm

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by cmpman1974 » Thu Feb 02, 2023 4:53 pm

Being overweight and suspecting too high retention of CO2 and daytime fatigue. Highest CO2 in daytime was 32 on lab test and normal range on Arterial Blood Gas analysis. There could very well be more I don't understand that explains justification.
palerider wrote:
Thu Feb 02, 2023 2:05 pm
cmpman1974 wrote:
Thu Feb 02, 2023 10:22 am
Old machine was Resmed Air Curve 10 VAuto.
And what was the 'justification' of switching from a VAuto to a ST-A?

cmpman1974
Posts: 10
Joined: Wed Feb 01, 2023 10:32 pm

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by cmpman1974 » Thu Feb 02, 2023 5:23 pm

Oscar Report.png
Oscar Report.png (38.51 KiB) Viewed 2404 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by palerider » Fri Feb 03, 2023 12:36 am

cmpman1974 wrote:
Thu Feb 02, 2023 4:53 pm
Being overweight and suspecting too high retention of CO2 and daytime fatigue. Highest CO2 in daytime was 32 on lab test and normal range on Arterial Blood Gas analysis. There could very well be more I don't understand that explains justification.
palerider wrote:
Thu Feb 02, 2023 2:05 pm
cmpman1974 wrote:
Thu Feb 02, 2023 10:22 am
Old machine was Resmed Air Curve 10 VAuto.
And what was the 'justification' of switching from a VAuto to a ST-A?
Sounds like BS to me.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

cmpman1974
Posts: 10
Joined: Wed Feb 01, 2023 10:32 pm

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by cmpman1974 » Mon Feb 06, 2023 10:40 am

Does anyone else have additional feedback? Is this a very active group or just a few posters?

User avatar
Pugsy
Posts: 65121
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by Pugsy » Mon Feb 06, 2023 10:46 am

Actually a fairly active forum here but extremely limited on anyone who has any experience with your machine or your problem.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
robysue1
Posts: 1311
Joined: Sun Sep 18, 2022 3:39 pm
Location: Buffalo, NY

Re: Looking for Advice on Resmed Air Curve 10 BiPAP - ST-A

Post by robysue1 » Mon Feb 06, 2023 4:32 pm

cmpman1974 wrote:
Thu Feb 02, 2023 5:23 pm
Oscar Report.png
Those leak numbers are pretty bad and I would think they're also significant.

Not only can large leaks affect the efficacy of therapy directly, but leaks, even small ones, can also negatively affect overall sleep quality.

Any chance you are waking up because of the leaks?
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.

Correct number of posts is 7250 as robysue + what I have as robysue1

Profile pic: Frozen Niagara Falls