Hello,
Thank you to whoever replies to this. I was diagnosed with an AHI of 11, purely hypopneas- no apneas or RERAs were noted during my in-lab sleep study. I have severe symptoms: drowsiness, sleepiness and general body fatigue. This all started about three years ago and has been progressively getting worse and worse. I started using APAP full-time from the 1st of January, before that I was getting acclimated and couldn't use the mask a full night. However, I progressed to now being able to fall asleep and wake up with the mask secure without any issues. I use P10 nasal pillows which I find comfortable.
Regarding APAP settings, my doctor hasn't helped me at all. I had no titration study done. When I was given my diagnosis near the end of last year the doctor simply said "try out a fixed pressure of 4" without any evidence as to why that would help- it didn't. My reported AHI was too high and I couldn't use the CPAP a full night so I changed it to APAP mode, min 4.6 and max 8. EPR 3, full-time. With these settings I'm able to use the APAP throughout the night.I haven't been able to get an appointment with the sleep department since then (NHS).
I'm still very tired, and although it could be that my body is recovering from sleep debt, I have a feeling that something isn't quite right with my CPAP therapy. And I don't have the knowledge to be able to tell what's wrong. Could any of you please help me? OSCAR screenshots below.
Notes:
- Fall asleep and wake up fine with the P10 Nasal pillows
- I have woken up with a dry mouth for the last three days. All the other times I haven't noticed a dry mouth at all.
- I seem to still be waking up during or after dreams, as I have a vivid recall to my dreams.
Current settings:
- Min 4.6
- Max 8
- EPR 3, full-time
Thanks!
Thank you to whoever replies to this. I was diagnosed with an AHI of 11, purely hypopneas- no apneas or RERAs were noted during my in-lab sleep study. I have severe symptoms: drowsiness, sleepiness and general body fatigue. This all started about three years ago and has been progressively getting worse and worse. I started using APAP full-time from the 1st of January, before that I was getting acclimated and couldn't use the mask a full night. However, I progressed to now being able to fall asleep and wake up with the mask secure without any issues. I use P10 nasal pillows which I find comfortable.
Regarding APAP settings, my doctor hasn't helped me at all. I had no titration study done. When I was given my diagnosis near the end of last year the doctor simply said "try out a fixed pressure of 4" without any evidence as to why that would help- it didn't. My reported AHI was too high and I couldn't use the CPAP a full night so I changed it to APAP mode, min 4.6 and max 8. EPR 3, full-time. With these settings I'm able to use the APAP throughout the night.I haven't been able to get an appointment with the sleep department since then (NHS).
I'm still very tired, and although it could be that my body is recovering from sleep debt, I have a feeling that something isn't quite right with my CPAP therapy. And I don't have the knowledge to be able to tell what's wrong. Could any of you please help me? OSCAR screenshots below.
Notes:
- Fall asleep and wake up fine with the P10 Nasal pillows
- I have woken up with a dry mouth for the last three days. All the other times I haven't noticed a dry mouth at all.
- I seem to still be waking up during or after dreams, as I have a vivid recall to my dreams.
Current settings:
- Min 4.6
- Max 8
- EPR 3, full-time
Thanks!
- Attachments
-
- 2.png (131.74 KiB) Viewed 287 times
-
- 5.png (131.84 KiB) Viewed 287 times
-
- 6.png (134.58 KiB) Viewed 287 times
Re: I have been using APAP since New Year and still feel sleepy. Need help with settings
Too Sleepy,
Welcome to the forum.
Congratulations on making the decision to just take things into your own hands and start making reasonable changes to the settings so that you can get some sleep and benefit from your APAP.
You are still waking up several times a night to turn the machine off and back on: I can pick that up from the Sessions data.
On Jan 17 your total time using the machine was under 6 hours, so you may simply have not gotten enough sleep (with the machine) that night.
In other words, from the Oscar data you've posted, it's hard to pinpoint what your feeling that "something isn't quite right" might be related to.
So let me ask you some questions that don't require you to analyze the data in order to answer:
1) Are you getting to sleep comfortably at the beginning of the night when your pressure is close to 4.6cm? If not, describe what's not comfortable as best you can.
2) How many times do you typically wake up during the night? When you wake up, do you know why you're waking up? In particular, are you waking up uncomfortable for any reason?
3) Are you aware of leaks at any point during the night? Or are you surprised when you look at the data in Oscar and the data says you have leaks?
The dry mouth may be related to leaks. Even small leaks that you ordinarily would not need to worry about can cause dry mouth problems.
You can try consciously training your tongue to stay touching the top of your mouth right behind your upper front teeth. This is the normal natural resting spot of the tongue, but when it slips down from there in the night, air can enter the oral cavity from the nasal cavity. Typically when too much air builds up in the mouth, you either get a small mouth leak from air leaking through closed lips OR you open your mouth and get a larger leak OR you swallow the air and (potentially) wind up with aerophagia---i.e. enough air in your stomach to cause discomfort.
You can also try using Biotene or other products designed to help fight dry mouth problems. Xylimelts is another product that is often recommended.
Increasing the humidifier setting can also help with dry mouth problems. If you are currently using a heated hose on Auto temp and Auto humidity, you'll need to switch to Manual in order to select a higher humidity setting. Keep in mind, the higher the humidity setting at a given temperature, the more likely you are to have problems with rainout. Sometimes you just have to deal with a compromise between the humidity the nose & mouth wants and the problems caused by rainout. But most people are able to find a sweet spot where there's enough humidity to prevent the dry nose & mouth problems, but not enough to cause problems with rainout.
Welcome to the forum.
Severity of symptoms is not always correlated with severity of the undiagnosed sleep apnea. But most people who have severe symptoms find the transition to xPAP therapy a bit easier than those of us who didn't have severe symptoms.Too Sleepy wrote: ↑Sun Jan 22, 2023 10:03 amI was diagnosed with an AHI of 11, purely hypopneas- no apneas or RERAs were noted during my in-lab sleep study. I have severe symptoms: drowsiness, sleepiness and general body fatigue. This all started about three years ago and has been progressively getting worse and worse.
So you're about 3 weeks into full time therapy.I started using APAP full-time from the 1st of January, before that I was getting acclimated and couldn't use the mask a full night. However, I progressed to now being able to fall asleep and wake up with the mask secure without any issues. I use P10 nasal pillows which I find comfortable.
Far too many new CPAPers (on both sides of the Atlantic) have an experience similar to yours.Regarding APAP settings, my doctor hasn't helped me at all. I had no titration study done. When I was given my diagnosis near the end of last year the doctor simply said "try out a fixed pressure of 4" without any evidence as to why that would help- it didn't. My reported AHI was too high and I couldn't use the CPAP a full night so I changed it to APAP mode, min 4.6 and max 8. EPR 3, full-time. With these settings I'm able to use the APAP throughout the night.I haven't been able to get an appointment with the sleep department since then (NHS).
Congratulations on making the decision to just take things into your own hands and start making reasonable changes to the settings so that you can get some sleep and benefit from your APAP.
The data that you have posted looks excellent in terms of efficacy, but given how far you've zoomed in on the detailed data for Jan 17 and Jan 20, it's hard to figure out what the overall night's leaks look like or what the pressure was doing on those nights.I'm still very tired, and although it could be that my body is recovering from sleep debt, I have a feeling that something isn't quite right with my CPAP therapy. And I don't have the knowledge to be able to tell what's wrong. Could any of you please help me? OSCAR screenshots below.
You are still waking up several times a night to turn the machine off and back on: I can pick that up from the Sessions data.
On Jan 17 your total time using the machine was under 6 hours, so you may simply have not gotten enough sleep (with the machine) that night.
In other words, from the Oscar data you've posted, it's hard to pinpoint what your feeling that "something isn't quite right" might be related to.
So let me ask you some questions that don't require you to analyze the data in order to answer:
1) Are you getting to sleep comfortably at the beginning of the night when your pressure is close to 4.6cm? If not, describe what's not comfortable as best you can.
2) How many times do you typically wake up during the night? When you wake up, do you know why you're waking up? In particular, are you waking up uncomfortable for any reason?
3) Are you aware of leaks at any point during the night? Or are you surprised when you look at the data in Oscar and the data says you have leaks?
You may be experiencing both sleep rebound and REM rebound: Now that your sleep is not being interrupted by the hypopneas all night, the body may be craving more sleep and spending more time in REM. This should start to resolve in a few more weeks.Notes:
- Fall asleep and wake up fine with the P10 Nasal pillows
- I have woken up with a dry mouth for the last three days. All the other times I haven't noticed a dry mouth at all.
- I seem to still be waking up during or after dreams, as I have a vivid recall to my dreams.
The dry mouth may be related to leaks. Even small leaks that you ordinarily would not need to worry about can cause dry mouth problems.
You can try consciously training your tongue to stay touching the top of your mouth right behind your upper front teeth. This is the normal natural resting spot of the tongue, but when it slips down from there in the night, air can enter the oral cavity from the nasal cavity. Typically when too much air builds up in the mouth, you either get a small mouth leak from air leaking through closed lips OR you open your mouth and get a larger leak OR you swallow the air and (potentially) wind up with aerophagia---i.e. enough air in your stomach to cause discomfort.
You can also try using Biotene or other products designed to help fight dry mouth problems. Xylimelts is another product that is often recommended.
Increasing the humidifier setting can also help with dry mouth problems. If you are currently using a heated hose on Auto temp and Auto humidity, you'll need to switch to Manual in order to select a higher humidity setting. Keep in mind, the higher the humidity setting at a given temperature, the more likely you are to have problems with rainout. Sometimes you just have to deal with a compromise between the humidity the nose & mouth wants and the problems caused by rainout. But most people are able to find a sweet spot where there's enough humidity to prevent the dry nose & mouth problems, but not enough to cause problems with rainout.
_________________
Machine: DreamStation BiPAP® Auto Machine |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: Also use a P10 mask |
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
-
- Posts: 5
- Joined: Sun Jan 22, 2023 9:57 am
Re: I have been using APAP since New Year and still feel sleepy. Need help with settings
Hello robysue1,
Thank you very much for your reply, I really appreciate it.
- Apologies for the zoomed in data, I have attached screenshots of those nights that are zoomed out so you can see the entirety of the night. Do these give a better idea of what's going on?
- Unfortunately, anything to do with humidity and temperature in manual mode causes rainout. Even at the highest tube temperature I experience rainout if I've manually set the humidity. I now have it all set to auto and it's going great. My nose is clear 95% of the time.
1) Yes I am getting to sleep comfortably at night with 4.6cm of pressure. In fact, I've tried to increase the pressure to 4.8cm (my average pressure is 5.3 and I thought it'd be best to get the minimum as close to that as possible) and I can't handle it. It appears that any higher than 4.6cm and I feel like I'm struggling to inhale all that air even with EPR on. In terms of max pressure: previously I set it at 9cm and a few times I woke up during a dream, looked at the CPAP and saw pressure at 9, and concluded that my breathing became more restricted during REM sleep prompting a pressure increase. However, that pressure increase must have woken me up.
2) Before therapy I wouldn't wake up during the night at all. However, the first one to three weeks of therapy there's been times where I would wake up 1-3 times at night. Some of those times I need to pee; that was new to me. I also seem to still have vivid recall of my dreams. However, recently I would say that the wake ups are decreasing. However, I seem to sleeping more. I sleep 6-8 hours, wake up and pee, then sleep another 4-6 hours.
3) I am not aware of any leaks when I sleep. The high leaks in OSCAR surprised me. I have been waking up with dry mouth recently which makes me think that perhaps I am opening my mouth which may explain those leaks. I am going to start mouth-taping from now on and see how that goes. During the day my tongue naturally touches my palate.
Thank you very much for your reply, I really appreciate it.
- Apologies for the zoomed in data, I have attached screenshots of those nights that are zoomed out so you can see the entirety of the night. Do these give a better idea of what's going on?
- Unfortunately, anything to do with humidity and temperature in manual mode causes rainout. Even at the highest tube temperature I experience rainout if I've manually set the humidity. I now have it all set to auto and it's going great. My nose is clear 95% of the time.
1) Yes I am getting to sleep comfortably at night with 4.6cm of pressure. In fact, I've tried to increase the pressure to 4.8cm (my average pressure is 5.3 and I thought it'd be best to get the minimum as close to that as possible) and I can't handle it. It appears that any higher than 4.6cm and I feel like I'm struggling to inhale all that air even with EPR on. In terms of max pressure: previously I set it at 9cm and a few times I woke up during a dream, looked at the CPAP and saw pressure at 9, and concluded that my breathing became more restricted during REM sleep prompting a pressure increase. However, that pressure increase must have woken me up.
2) Before therapy I wouldn't wake up during the night at all. However, the first one to three weeks of therapy there's been times where I would wake up 1-3 times at night. Some of those times I need to pee; that was new to me. I also seem to still have vivid recall of my dreams. However, recently I would say that the wake ups are decreasing. However, I seem to sleeping more. I sleep 6-8 hours, wake up and pee, then sleep another 4-6 hours.
3) I am not aware of any leaks when I sleep. The high leaks in OSCAR surprised me. I have been waking up with dry mouth recently which makes me think that perhaps I am opening my mouth which may explain those leaks. I am going to start mouth-taping from now on and see how that goes. During the day my tongue naturally touches my palate.
-
- Posts: 5
- Joined: Sun Jan 22, 2023 9:57 am
Re: I have been using APAP since New Year and still feel sleepy. Need help with settings
Screenshots here
- Attachments
-
- 19th.png (133.9 KiB) Viewed 272 times
-
- 18th.png (144.16 KiB) Viewed 272 times
-
- 17th of Jan.png (134.46 KiB) Viewed 272 times
-
- Posts: 5
- Joined: Sun Jan 22, 2023 9:57 am
Re: I have been using APAP since New Year and still feel sleepy. Need help with settings
Do the leaks seem like a concern?
Re: I have been using APAP since New Year and still feel sleepy. Need help with settings
are the leaks waking you up?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"the world is full of good people. if you can't find one, be one."-sleepgeek(avatar)
"I'd rather laugh with the sinners than cry with the saints
The sinners are much more fun
You know that only the good die young"-billy joel
"I'd rather laugh with the sinners than cry with the saints
The sinners are much more fun
You know that only the good die young"-billy joel
-
- Posts: 5
- Joined: Sun Jan 22, 2023 9:57 am
Re: I have been using APAP since New Year and still feel sleepy. Need help with settings
Not that I remember, no. But my concern is micro-arousals. Just like how apnea events wake you up without consciously remembering it.
Re: I have been using APAP since New Year and still feel sleepy. Need help with settings
You can't be getting the necessary therapy with that amount of mouth leaking.
And its quite possilble the leaks are causing arousals you aren't conscious of.
And its quite possilble the leaks are causing arousals you aren't conscious of.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Machine: AirSense??? 10 AutoSet??? For Her CPAP Machine with HumidAir??? Heated HumidifierSoftware: Oscar; alternating masks |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Antoine de Saint-Exupery
Re: I have been using APAP since New Year and still feel sleepy. Need help with settings
Well, you see how your pressure line looks like a roller coaster? you're having repeated episodes of breathing problems.
Set your min pressure to 7 and max to 20 and you'll sleep better.